Re: [Elecraft] P3 lvl cal?

2013-04-15 Thread Sam Morgan

I have been using the following work around for this discrepancy.
I calibrated the K3's S-Meter using the XG3.
I also calibrated the P3 using the XG3 and found the same problem you see.

I have reset the P3's Lvl Cal to show the same S-meter readings as the K3
*WHEN* the P3 is in [PEAK] mode.
this way no matter what the span, an S9-20 on the K3 is about the same S9-20 on 
the P3.


I'm sure this is not a scientific way to do it, but it works for me.

ymmv
--
73  GB
Sam K5OAI

quote author=CX7TT
Hello All,

Question about the Lvl cal operation of the P3. I have noticed a 2 S
unit diff between the K3 and P3.  I used the XG-1 to make sure the K3
was set properly. Using Lvl cal, I noted the default was -5.4 and my  P3
was -9.6.  To get a K3 S meter reading of S4 to equal S4 on the P3, I
had to increase the Transfer gain to -30.0!

This seems excessive. Would appreciate some guidance on how to resolve
this issue.

73
Tom
CX7TT
/quote
--

GB  73
K5OAI
Sam Morgan
__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html


Re: [Elecraft] P3 lvl cal?

2013-04-15 Thread Matt Zilmer
Good points, all.  Thanks Alan.

73,
matt W6NIA

On Sun, 14 Apr 2013 18:13:35 -0700, you wrote:

Hi Matt,

If they matched it would be a coincidence.  For example, if the P3 is 
set to 45 kHz span, then each frequency point on the P3 display is about 
100 Hz wide.  If the K3 bandwidth is 2.7 kHz, then on a wideband signal 
like band noise or SSB, 27 times (14 dB) as much signal is getting to 
the K3's S meter as gets to each frequency point of the P3.  So you'd 
expect the average signal level on the P3 to be about 14 dB (2 or 3 
S-units) lower.

To illustrate this without a P3, try this:  Set the K3 to the normal 2.7 
kHz bandwidth, tune in an SSB signal and note the S meter reading.  Now 
change the K3 bandwidth to 100 Hz and note the S meter reading.  It will 
be much lower.

Alan N1AL


On 4/14/2013 4:37 PM, Matt Zilmer wrote:
 Hi Alan,

 I would guess that running Peak on the P3 and having the S meter on
 the K3 set to PEAK ON should give roughly the same results.  Correct?
 Or does the same discrepancy you advised below exist?  Peak and PEAK
 ON similarity is what I'm seeing here on the K3 / P3.

 73,
 matt W6NIA

 On Sun, 14 Apr 2013 16:19:17 -0700, you wrote:

 Tom,

 There are several reasons why the signal level can be different on the
 P3 and K3, as explained in the How to Set Up and Interpret the P3
 Display section of the P3 Owner's Manual.  The main reason is that the
 effective bandwidth of the P3 is narrower than the K3, so more of the
 signal gets through to the K3 S meter than a single pixel on the P3
 display.  For that reason, the signal or noise level on the P3 can
 appear quite a bit lower than on the K3 S meter.

 Alan N1AL


 On 4/14/2013 3:46 PM, cx...@4email.net wrote:
 Hello All,

 Question about the Lvl cal operation of the P3. I have noticed a 2 S
 unit diff between the K3 and P3.  I used the XG-1 to make sure the K3
 was set properly. Using Lvl cal, I noted the default was -5.4 and my  P3
 was -9.6.  To get a K3 S meter reading of S4 to equal S4 on the P3, I
 had to increase the Transfer gain to -30.0!

 This seems excessive. Would appreciate some guidance on how to resolve
 this issue.

 73
 Tom
 CX7TT


 __
 Elecraft mailing list
 Home: http://mailman.qth.net/mailman/listinfo/elecraft
 Help: http://mailman.qth.net/mmfaq.htm
 Post: mailto:Elecraft@mailman.qth.net

 This list hosted by: http://www.qsl.net
 Please help support this email list: http://www.qsl.net/donate.html


 __
 Elecraft mailing list
 Home: http://mailman.qth.net/mailman/listinfo/elecraft
 Help: http://mailman.qth.net/mmfaq.htm
 Post: mailto:Elecraft@mailman.qth.net

 This list hosted by: http://www.qsl.net
 Please help support this email list: http://www.qsl.net/donate.html
 __
 Elecraft mailing list
 Home: http://mailman.qth.net/mailman/listinfo/elecraft
 Help: http://mailman.qth.net/mmfaq.htm
 Post: mailto:Elecraft@mailman.qth.net

 This list hosted by: http://www.qsl.net
 Please help support this email list: http://www.qsl.net/donate.html


__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html
__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html


Re: [Elecraft] P3 lvl cal?

2013-04-14 Thread Alan Bloom

Tom,

There are several reasons why the signal level can be different on the 
P3 and K3, as explained in the How to Set Up and Interpret the P3 
Display section of the P3 Owner's Manual.  The main reason is that the 
effective bandwidth of the P3 is narrower than the K3, so more of the 
signal gets through to the K3 S meter than a single pixel on the P3 
display.  For that reason, the signal or noise level on the P3 can 
appear quite a bit lower than on the K3 S meter.


Alan N1AL


On 4/14/2013 3:46 PM, cx...@4email.net wrote:

Hello All,

Question about the Lvl cal operation of the P3. I have noticed a 2 S
unit diff between the K3 and P3.  I used the XG-1 to make sure the K3
was set properly. Using Lvl cal, I noted the default was -5.4 and my  P3
was -9.6.  To get a K3 S meter reading of S4 to equal S4 on the P3, I
had to increase the Transfer gain to -30.0!

This seems excessive. Would appreciate some guidance on how to resolve
this issue.

73
Tom
CX7TT


__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html



__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html


Re: [Elecraft] P3 lvl cal?

2013-04-14 Thread Matt Zilmer
Hi Alan,

I would guess that running Peak on the P3 and having the S meter on
the K3 set to PEAK ON should give roughly the same results.  Correct?
Or does the same discrepancy you advised below exist?  Peak and PEAK
ON similarity is what I'm seeing here on the K3 / P3.

73,
matt W6NIA

On Sun, 14 Apr 2013 16:19:17 -0700, you wrote:

Tom,

There are several reasons why the signal level can be different on the 
P3 and K3, as explained in the How to Set Up and Interpret the P3 
Display section of the P3 Owner's Manual.  The main reason is that the 
effective bandwidth of the P3 is narrower than the K3, so more of the 
signal gets through to the K3 S meter than a single pixel on the P3 
display.  For that reason, the signal or noise level on the P3 can 
appear quite a bit lower than on the K3 S meter.

Alan N1AL


On 4/14/2013 3:46 PM, cx...@4email.net wrote:
 Hello All,

 Question about the Lvl cal operation of the P3. I have noticed a 2 S
 unit diff between the K3 and P3.  I used the XG-1 to make sure the K3
 was set properly. Using Lvl cal, I noted the default was -5.4 and my  P3
 was -9.6.  To get a K3 S meter reading of S4 to equal S4 on the P3, I
 had to increase the Transfer gain to -30.0!

 This seems excessive. Would appreciate some guidance on how to resolve
 this issue.

 73
 Tom
 CX7TT


 __
 Elecraft mailing list
 Home: http://mailman.qth.net/mailman/listinfo/elecraft
 Help: http://mailman.qth.net/mmfaq.htm
 Post: mailto:Elecraft@mailman.qth.net

 This list hosted by: http://www.qsl.net
 Please help support this email list: http://www.qsl.net/donate.html


__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html
__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html


Re: [Elecraft] P3 lvl cal?

2013-04-14 Thread Alan Bloom

Hi Matt,

If they matched it would be a coincidence.  For example, if the P3 is 
set to 45 kHz span, then each frequency point on the P3 display is about 
100 Hz wide.  If the K3 bandwidth is 2.7 kHz, then on a wideband signal 
like band noise or SSB, 27 times (14 dB) as much signal is getting to 
the K3's S meter as gets to each frequency point of the P3.  So you'd 
expect the average signal level on the P3 to be about 14 dB (2 or 3 
S-units) lower.


To illustrate this without a P3, try this:  Set the K3 to the normal 2.7 
kHz bandwidth, tune in an SSB signal and note the S meter reading.  Now 
change the K3 bandwidth to 100 Hz and note the S meter reading.  It will 
be much lower.


Alan N1AL


On 4/14/2013 4:37 PM, Matt Zilmer wrote:

Hi Alan,

I would guess that running Peak on the P3 and having the S meter on
the K3 set to PEAK ON should give roughly the same results.  Correct?
Or does the same discrepancy you advised below exist?  Peak and PEAK
ON similarity is what I'm seeing here on the K3 / P3.

73,
matt W6NIA

On Sun, 14 Apr 2013 16:19:17 -0700, you wrote:


Tom,

There are several reasons why the signal level can be different on the
P3 and K3, as explained in the How to Set Up and Interpret the P3
Display section of the P3 Owner's Manual.  The main reason is that the
effective bandwidth of the P3 is narrower than the K3, so more of the
signal gets through to the K3 S meter than a single pixel on the P3
display.  For that reason, the signal or noise level on the P3 can
appear quite a bit lower than on the K3 S meter.

Alan N1AL


On 4/14/2013 3:46 PM, cx...@4email.net wrote:

Hello All,

Question about the Lvl cal operation of the P3. I have noticed a 2 S
unit diff between the K3 and P3.  I used the XG-1 to make sure the K3
was set properly. Using Lvl cal, I noted the default was -5.4 and my  P3
was -9.6.  To get a K3 S meter reading of S4 to equal S4 on the P3, I
had to increase the Transfer gain to -30.0!

This seems excessive. Would appreciate some guidance on how to resolve
this issue.

73
Tom
CX7TT


__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html



__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html

__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html



__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html


Re: [Elecraft] P3 LVL CAL ?

2011-02-15 Thread R. Kevin Stover
On Mon, 14 Feb 2011 22:09:01 -0800 (PST)
K7WIA emoss98...@msn.com wrote:

 
 I calibrated my K3 with the K3 utility with the XG2 , and set the Lvl
 Cal on the P3 to -10.6  for a S9 sig.
 but then when I went to the 6 Meter band I had to set the Lvl Cal to
 -22 to get a S9 sig on the P3,
 it's like there is not enough transfer gain when on the 6 meter
 band..even with the preamp on.
 so when changing from HF to 6 meters I have to change the P3. because
 the setting does not hold per band.
 could be that the P3 needs a spot to Cal just for 6 meters like it
 has for when using Xverters?
 
 Ed   K7WIA

I believe I saw a thread about this a couple weeks ago.

If I'm not mistaken Alan originally set the S9 level on the P3 for 6
meters as -93 dBm as it should be for a VHF band.

On HF it's set for -73 dBm.

I believe it was decided to change the six meter setting to -73 dBm. I
don't have any idea whether that's been done or not.





-- 
R. Kevin Stover
AC0H
__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html


Re: [Elecraft] P3 LVL CAL ?

2011-02-15 Thread K7WIA

Ok..every thing is working right.   Don't quite know what happend, we were
running the latest software.
but we had a big storm with wind yesterday and the lights were blinking, 
this morning re-loaded the latest software for the K3  P3 and re calibrated
and it seems to be OK.
Ed  K7WIA
-- 
View this message in context: 
http://elecraft.365791.n2.nabble.com/P3-LVL-CAL-tp6026432p6028250.html
Sent from the [K3] mailing list archive at Nabble.com.
__
Elecraft mailing list
Home: http://mailman.qth.net/mailman/listinfo/elecraft
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:Elecraft@mailman.qth.net

This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html