Brian,
It is called out in the IEC 60335 standards. I don’t recall if it is in -1 or
the -2s.
Josh
Joshua Wiseman
Systems Engineering
Staff Engineer, Product Safety/EMC
Ortho Clinical Diagnostics
[cid:image001.png@01D69806.2DDD1E60]
From: Brian Kunde
Sent: Thursday, October 1, 2020 11:30 AM
Brian
It’s required by Annex I of the LVD:
2. Protection against hazards arising from the electrical equipment
Measures of a technical nature shall be laid down in accordance with point 1,
in order to ensure that:
(b) temperatures, arcs or radiation which would cause a danger, are not
Bernd
>So you may either create your own exhaustive risk analysis for your products,
>or claim compliance with the applicable product safety standard EN 62368-1, to
>demonstrate that your complete product does not present a risk and complies
>with the safety objectives of the LVD
A Risk
Hi Rich,
This may fall into ‘anecdotal’ but I recall CRT televisions could store charge.
I can’t recall if this was in an intentional capacitor, or capacitance inherent
in the tube, etc.
As a very practical, first hand experience matter, I used to work on small ion
implanters (essentially
Greetings,
Would any list member have a comment on a test requirement for Grille Cloth in
UL/IEC/EN 62368-1 ?
My word search finds no mention of "Grille Cloth"
In the past UL 60065 wanted a tablet test performed on cloth.
"Grille covering material, cloth, and reticulated foam"
The part is to
When do you know if a product needs to be tested to these standards or
not? What requires this testing? Is there a criteria that can be used?
Does the SCOPE of these standards make it clear if and when the standards
need to be applied to a product? We have had customers request it
but most
Scott
There no “solid guidance”
EN 50663:2017 is a much more recent standard and makes reference to both EN
62311:2008 and EN 62479:2010 for assessment, but does not allow use of EN 62479
table B.1 for using higher transmit powers for values of Pmax
Best regards
Charlie
Charlie Blackham
7 matches