Kelly, You are not going to stay within the temp ranges if putting pellets
directly into the embedding center tank. A couple solutions: 1) write into your
procedure that the temperature will go down when pellets are added and that
once melted should be in whatever range it calls for (and don't turn it up,
because that means you have to keep adjusting it and someone may forget and as
you say, exceed the temperature - you just need to put in far enough ahead to
let it melt. Or, 2) Get an bulk paraffin dispenser and melt the paraffin in
that, then draw off liquid paraffin for your embedding center. This is pretty
common for tissue processor and embedding centers to avoid having to use them
to melt the paraffin. Gets you back to work faster as well.
Tim Morken
Supervisor, Electron Microscopy/Neuromuscular Special Studies
Department of Pathology
UC San Francisco Medical Center
-Original Message-
From: E. Wayne Johnson via Histonet
Sent: Tuesday, December 08, 2020 1:26 PM
To: Pairan, Kelly ;
'histonet@lists.utsouthwestern.edu'
Subject: Re: [Histonet] Maximum and Minimum Temperatures on Embedding Center
Paraffin Tanks
You could warm the paraffin in an oven overnight to melt it so that the heat of
fusion is not extracted from the embedding center paraffin tanks when it is
added as melted paraffin at 58-65 degrees.
Pairan, Kelly via Histonet wrote:
> Good Afternoon,
> Recently we starting taking the daily maximum and minimum temperature ranges
> for our embedding center paraffin tanks. The melting point of our wax is
> 56oC and our current temperature range is 58 to 65oC. The problem we are
> having is that when we add wax, the temp dips below the range until it melts.
> If we turn it up, it exceeds the max range. Any suggestions?
>
> Thanks,
> Kelly
> ___
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthweste
> rn.edu_mailman_listinfo_histonet=DwICAg=iORugZls2LlYyCAZRB3XLg=7
> cy9qXFa73jDX2Iixpjkq1XlWAfHgLLHm33agI_sCKA=4_M_9rKMFL-Sh7pBndIIL-bv0
> oW9Qdhfz4CMKoyUJz0=sXXQcK8dPXYAYVBhgQQkD_qJtcBV3mOUyb7ehqCprQg=
>
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
https://urldefense.proofpoint.com/v2/url?u=http-3A__lists.utsouthwestern.edu_mailman_listinfo_histonet=DwICAg=iORugZls2LlYyCAZRB3XLg=7cy9qXFa73jDX2Iixpjkq1XlWAfHgLLHm33agI_sCKA=4_M_9rKMFL-Sh7pBndIIL-bv0oW9Qdhfz4CMKoyUJz0=sXXQcK8dPXYAYVBhgQQkD_qJtcBV3mOUyb7ehqCprQg=
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet