Hello Rich:
IEC 601-1 (Medical Electrical Equipment) in Clause 57.10 contains
requirements you allude to in the last paragraph of your message.
Bare live parts are subjected to a force of 2 N. In addition, metal
enclosures are subjected to an inward force of
When I sent this out, I selected reply to sender, rather than to all.
Here's what Rich got. For your perusal.
Peter L. Tarver
Nortel
ptar...@nt.com
--
From: Tarver, Peter (P.T.)
Sent: Friday, December 12, 1997 8:30 AM
--
From: Rich Nute[SMTP:ri...@sdd.hp.com]
Sent:
Rich Nute wrote:
While we are on the topic of component insulation, the spacings
must be measured AFTER the component is bent to its worst-case
position. Somewhere, in some safety standard, there exists a
requirement to verify internal spacings after the component has
been subjected to a
Prior to the EMC Directive, Germany required Class A ITE to be licensed.
Article 6 of the EMC Directive gives the right to a country to control
EMC compatibility by limiting the use of equipment at various sites.
This appears to allow countries to limit the use of Class A equipment.
Of course, the
Another source of standards information can be found on the web at
DGIII's reference to harmonized standards www2.echo.lu/nasd
Be careful of the date of the update (which is displayed on the page).
They have a tendency to let it fall behind. At the moment it is fairly
current. Also be careful of
Eric Lifesey nicely summarized the requirements of IEC (also EN) 61326 in his
e-mail copied below. I would like to add however that harmonics are indeed
covered in the emission section of the standard. The emission
limits as listed in Table 4 for Class B (non-industrial locations) include
Heber:
Does anyone know where I can find a chart showing which EN or IEC EMC
standards apply to what equipment?
Attached is a chart, in WordPerfect 6.1 format. You should also be able to
read it with other word processors. For the life of me, I can't remember
how this file materialized on my
Doug -
Firstly, I do not think that the EN has been listed in the Official
Journal (OJ) and so cannot be used (yet) to show compliance with the EU
EMC Directive.
As far as test plans go, they are required by the IEC 801 and EN61000
series of immunity specs (look in the Test Procedure
Hi George:
Here is what I sent to this group a week or so ago:
5. What are the IEC 950 enclosure flammability class requirements?
In Sub-clause 4.4.4, enclosure flammability class is specified
depending on the function of the enclosure:
Mechanical enclosure:HB
From: Arlen Olive arl...@futuretel.com
I've been working on a multimedia board in a standalone box
that connects to a PC parallel port. It also has an external
DC power supply, and connections for audio and video
input and output.
To pass FCC class B, I had to:
1. Shield the box
Hello from San Diego:
Eric Henning asks about clearance distances from an insulated
component to a grounded enclosure.
The issue is: What safety class is the insulation?
Insulation safety classes are:
Operational or functional (for functional purposes and which
is not
Thanks to Doug Frazee for pointing out that AC Harmonics and Flicker ARE in
IEC 61326-1 (contrary to my previous comment), but only as a requirement
for Class B emissions.
These tests are tucked into the emission limits table (Table 4 on page 29)
and buried between the typical emission limit
Moshe,
We currently test all of our new products to EN 61326-1 Class B.
The radiated emission limits, as well as line conducted limits, are the same
for
Class B equipment as defined in EN 55011.
The DOW is July 1, 2001.
If you plan to be selling your product past July 1, 2001, it may be best