Hi All,
I am pulling up an old thread here, because it may relate to my question.
Is there any requirement or guidance for sound pressure level limits from
ultrasonic equipment operating 40-50 kHz and used in a residential
environment?
61010 has a good reference in section 12.5.2, and 60950 has
Good morning,
I wanted to share a position announcement with the group. OSHA's
Nationally Recognized Testing Laboratory (NRTL) office is looking to hire
another engineer/laboratory assessor. Complete details of the position can
be found at this link
Hi Lauren:
I work mostly with EN 60950, but the problem you describe with the 200V
circuit embedded in an otherwise safe circuit is a problem I
have encountered several times with 60950. So, I will just make a
few general observations that may be helpful:
1) In order for the 200V circuit to
In message
calvemi_04vk4qkvyehcfqdwhkbvepr1no7-dfgolazu5em2...@mail.gmail.com,
dated Thu, 27 Mar 2014, S Drysdale sdd...@gmail.com writes:
I am pulling up an old thread here, because it may relate to my
question. Is there any requirement or guidance for sound pressure
level limits from
Hi Lauren,
I just joined on to this list, so I'm not sure where you're at in the
process, but based on your description of this circuit you should be
looking at 6.7.3 for creepage and clearance. This section is for secondary
circuits like those supplied by bricks on strings.
You should
Hello all,
I'm wondering if anyone has experience with test equipment intended to put
out a high voltage to a set of accessible test leads, e.g. a hipot tester.
If so, for compliance to 61010-1, how do you get around the access to
hazardous live parts issue?
Thanks,
*Paul MiltonGM
There are sections of the standard(s) that address construction and performance
where, for operating reasons, haz V must be exposed.
Brian
From: Paul Milton [mailto:p...@gmcompliance.com]
Sent: Thursday, March 27, 2014 11:48 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: [PSES] High voltage output
Hi Paul,
This is allowed but you cannot use the 61010-1 Part 1 basic standard alone.
You must bring in a Part 2, particular standard to be used in
conjunction with Part 1. Select the Part 2 for the type of handheld probes
you are using. Take a look at EN/IEC 61010-2-30 and EN/IEC 61010-2-32.
We have a dielectric tester which makes use of retractable high voltage test
probes.
The probe requires an operator to depress a plunger to access the conductive
tip.
Google comes up with some options.
Donald McElheran
Product Compliance Specialist
Ross Video | Production Technology
In message
cabyvtvnveyhm5sqvhvmgpbdeqptxj0x3jl7dme2vtndarmc...@mail.gmail.com,
dated Thu, 27 Mar 2014, Doug Powell doug...@gmail.com writes:
. In the past, this has been a serious issue when the timing of
mandatory compliance with the Part 1 Ed 3 is in effect and the planned
update to a
Hi Everyone,
I have been back to writing again. Here is my latest article:
Technical
Tidbit - March-April 2014
Troubleshooting Radiated and Conducted Immunity Problems in the
Development Lab
This
month's Technical Tidbit
Hi Lauren:
You ask some questions whose answers cannot be short.
Transient overvoltages are normal in any power distribution
system. They arise mainly from two sources:
1) lightning;
2) switching equipment on and off.
Transient overvotages are either normal mode or common
mode. Normal mode