Re: [volt-nuts] fluke 731b battery pack
If even with a fully charged pack the batt charge side gets high enough to feed into the ref you need to check the current value of the current limiting resistor in the charge circuit. Don't have the exact details in my head at the moment (was about 3 years ago) but in both of my 731Bs that resistor had dropped in value by more than a little and was floating the batteries at a higher voltage. I replaced with a correct value over-rated metal film type and no more issue, max floating voltage for ~ C/20 maint to the pack is below cutoff. If it is climbing high enough compared to the other feed, then resistors have drifted and the batt circuit is running to hot or the filter AC -> ref supply is running too low. On 8/3/16 13:30, Orin Eman wrote: > On Wed, Aug 3, 2016 at 2:08 PM, Davidwrote: > >> >> I looked at the schematic and is seems feasible without excessive >> effort. Either the existing simple series preregulator can be >> modified or replaced allowing it to both charge the battery (through a >> blocking diode) and power the instrument or a completely separate >> power charging circuit can be added in parallel. >> > > > The existing charging circuit _is_ in parallel with the pre-regulator; they > are separated by CR5 and CR6 and the pre-regulator/battery outputs are > combined by CR8 and CR9. The only problem with the existing circuit is > with a fresh, strong, fully charged battery pack, charging pulses will leak > past CR8 with only R1/C1 to filter them. > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] fluke 731b battery pack
Most NiMH cells have a capacity of around twice what a NiCad has, so a NiCad charger that works at a C/10 charge rate will be charging NiMH cells at C/20. That can be an acceptable trickle rate for NiMH cells, but I prefer something in the C/40 neighborhood. I put some in an electric shaver (really crude charger) and they last longer than NiCads. Proper low self discharge NiMh cells have about 1/10th the self-discharge rate of NiCads. I have some in some test equipment that seldom gets used and after over a year of storage were still going strong and provided hours of run time. NiCads would be dead after a couple of months of storage. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
If they are integrated with the device, the manufacturer must certify they have tested the device and proved it can not cause a fire. If you change the NiCd battery to a LiFe system, you will fail that requirement. When you ship, you must certify there is no hazmat item in your package. There is a minimum Wh capacity allowed but I am not sure what it is. Not much from memory. - Original Message - From: "David" <davidwh...@gmail.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Wednesday, August 03, 2016 5:29 PM Subject: Re: [volt-nuts] fluke 731b battery pack Does that restriction apply to batteries which are integrated with a device? I thought it only applied to separate cells. On Wed, 03 Aug 2016 17:20:00 -0400, you wrote: I like the LiFe idea except for the restrictions on shipping by air or USPS. That shouldn't be too much of a problem for most that have access to a local cal lab. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
On Wed, 3 Aug 2016 14:30:33 -0700, you wrote: >On Wed, Aug 3, 2016 at 2:08 PM, Davidwrote: > >> >> I looked at the schematic and is seems feasible without excessive >> effort. Either the existing simple series preregulator can be >> modified or replaced allowing it to both charge the battery (through a >> blocking diode) and power the instrument or a completely separate >> power charging circuit can be added in parallel. > >The existing charging circuit _is_ in parallel with the pre-regulator; they >are separated by CR5 and CR6 and the pre-regulator/battery outputs are >combined by CR8 and CR9. The only problem with the existing circuit is >with a fresh, strong, fully charged battery pack, charging pulses will leak >past CR8 with only R1/C1 to filter them. Ah, I did not see that A1S1A is the same as switch A1S1B; I am used to seeing a little dashed line connecting all of the elements of a single switch. I guess that goes along with the symbol they used for constant current diode CR10 which I have not seen in a long time. Anyway, the changes I suggested still apply. Actually, using a battery which accepts a constant voltage charge simplifies the requirements. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
On Wed, Aug 3, 2016 at 2:08 PM, Davidwrote: > > I looked at the schematic and is seems feasible without excessive > effort. Either the existing simple series preregulator can be > modified or replaced allowing it to both charge the battery (through a > blocking diode) and power the instrument or a completely separate > power charging circuit can be added in parallel. > The existing charging circuit _is_ in parallel with the pre-regulator; they are separated by CR5 and CR6 and the pre-regulator/battery outputs are combined by CR8 and CR9. The only problem with the existing circuit is with a fresh, strong, fully charged battery pack, charging pulses will leak past CR8 with only R1/C1 to filter them. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
Does that restriction apply to batteries which are integrated with a device? I thought it only applied to separate cells. On Wed, 03 Aug 2016 17:20:00 -0400, you wrote: >I like the LiFe idea except for the restrictions on shipping by air or USPS. > >That shouldn't be too much of a problem for most that have access to a local >cal lab. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
I like the LiFe idea except for the restrictions on shipping by air or USPS. That shouldn't be too much of a problem for most that have access to a local cal lab. - Original Message - From: "David" <davidwh...@gmail.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Wednesday, August 03, 2016 5:08 PM Subject: Re: [volt-nuts] fluke 731b battery pack On Wed, 3 Aug 2016 16:34:31 -0400, you wrote: Eric wrote: NiMH cells seem to work in the 731b so I'd prefer to use them over NiCd because of the higher capacity and lower toxicity of potential leaks. Is there a downside to doing this? In my experience, good NiCDs are preferable to NiMHs. Good NiCds have substantially lower self-leakage than NiMHs (this is true even of the "NEW!! Low-discharge!!" NiMHs). NiCds also don't degrade nearly as fast if they are left too long on trickle charge. Both of these features translate directly into increased life for the NiCds. NiCds are also quieter, and capable of larger current drains, due to their lower internal (series) resistance (high current is not really an issue in your application). That is my experience as well however I do like the better low discharge NiMH cells. Some are better than other though. Consumer-type NiCds may suffer from the problem PHK noted (poor quality due to low production volumes), but there are still many industrial and military applications that specify NiCDs for some or all of the reasons given above. Excellent NiCds are readily available -- just look for aerospace-grade parts rather than consumer batteries. Low leakage and high temperature NiCd cells seem to last a lot longer than other types and you will not find any consumer versions of these but the price is high. Many built-in charging circuits are crude and leave the batteries on a trickle current that is really too high, particularly given the temperatures inside electronic instruments. So whichever batteries you choose, plan on redesigning the charging circuit. That brings up the possibility of using either LiFePO4 or SLA (sealed lead-acid) batteries -- if you have to redesign the charging circuit anyway, you can just as easily design it for LiFePO4 or SLA. I have not evaluated the 731B power supply in particular, but LiFePO4 would be my presumptive choice unless I encountered an insurmountable obstacle. Best regards, Charles I looked at the schematic and is seems feasible without excessive effort. Either the existing simple series preregulator can be modified or replaced allowing it to both charge the battery (through a blocking diode) and power the instrument or a completely separate power charging circuit can be added in parallel. The difficulty of maintaining charge in a backup application using NiMH cells would lead me to consider LiFePO4 cells instead. The only serious difficulty would be preventing excessive discharge which will ruin a lithium (or PbSO4) based rechargeable battery in short order. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
The battery in the 731B is charged via a diode and 510 ohm resistor from the rectified DC, before the smoothing capacitor. The smoothing capacitor and power supply is fed by another diode, so the battery is charged with a pulse waveform. If you have no battery, then this pulse charge waveform will get around the power supply regulator (since the regulated supply and battery are diode switched) and increase noise on the output. I therefore don't recommend running without a battery installed... (though you could remove CR5 or or R30). Even with the battery installed, if it's good, at full charge, you can still get charging pulses bypassing the regulator. I built a high-side switch to switch between the battery and power supply on my 731B. I got the local Batteries Plus to build a new battery pack. They usually have NiCd cells of the right size and will build a battery pack while you wait. With an NiCd pack so easily available, I saw no reason to change battery type. Orin. On Wed, Aug 3, 2016 at 1:34 PM, Charles Steinmetzwrote: > Eric wrote: > > NiMH cells seem to work in the 731b so I'd prefer to use them over NiCd >> because of the higher capacity and lower toxicity of potential leaks. Is >> there a downside to doing this? >> > > In my experience, good NiCDs are preferable to NiMHs. Good NiCds have > substantially lower self-leakage than NiMHs (this is true even of the > "NEW!! Low-discharge!!" NiMHs). NiCds also don't degrade nearly as fast if > they are left too long on trickle charge. Both of these features > translate directly into increased life for the NiCds. NiCds are also > quieter, and capable of larger current drains, due to their lower internal > (series) resistance (high current is not really an issue in your > application). > > Consumer-type NiCds may suffer from the problem PHK noted (poor quality > due to low production volumes), but there are still many industrial and > military applications that specify NiCDs for some or all of the reasons > given above. Excellent NiCds are readily available -- just look for > aerospace-grade parts rather than consumer batteries. > > Many built-in charging circuits are crude and leave the batteries on a > trickle current that is really too high, particularly given the > temperatures inside electronic instruments. So whichever batteries you > choose, plan on redesigning the charging circuit. > > That brings up the possibility of using either LiFePO4 or SLA (sealed > lead-acid) batteries -- if you have to redesign the charging circuit > anyway, you can just as easily design it for LiFePO4 or SLA. > > I have not evaluated the 731B power supply in particular, but LiFePO4 > would be my presumptive choice unless I encountered an insurmountable > obstacle. > > Best regards, > > Charles > > > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
On Wed, 3 Aug 2016 16:34:31 -0400, you wrote: >Eric wrote: > >> NiMH cells seem to work in the 731b so I'd prefer to use them over NiCd >> because of the higher capacity and lower toxicity of potential leaks. Is >> there a downside to doing this? > >In my experience, good NiCDs are preferable to NiMHs. Good NiCds have >substantially lower self-leakage than NiMHs (this is true even of the >"NEW!! Low-discharge!!" NiMHs). NiCds also don't degrade nearly as fast >if they are left too long on trickle charge. Both of these features >translate directly into increased life for the NiCds. NiCds are also >quieter, and capable of larger current drains, due to their lower >internal (series) resistance (high current is not really an issue in >your application). That is my experience as well however I do like the better low discharge NiMH cells. Some are better than other though. >Consumer-type NiCds may suffer from the problem PHK noted (poor quality >due to low production volumes), but there are still many industrial and >military applications that specify NiCDs for some or all of the reasons >given above. Excellent NiCds are readily available -- just look for >aerospace-grade parts rather than consumer batteries. Low leakage and high temperature NiCd cells seem to last a lot longer than other types and you will not find any consumer versions of these but the price is high. >Many built-in charging circuits are crude and leave the batteries on a >trickle current that is really too high, particularly given the >temperatures inside electronic instruments. So whichever batteries you >choose, plan on redesigning the charging circuit. > >That brings up the possibility of using either LiFePO4 or SLA (sealed >lead-acid) batteries -- if you have to redesign the charging circuit >anyway, you can just as easily design it for LiFePO4 or SLA. > >I have not evaluated the 731B power supply in particular, but LiFePO4 >would be my presumptive choice unless I encountered an insurmountable >obstacle. > >Best regards, > >Charles I looked at the schematic and is seems feasible without excessive effort. Either the existing simple series preregulator can be modified or replaced allowing it to both charge the battery (through a blocking diode) and power the instrument or a completely separate power charging circuit can be added in parallel. The difficulty of maintaining charge in a backup application using NiMH cells would lead me to consider LiFePO4 cells instead. The only serious difficulty would be preventing excessive discharge which will ruin a lithium (or PbSO4) based rechargeable battery in short order. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
NiMH does not do too well in float charge conditions and that is the main state a 731B will spend its time. Regards, Tom - Original Message - From: "Eric Garner" <garn...@gmail.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Wednesday, August 03, 2016 1:27 PM Subject: [volt-nuts] fluke 731b battery pack So I managed to get one of the Fluke 731b's mentioned on the list a while back. It was a little more chewed up than the description in the auction led me to believe: https://plus.google.com/u/0/112871219575380932919/posts/iDWyzwYYRdj?pid=6314634305601690930=112871219575380932919 but it appears to be functional. The NiCd pack appears to have leaked some time in the past and was removed, which is fine since I was going to replace the batteries anyway. Looking through the archives, replacement battery packs for the 731b has been a popular topic but I have some questions that remain open. 1. NiMH cells seem to work in the 731b so I'd prefer to use them over NiCd because of the higher capacity and lower toxicity of potential leaks. Is there a downside to doing this? 2. If I go with NiMH, is there a manufacturer that's prefered? reading reviews of NiMH cells online leads me to believe that some manufacturers (Tenergy) have lower quality and are less reliable. 3. If I go with NiCd would you expect problems if I used larger cells (4/5A vs 2/3A) thanks --Eric _ Eric Garner ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] fluke 731b battery pack
Are the low discharge NiMH cells available now any more rugged than regular NiCd cells as far as charging current? One solution I would consider is using LiFePO4 cells and modifying the charger for current limited constant voltage. On Wed, 03 Aug 2016 19:37:12 +, you wrote: > >In message
Re: [volt-nuts] fluke 731b battery pack
Eric: I have been using NiMH AA batteries in my 731B for at least 15 years now. I am sure that the specs say don't use them in a circuit designed for NiCd but I haven't had any problems with over heating or overcharging. You have to do some mods to the battery mounting to get them in there but once done you are good for a lifetime. I have never checked to see how long the NiMH batteries would run for but my guess would be longer than the original NiCds. Bill - Original Message - From: "Eric Garner" <garn...@gmail.com> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Wednesday, August 03, 2016 10:27 AM Subject: [volt-nuts] fluke 731b battery pack > So I managed to get one of the Fluke 731b's mentioned on the list a while > back. It was a little more chewed up than the description in the auction > led me to believe: > > https://plus.google.com/u/0/112871219575380932919/posts/iDWyzwYYRdj?pid=6314634305601690930=112871219575380932919 > > but it appears to be functional. > > The NiCd pack appears to have leaked some time in the past and was removed, > which is fine since I was going to replace the batteries anyway. > > Looking through the archives, replacement battery packs for the 731b has > been a popular topic but I have some questions that remain open. > > 1. NiMH cells seem to work in the 731b so I'd prefer to use them over NiCd > because of the higher capacity and lower toxicity of potential leaks. Is > there a downside to doing this? > > 2. If I go with NiMH, is there a manufacturer that's prefered? reading > reviews of NiMH cells online leads me to believe that some manufacturers > (Tenergy) have lower quality and are less reliable. > > 3. If I go with NiCd would you expect problems if I used larger cells (4/5A > vs 2/3A) > > > > > thanks > > --Eric > _ > Eric Garner > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] fluke 731b battery pack
So I managed to get one of the Fluke 731b's mentioned on the list a while back. It was a little more chewed up than the description in the auction led me to believe: https://plus.google.com/u/0/112871219575380932919/posts/iDWyzwYYRdj?pid=6314634305601690930=112871219575380932919 but it appears to be functional. The NiCd pack appears to have leaked some time in the past and was removed, which is fine since I was going to replace the batteries anyway. Looking through the archives, replacement battery packs for the 731b has been a popular topic but I have some questions that remain open. 1. NiMH cells seem to work in the 731b so I'd prefer to use them over NiCd because of the higher capacity and lower toxicity of potential leaks. Is there a downside to doing this? 2. If I go with NiMH, is there a manufacturer that's prefered? reading reviews of NiMH cells online leads me to believe that some manufacturers (Tenergy) have lower quality and are less reliable. 3. If I go with NiCd would you expect problems if I used larger cells (4/5A vs 2/3A) thanks --Eric _ Eric Garner ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Fluke 731b Rack Mounting
Hello All, I just picked up a 731b that I would like to mount in a rack with a back up UPS. I know that Fluke made the "Rack Mounting Kit M03-205-605" These were used to house up to four of these standards in the space of 3U, I think. The present day version of this is the Fluke 734a-7001. Does anyone know where to find one of these older rack options or know if the present day 734a would fit the older 731b standards? Cheers, David ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 731B
I would never adjust any pots on a working fluke voltage standard. When the pot moves you introduce more drift. Just keep track of differences. By the same token I do not turn off my 3458As as long as the line power is up. There was a time I had a rack of 15 3458As running waiting to be sold. We have a power outage about 1 time per year long enough to trip our julian date and time clock. We use it to track power up time. I do not see anyone posting 3458A in stat mode data: mean, SDEV, High, and Low, Range, and Number of samples over extended time. then reverse polarity and do it again. Also DC cal values before and after Auto cal and Keysight cal would be nice to know. Also using Before and after cal data from Keysight to correct values. Do not adjust the Flukes but keep a spreadsheet of best guess of its value. Something like mean of present guess + before cal data correction + after cal data correction. Example: Say you present guess (guess1) is 9.468 cal the meter using 9.468 and then read the cal var 1 (I think that is the one for the reference board) Your as found data should indicate how far off your reference is. correct and call that guess2 After you get the meter back read cal var 1 and then use successive approximation to when caling to get the same value in var 1. That should be guess3 So new guess is mean (guess1, guess2, guess3). If you have better confidence in one guess or less confidence in an other the values could be weighted. You never really know what the actual value is a given time but do know it is changing. You can guess at it's value like 9.468 is just a guess even if the real value is 9.4681 or even 9.468 with what we have you just cannot know for sure. There can alway be another digit. Anyone want to poke some holes in what I wrote? Does anyone have a better way to come up with a Cal value for a fluke 10 volt reference without sending it out for calibration? Where could send it to get a better calibration and what would it cost? On Thu, Mar 26, 2015 at 8:03 PM, J. L. Trantham jlt...@att.net wrote: A couple of years ago, I sat my 735C, a (at that time) recently Agilent calibrated 3458A, and two 731B's on a bench and have left the references on, with battery back-up for any power failure. I have turned the 3458A off and on a number of times in the interim. When I set all this up, I adjusted the 735C to 10.00 VDC as best I could using the then recently calibrated 3458A and adjusted the 731B's to match the 735C using the 3458A as a 'null' meter. Tonight, after the 3458A had been on for several weeks, I did the 'AutoCal All' on the 3458A and read each of the references. The 735C measures about 10.16 VDC, and the 731B's measure about 10.05 VDC each, within about 2 uV of each other. In the next couple of months, I plan to send the 3458A back to Agilent/Keysight for recalibration and see what the meter measures 'as received' and 'as left'. We'll see how close it is to being 'in tolerance' 'as received'. Hope this helps. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Orin Eman Sent: Monday, March 23, 2015 12:36 AM To: Discussion of precise voltage measurement Cc: richiem5...@gmail.com Subject: Re: [volt-nuts] Fluke 731B Richard, Some really good information has already been posted. Here is what I have found. I posted a comparison of the 731B against the Geller Labs SVR-T here: http://www.eevblog.com/forum/testgear/geller-labs-svr-and-svr-t/msg282582/#msg282582 There were within 1ppm of each other over 48 hours and 3.8 deg K. I'm currently running a comparison against the reference from an HP 3458A, using an Agilent 34461A in _ratio_ mode. A screen capture follows if the list lets it through. There is only 5 hours here as I just updated the 34461A firmware (it adds another digit to the scaling limits on the trend chart and is definitely worthwhile). One thing about the 731B is that it if the NiCd pack is dead or fully charged(!), there will be much more noise on the output. I designed a 'high-side' switch to switch between AC and battery power and it improves noise when the NiCd is fully charged. (The original design is a diode switch and if you have a NiCd pack which has a high voltage when fully charged, the half-wave rectified AC that is used to charge the NiCd pack leaks through to the voltage reference power supply.) A big WARNING. If a 731B still has its NiCd pack installed, replace it. I had Batteries Plus make me a replacement pack. About $30 as I recall. If the NiCd pack isn't installed, you need to remove the diode that goes from the battery pack to the voltage reference supply, or you'll get unfiltered half-wave rectified AC on your voltage reference power supply and it does get into the output! My 731B seems to be doing quite well IMO, being within 1ppm of the new at the time Geller SVR-T. It has also been
Re: [volt-nuts] Fluke 731B
John, Thanks for the thoughts. I basically agree with the idea of not 'adjusting' the standards but when I got them, they all had 'problems', including leaking NiCd batteries that required clean up. Not knowing how long they had been off and having only calibrated DMM's (3458A and 7081) I chose to get the standards up and running then 'adjust' them to begin a 'new era' of standards, at least in my set-up. I also have two 732A's that have been brought back to life and they are sitting on now for about 12 to 18 months. I have not tried to read them in a while. I have been collecting other gear including Fluke 720A, 721A, 750A, and 752A and a Fluke 845AB. I also have a Fluke 887AB that I need to get working. My problem is that work is about 80 to 100 hours per week and I have little time to spend on this. But, once retired. I have left my 3458A's on for extended periods of time (weeks) and have been very pleased that when I turn them back on after weeks off, they come back within a couple of uV to the last readings of the 735C suggesting everything is fairly stable or at least drifting at the same rate and in the same direction. I'm a bit nervous about leaving the 3458A's on all the time due to the potential for the displays to fade or fail. So far, that's not been a problem. As far as calibration of the Fluke references, I would think Fluke would be the best bet but there is the problem of getting them there and back while still powered. I've been working on a method to do that using SLA batteries or using the Fluke 732A-7003 battery charger/pack. I think Fluke made a case that would house the 732A and the 732A-7003, both filled with batteries and completely charged, to ship the units in for calibration. My plan is to construct a 'case' to house two 12V 18AH SLA's and a charger that can be plugged in once the unit arrives at the calibration facility. I have acquired a couple of dead, industrial strength, UPS's that, once I get some time, I would like to rebuild and power my 'metrology section' with to deal with any power interruptions. I have a stand-by generator system as well. However, it all takes time to get up and running and I don't have much of that right now. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of John Phillips Sent: Friday, March 27, 2015 1:41 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Fluke 731B I would never adjust any pots on a working fluke voltage standard. When the pot moves you introduce more drift. Just keep track of differences. By the same token I do not turn off my 3458As as long as the line power is up. There was a time I had a rack of 15 3458As running waiting to be sold. We have a power outage about 1 time per year long enough to trip our julian date and time clock. We use it to track power up time. I do not see anyone posting 3458A in stat mode data: mean, SDEV, High, and Low, Range, and Number of samples over extended time. then reverse polarity and do it again. Also DC cal values before and after Auto cal and Keysight cal would be nice to know. Also using Before and after cal data from Keysight to correct values. Do not adjust the Flukes but keep a spreadsheet of best guess of its value. Something like mean of present guess + before cal data correction + after cal data correction. Example: Say you present guess (guess1) is 9.468 cal the meter using 9.468 and then read the cal var 1 (I think that is the one for the reference board) Your as found data should indicate how far off your reference is. correct and call that guess2 After you get the meter back read cal var 1 and then use successive approximation to when caling to get the same value in var 1. That should be guess3 So new guess is mean (guess1, guess2, guess3). If you have better confidence in one guess or less confidence in an other the values could be weighted. You never really know what the actual value is a given time but do know it is changing. You can guess at it's value like 9.468 is just a guess even if the real value is 9.4681 or even 9.468 with what we have you just cannot know for sure. There can alway be another digit. Anyone want to poke some holes in what I wrote? Does anyone have a better way to come up with a Cal value for a fluke 10 volt reference without sending it out for calibration? Where could send it to get a better calibration and what would it cost? On Thu, Mar 26, 2015 at 8:03 PM, J. L. Trantham jlt...@att.net wrote: A couple of years ago, I sat my 735C, a (at that time) recently Agilent calibrated 3458A, and two 731B's on a bench and have left the references on, with battery back-up for any power failure. I have turned the 3458A off and on a number of times in the interim. When I set all this up, I adjusted the 735C to 10.00 VDC as best I could using the then recently calibrated 3458A and adjusted
Re: [volt-nuts] Fluke 731B
A couple of years ago, I sat my 735C, a (at that time) recently Agilent calibrated 3458A, and two 731B's on a bench and have left the references on, with battery back-up for any power failure. I have turned the 3458A off and on a number of times in the interim. When I set all this up, I adjusted the 735C to 10.00 VDC as best I could using the then recently calibrated 3458A and adjusted the 731B's to match the 735C using the 3458A as a 'null' meter. Tonight, after the 3458A had been on for several weeks, I did the 'AutoCal All' on the 3458A and read each of the references. The 735C measures about 10.16 VDC, and the 731B's measure about 10.05 VDC each, within about 2 uV of each other. In the next couple of months, I plan to send the 3458A back to Agilent/Keysight for recalibration and see what the meter measures 'as received' and 'as left'. We'll see how close it is to being 'in tolerance' 'as received'. Hope this helps. Joe -Original Message- From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Orin Eman Sent: Monday, March 23, 2015 12:36 AM To: Discussion of precise voltage measurement Cc: richiem5...@gmail.com Subject: Re: [volt-nuts] Fluke 731B Richard, Some really good information has already been posted. Here is what I have found. I posted a comparison of the 731B against the Geller Labs SVR-T here: http://www.eevblog.com/forum/testgear/geller-labs-svr-and-svr-t/msg282582/#msg282582 There were within 1ppm of each other over 48 hours and 3.8 deg K. I'm currently running a comparison against the reference from an HP 3458A, using an Agilent 34461A in _ratio_ mode. A screen capture follows if the list lets it through. There is only 5 hours here as I just updated the 34461A firmware (it adds another digit to the scaling limits on the trend chart and is definitely worthwhile). One thing about the 731B is that it if the NiCd pack is dead or fully charged(!), there will be much more noise on the output. I designed a 'high-side' switch to switch between AC and battery power and it improves noise when the NiCd is fully charged. (The original design is a diode switch and if you have a NiCd pack which has a high voltage when fully charged, the half-wave rectified AC that is used to charge the NiCd pack leaks through to the voltage reference power supply.) A big WARNING. If a 731B still has its NiCd pack installed, replace it. I had Batteries Plus make me a replacement pack. About $30 as I recall. If the NiCd pack isn't installed, you need to remove the diode that goes from the battery pack to the voltage reference supply, or you'll get unfiltered half-wave rectified AC on your voltage reference power supply and it does get into the output! My 731B seems to be doing quite well IMO, being within 1ppm of the new at the time Geller SVR-T. It has also been reading 9.6 to 9.7 on the 34461A for the last year (as long as I don't turn on the 8568A that is next to the 34461A... the 8568A blows warm air into the 34461A and then all bets are off). Orin. [image: Inline image 1] On Sat, Mar 21, 2015 at 1:02 PM, Richard Moore richiem5...@gmail.com wrote: Hi all — realisitcally, what sort of stability can be expected from a fully functional 731B regarding variation in ambient temp a few degrees C around 23, and for long-term drift? Fluke’s specs are very conservative, and I think the 731 is far better than the specs would lead you to believe, but I have no practical experience with these units. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 731B
Richard, Some really good information has already been posted. Here is what I have found. I posted a comparison of the 731B against the Geller Labs SVR-T here: http://www.eevblog.com/forum/testgear/geller-labs-svr-and-svr-t/msg282582/#msg282582 There were within 1ppm of each other over 48 hours and 3.8 deg K. I'm currently running a comparison against the reference from an HP 3458A, using an Agilent 34461A in _ratio_ mode. A screen capture follows if the list lets it through. There is only 5 hours here as I just updated the 34461A firmware (it adds another digit to the scaling limits on the trend chart and is definitely worthwhile). One thing about the 731B is that it if the NiCd pack is dead or fully charged(!), there will be much more noise on the output. I designed a 'high-side' switch to switch between AC and battery power and it improves noise when the NiCd is fully charged. (The original design is a diode switch and if you have a NiCd pack which has a high voltage when fully charged, the half-wave rectified AC that is used to charge the NiCd pack leaks through to the voltage reference power supply.) A big WARNING. If a 731B still has its NiCd pack installed, replace it. I had Batteries Plus make me a replacement pack. About $30 as I recall. If the NiCd pack isn't installed, you need to remove the diode that goes from the battery pack to the voltage reference supply, or you'll get unfiltered half-wave rectified AC on your voltage reference power supply and it does get into the output! My 731B seems to be doing quite well IMO, being within 1ppm of the new at the time Geller SVR-T. It has also been reading 9.6 to 9.7 on the 34461A for the last year (as long as I don't turn on the 8568A that is next to the 34461A... the 8568A blows warm air into the 34461A and then all bets are off). Orin. [image: Inline image 1] On Sat, Mar 21, 2015 at 1:02 PM, Richard Moore richiem5...@gmail.com wrote: Hi all — realisitcally, what sort of stability can be expected from a fully functional 731B regarding variation in ambient temp a few degrees C around 23, and for long-term drift? Fluke’s specs are very conservative, and I think the 731 is far better than the specs would lead you to believe, but I have no practical experience with these units. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] Fluke 731B power switch replacement: TE Connectivity / Alcoswitch MTL206N
Original was MSTL206N, count not cross-ref the maker. TE MTL206N is an exact match drop-in, same locking toggle with same shape/size everything. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 731B parts xref
Hi Fred, Well, it seems to me that the only time the relay is an issue is if the coarse adjust pot is not fully clockwise. If you look at the schematic, the relay is bypassed with R101 (coarse) is in the full CW position (a detented switch position) actuating S101. Near as I can tell, the coarse CW position completely bypasses the protection circuit. So, you should not need batteries. The coarse/fine controls are used when you have a fixed voltage source. As you indicate you have 332 then you can adjust it and do not need the coarse/fine function. Of course you would be responsible for making sure of INPUT / OUTPUT switches are set to the right scales for the voltages being used. But let me make the standard disclaimer - - - It has been a long time since I used my 750A - - - so I could be wrong ! Please let me know if I am all wet. BillWB6BNQ Fred Schneider wrote: I have a 750 too but it needs some rare 6 V batterys to make -6-0-+6V. The adaption to 10V sounds interesting but i do have standard cells and like to use it for that. I do not use the 750 at the moment because without the batterys it won't work. You have an alernative solution for that too ? I measure the cells with a 10 GOhm 7,5 digit meter, write down the reading, then zero the meter in the 1V range with that, set my 720 to that value and connect the meter, then connect my 332 and set the 10V or 100V so the meter reads zero again. And finaly undo the zero and measure the output of the 720 again to heck if the reading is the same as I wrote down. Fred PA4TIM Op 22 nov. 2012 om 07:43 heeft WB6BNQ wb6...@cox.net het volgende geschreven: Hi Chris, If you study the schematic you will see that S3, the front panel lock-open-momentary switch, is in series between the standard cell (J10) and the null meter (J8). So what is happening is the voltage from the 750A goes to one side of the null meter and the reference (Standard Cell) goes to the other side of the null meter. Remove the wire from S1201 going to J7 and cover it with heat shrink. Then connect a wire between J7 and the 10 volt TAP (top of R12). From this point on, the Standard Cell banana jacks is where you connect the 731B 10 volt source, the null meter goes to its jacks and the S3 switch is still normal. The kelvin decades for the 1.xx voltage are no longer in the circuit and have no effect. Should you ever want to sell the 750A (you shouldn't) it is easy to reverse back to the factory arrangement. If you don't have a manual for the 750A, let me know and I will forward it to you. BillWB6BNQ Christopher Brown wrote: On 11/21/12 6:44 PM, WB6BNQ wrote: Hi Chris, Do you mean the Fluke 750A ? It has connections for standard cells. The 752 is a ratio device designed to be used with a 10 volt reference standard. No need for the 1.xx outputs from the 731B or the use of standard cells.. Your are correct, not the 752 (would like a 752 and a 720) but do have 2 750A. The 750A standard cell arrangement can be bypassed and set up for use with just a 10 volt reference using the original banana jacks for the reference and null meter as normal. All that is needed is to (non destructively) remove one connection and add one wire. BillWB6BNQ That sounds like a very worthwhile mod. How is it done and how does it change operation? As far as the 731B(s), I have 2. One is in good working order and overall great shape. Came out of a standards lab in the czech republic that recently switched to a 732A. Have a replacement NiCd pack currently cycling. Other than cleaning, and replacing the dead pack the only changes planned are replacing the IEC socket with a small filtered socket (have a box of them) and replacing R30 (if needed) to adjust the float current for the new pack (old is 450mah, new is 2100mah. For the second (received this morning), intend to do the same, and repair. I may just hard-wire around the failed switch, but beyond that, not fixing it would just bug me. The ref board is in good shape and does not appear to have been messed with. The main issues being the pot and the P/S section. Near as I can tell the Fairchild 1N5248B is a good match for CR1, and for 2 cents each I can just grab a batch and select. I have a bunch of 4148 and 1n914 fairchild parts around here somewhere, curious as to suggestion of 4448 over 4148? Looks like 2pf v.s. 4pf and the 4148 may start conducting a few hundred mv earlier from the datasheet. Thanks Chris - WL7CLA ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com
Re: [volt-nuts] Fluke 731B parts xref
On 11/21/12 6:44 PM, WB6BNQ wrote: Hi Chris, Do you mean the Fluke 750A ? It has connections for standard cells. The 752 is a ratio device designed to be used with a 10 volt reference standard. No need for the 1.xx outputs from the 731B or the use of standard cells.. Your are correct, not the 752 (would like a 752 and a 720) but do have 2 750A. The 750A standard cell arrangement can be bypassed and set up for use with just a 10 volt reference using the original banana jacks for the reference and null meter as normal. All that is needed is to (non destructively) remove one connection and add one wire. BillWB6BNQ That sounds like a very worthwhile mod. How is it done and how does it change operation? As far as the 731B(s), I have 2. One is in good working order and overall great shape. Came out of a standards lab in the czech republic that recently switched to a 732A. Have a replacement NiCd pack currently cycling. Other than cleaning, and replacing the dead pack the only changes planned are replacing the IEC socket with a small filtered socket (have a box of them) and replacing R30 (if needed) to adjust the float current for the new pack (old is 450mah, new is 2100mah. For the second (received this morning), intend to do the same, and repair. I may just hard-wire around the failed switch, but beyond that, not fixing it would just bug me. The ref board is in good shape and does not appear to have been messed with. The main issues being the pot and the P/S section. Near as I can tell the Fairchild 1N5248B is a good match for CR1, and for 2 cents each I can just grab a batch and select. I have a bunch of 4148 and 1n914 fairchild parts around here somewhere, curious as to suggestion of 4448 over 4148? Looks like 2pf v.s. 4pf and the 4148 may start conducting a few hundred mv earlier from the datasheet. Thanks Chris - WL7CLA ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] Fluke 731B parts xref
Hi Chris, If you study the schematic you will see that S3, the front panel lock-open-momentary switch, is in series between the standard cell (J10) and the null meter (J8). So what is happening is the voltage from the 750A goes to one side of the null meter and the reference (Standard Cell) goes to the other side of the null meter. Remove the wire from S1201 going to J7 and cover it with heat shrink. Then connect a wire between J7 and the 10 volt TAP (top of R12). From this point on, the Standard Cell banana jacks is where you connect the 731B 10 volt source, the null meter goes to its jacks and the S3 switch is still normal. The kelvin decades for the 1.xx voltage are no longer in the circuit and have no effect. Should you ever want to sell the 750A (you shouldn't) it is easy to reverse back to the factory arrangement. If you don't have a manual for the 750A, let me know and I will forward it to you. BillWB6BNQ Christopher Brown wrote: On 11/21/12 6:44 PM, WB6BNQ wrote: Hi Chris, Do you mean the Fluke 750A ? It has connections for standard cells. The 752 is a ratio device designed to be used with a 10 volt reference standard. No need for the 1.xx outputs from the 731B or the use of standard cells.. Your are correct, not the 752 (would like a 752 and a 720) but do have 2 750A. The 750A standard cell arrangement can be bypassed and set up for use with just a 10 volt reference using the original banana jacks for the reference and null meter as normal. All that is needed is to (non destructively) remove one connection and add one wire. BillWB6BNQ That sounds like a very worthwhile mod. How is it done and how does it change operation? As far as the 731B(s), I have 2. One is in good working order and overall great shape. Came out of a standards lab in the czech republic that recently switched to a 732A. Have a replacement NiCd pack currently cycling. Other than cleaning, and replacing the dead pack the only changes planned are replacing the IEC socket with a small filtered socket (have a box of them) and replacing R30 (if needed) to adjust the float current for the new pack (old is 450mah, new is 2100mah. For the second (received this morning), intend to do the same, and repair. I may just hard-wire around the failed switch, but beyond that, not fixing it would just bug me. The ref board is in good shape and does not appear to have been messed with. The main issues being the pot and the P/S section. Near as I can tell the Fairchild 1N5248B is a good match for CR1, and for 2 cents each I can just grab a batch and select. I have a bunch of 4148 and 1n914 fairchild parts around here somewhere, curious as to suggestion of 4448 over 4148? Looks like 2pf v.s. 4pf and the 4148 may start conducting a few hundred mv earlier from the datasheet. Thanks Chris - WL7CLA ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.