Re: [Ifeffit] S02 selection from reviewer

2021-10-02 Thread Jeff Catalano

Peng,

I think what everyone has said is quite useful, but there is likely 
another aspect to consider as you make revisions. Based on the reviewer 
comments, it sounds like you are trying to interpret your fitted N 
values as being statistically distinct. My guess is that is not correct, 
as Matt suggested, and the fact that the N is lower in fluorescence 
suggests that overabsorption is attenuating the EXAFS amplitude, as 
Matthew suggested.


If you really want to make the claim of differences in coordination 
environment then you should also look at R. A lower coordination number 
should be accompanied by a decrease in R. How R shifts will be impacted 
by the resolution of the data, determined by your k-range, but I think 
if fitting a single Sb-O path then in most cases you should see R change 
if N is truly going down. It can get more complicated, clearly, and the 
S02 you are using may not be a perfect value, as Shelly's paper 
indicates, but if N is truly going down in one sample then R definitely 
should be changing.


I do not think about Sb very often, so if the only way for coordination 
to change is from a change in redox state then you should see a shift in 
the XANES spectrum. However, be cautious here because overabsorption 
will increase the intensity of a XANES spectrum for normalized 
absorbances below 1 and decrease the intensity above 1. If the XANES 
fine structure features of all samples are the same (in position) but 
you see what I describe in the overall absorbance, then your 
fluorescence samples is likely just suffering from overabsorption 
effects and your N will appear smaller. This gives a lower apparent S02, 
which is likely what the reviewer is getting at.


My sense is that what the reviewer is really criticizing is your 
interpretations based on differences in fitted N values, which may not 
be valid.


Jeff



On 10/2/2021 9:48 AM, Kelly, Shelly D. wrote:

Hi Peng:

It might be helpful to understanding some of Matt's points regarding S02 
transferability, Ei and energy resolution by looking at this paper.

Comparison of EXAFS foil spectra from around the world
DOI: 10.1088/1742-6596/190/1/012032

Kind regards,
Shelly

-Original Message-
From: Ifeffit  On Behalf Of Matthew 
Marcus
Sent: Saturday, October 2, 2021 9:04 AM
To:ifeffit@millenia.cars.aps.anl.gov
Subject: Re: [Ifeffit] S02 selection from reviewer

Since S02 is a parameter in the description of EXAFS and not of the experiment, 
it's independent of technique.  Overabsorption (misnamed
'self-absorption') can reduce the *measured* amplitude, an effect which can be 
fudged in analysis by reducing S02.  If the sample is truly homogeneous (on the 
scale of the absorption length), then you can calculate the amount of 
overabsorption to see if it's significant.
However, many kinds of samples, such as concentrated powders mixed with a 
diluent such as BN, this condition is not met.  If the particles are large 
enough for each to have significant absorption edge jumps, then diluting them 
in BN doesn't fix the problem.
mam

On 10/2/2021 12:49 AM, Peng Liu wrote:

Dear IFEFFIT members,

I am sorry to bother you again. I asked about S02 selection for the
first major revision. I just received the second revision. The
reviewer is not satisfied with one S02 value for all our samples.
"

1. I am still not satisfied with selected SO2 value (it is set to 0.85).
SO2 is not transferable between different samples and detection methods.
It is not possible to use a value obtained from different compound
using transmission measurement mode to completely different other
compound measured using fluorescence mode. One method to fix SO2 value
is to measure diluted solution (to avoid self-absorption) of reference
material in fluorescence mode. Other is to use multiple spectra
fitting for all samples of interest (e.g. with Sb(V)) measured using
fluorescence mode where SO2 parameter is the same for all samples.

At the same time I am confident that CN values 5.6, 7.1 and 6.9
correspond to CN(Sb-O)=6. I suggest reconsidering the SO2 value for
measurements in fluorescence mode.

"

We do get the S02 from a similar reference material measured in
transmission mode, and our samples were all measured in fluorescence
mode. It is not possible to measure the diluted reference material in
fluorescence mode in one or two months. If you could give me some
suggestions, that would be great.


--
Best Regards,

Peng Liu

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe:http://millenia.cars.aps.anl.gov/mailman/options/ifeffit


___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe:http://millenia.cars.aps.anl.gov/mailman/options/ifeffit

___
Ifeffit mailing 

[Ifeffit] Substantial problems running ATOMS in Demeter 0.9.24

2016-06-03 Thread Jeff Catalano
The message sequence from this morning prompted me to inquire if anyone 
can provide help with ATOMS crashing on my own computer. I am running 
Demeter 0.9.24 on Windows 10 Pro 64. I did a fresh install of Demeter 
after having removed all older versions of Demeter 0.9, including the 
Demeter folders in C:\Users\\AppData\Roaming\. My problem is 
unfortunately simple and something I do not know how to fix: Whenever I 
hit the "Run Atoms" button, stand-alone atoms crashes. The program 
simply dies, and no files are written. I am able to enter text, save an 
atoms.inp file, and reload an atoms.inp file, but "Run Atoms" crashes 
the program. I have tested this with a simple compound, NaCl, using the 
attached atoms.inp file. Note that I don't want to work on NaCl, I just 
used this to test a simple structure. The log file associated with the 
crash is attached.  The log file says Windows 8, but I think this is 
just hows Windows 10 shows up as that is what I am running, from a clean 
Windows 10 install. Anyway, when I load this atoms.inp file or enter the 
same information by hand the program crashes. I was able to load and run 
this atoms.inp file in the old version of atoms (3.0.1) and produced the 
attached feff_oldATOMS.inp file.  I thus don't think it is a data entry 
problem.


I hope that I have provided enough information. I have no problems 
running ATHENA, ARTEMIS, or HEPHAESTUS. Given the log file I suspect 
this is something unique to my computer, but I am stumped as to what it 
could be as I am running a standard OS with a standard install on Demeter.


Thank you,
Jeff

--
Jeffrey G. Catalano, Associate Professor
Department of Earth and Planetary Sciences
Washington University
1 Brookings Drive, Campus Box 1169
Saint Louis, MO 63130
USA
http://aqgeochem.wustl.edu/

## This Atoms file was generated by Demeter 0.9.24
## Demeter written by and copyright (c) Bruce Ravel, 2006-2015

title = Halite
space = Fm3m
a =   5.64010b=   5.64010c =   5.64010
alpha =  90.0beta =  90.0gamma =  90.0
rmax  =   8.0core  = Cl
# polarization = 0.0  0.0  0.0
shift = 0.00.00.0
atoms
# el. x   y   ztag
  Na 0.0 0.0 0.0   Na
  Cl 0.5 0.5 0.5   Cl
Started at 2016-06-03T12:54:07
Win8Professional (64-bit)   6292002002561

PATH is:

C:\Windows\system32;C:\Windows;C:\Windows\System32\Wbem;C:\Users\catal\AppData\Roaming\DemeterPerl\c\bin;C:\Users\catal\AppData\Roaming\DemeterPerl\perl\site\bin;C:\Users\catal\AppData\Roaming\DemeterPerl\perl\bin;C:\Users\catal\AppData\Roaming\DemeterPerl\c\bin\gnuplot\bin

perl version: v5.18.2

@INC:

C:/Users/catal/AppData/Roaming/DemeterPerl/perl/site/lib/MSWin32-x64-multi-thread
C:/Users/catal/AppData/Roaming/DemeterPerl/perl/site/lib
C:/Users/catal/AppData/Roaming/DemeterPerl/perl/vendor/lib
C:/Users/catal/AppData/Roaming/DemeterPerl/perl/lib
.
Demeter version 0.9.24

Can't call method "workspace" on an undefined value at 
C:/Users/catal/AppData/Roaming/DemeterPerl/perl/site/lib/Demeter/UI/Atoms/Xtal.pm
 line 1004.

 * This feff6 input file was generated by Atoms 3.0.1
 * Atoms written by and copyright (c) Bruce Ravel, 1998-2001

 * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * 
 *   total mu*x=1: 3.89 microns,  unit edge step:21.71 microns
 *   specific gravity =  2.164
 * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * 
 *   Normalization correction:0.00105 ang^2
 * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * -- * 

 * -
 * The following crystallographic data were used:
 *
 * titleHalite
 * space = F m -3 m
 * a =5.64010   b =   5.64010   c =   5.64010
 * alpha =   90.0   beta =   90.0   gamma =  90.0
 * core =   Cl  edge =  K
 * atoms
 * ! elem   x  y  z   tagocc
 *   Na0.00.00.0  Na1.0
 *   Cl0.50.50.5  Cl1.0
 * -


 TITLE Halite

 HOLE 1   1.0   *  Cl K edge  (2822.0 eV), second number is S0^2

 * mphase,mpath,mfeff,mchi
 CONTROL   1  1 1 1
 PRINT 1  0 0 0

 RMAX8.0

 *CRITERIA curved   plane
 *DEBYEtemp debye-temp
 NLEG 4

 POTENTIALS
 *ipot   Z  element
0   17   Cl
1   11   Na
2   17   Cl

 ATOMS  * this list contains 93 atoms
 *   x  y  z  ipot  tag  distance
0.00.00.0  0 Cl  0.0
2.820050.00.0  1 Na_12.82005
   -2.820050.00.0  1 Na_12.82005
0.02.82005

Re: [Ifeffit] Unable to export flattened normalized XANES spectra from Athena 0.9.24

2015-12-09 Thread Jeff Catalano

Hi Bruce,

I appreciate the reply. It looks like what I want is the flattend column 
of data, which is not column 2 in the version 0.9.24 norm(E) files, it 
is column 3. I was unaware both flattened and unflatten normalized data 
were exported in the file. In Athena 0.8 if "Flatten normalized data" 
was selected then column 2 was this flattened data in a norm(E) file. I 
had no idea that the export format for norm(E) files changed in version 
0.9. To make sure I have this correct, below are the column labels from 
norm(E) files exported from Athena 0.8.061 and Athena 0.9.24:


0.8.061: #  energy norm bkg_norm der_norm
0.9.24: #   e normflatfbkg ndernsec

I have plotting code that expects columns 1 and 2 and I was unaware that 
flattened data is now output separately (in column 3). In version 0.8 a 
flattened normalized spectra was outputted if the box was checked, and 
if it was unchecked then column 2 was the unflattened data, from what I 
can tell. Can you please confirm that I understand this correctly?


I admit that I had not looked at the documentation before writing since 
I understood the Athena 0.8 file format and had not anticipated a format 
change in the new version.  It clearly would have been informative to 
look at the current documentation! Of course, as you note that the 
columns listed in the documentation for norm(E) files are wrong, with 
the current output format being columns 1 4 2 3 6 7 from what is listed 
in the documentation (column 5 does not appear in my file). Should this 
be looked at?  I actually would prefer the column order that is listed 
in the documentation. The file format is clearly not what you intended.


Thank you,
Jeff

On 12/8/2015 9:15 PM, Bruce Ravel wrote:


Here's what the manual has to say

http://bruceravel.github.io/demeter/aug/output/column.html

Looking at the code, however, it seems the document has swapped 
columns 2/3 and columns 4/5.  I think what you are looking for is in 
column 4, not column 2.  If that's right, I'll treat this as a 
documentation bug.  If that is not the case, then I'll prod you for 
more information so I can understand the software bug.


In either case, thanks for the report.

Cheers,
B


On 12/08/2015 06:09 PM, Jeff Catalano wrote:

I am looking for some help with a possible error on my part or bug
associated with exporting normalized mu(E) spectra from Athena 0.9.24
64-bit on Windows 10 Pro 64-bit. I have some Si XANES data with a
challenging background shape that is well removed through careful
normalization in Athena, as long as the "Flatten normalized data" option
is checked. When I save the current group as norm(E), what is saved is
the non-flattened normalized data. The checkbox for flattening the data
has no effect on what is exported despite changing what is plotted. I
have verified this by comparing the files exported the the flatten box
checked or unchecked. I have tried this with multiple data files loaded
in multiple formats (raw beamline data, simple two-column E and mu data,
PRJ files) and the effect is always the same. It is also unaffected by
the normalization ranges I select. No matter what options I select
Athena 0.9.24 will only save non-flattened norm(E) data.

This behavior is different from what I remember, and I have verified
that in version 0.8 of Athena the exported norm(E) was the flattened
data, provided that the flatten box was checked (of course). I have been
unable to determine if I have any errors in the settings in version
0.9.24 but all parameters that I can find are identical in both 0.9.24
and 0.8. I have also uninstalled and reinstalled Athena 0.9.24 and the
effect is still the same. Has anyone else noticed this problem? Is this
a bug?

Thank you,
Jeff







--
Jeffrey G. Catalano, Associate Professor
Department of Earth and Planetary Sciences
Washington University
1 Brookings Drive, Campus Box 1169
Saint Louis, MO 63130
USA
http://aqgeochem.wustl.edu/

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit


[Ifeffit] Unable to export flattened normalized XANES spectra from Athena 0.9.24

2015-12-08 Thread Jeff Catalano
I am looking for some help with a possible error on my part or bug 
associated with exporting normalized mu(E) spectra from Athena 0.9.24 
64-bit on Windows 10 Pro 64-bit. I have some Si XANES data with a 
challenging background shape that is well removed through careful 
normalization in Athena, as long as the "Flatten normalized data" option 
is checked. When I save the current group as norm(E), what is saved is 
the non-flattened normalized data. The checkbox for flattening the data 
has no effect on what is exported despite changing what is plotted. I 
have verified this by comparing the files exported the the flatten box 
checked or unchecked. I have tried this with multiple data files loaded 
in multiple formats (raw beamline data, simple two-column E and mu data, 
PRJ files) and the effect is always the same. It is also unaffected by 
the normalization ranges I select. No matter what options I select 
Athena 0.9.24 will only save non-flattened norm(E) data.


This behavior is different from what I remember, and I have verified 
that in version 0.8 of Athena the exported norm(E) was the flattened 
data, provided that the flatten box was checked (of course). I have been 
unable to determine if I have any errors in the settings in version 
0.9.24 but all parameters that I can find are identical in both 0.9.24 
and 0.8. I have also uninstalled and reinstalled Athena 0.9.24 and the 
effect is still the same. Has anyone else noticed this problem? Is this 
a bug?


Thank you,
Jeff

--
Jeffrey G. Catalano, Associate Professor
Department of Earth and Planetary Sciences
Washington University
1 Brookings Drive, Campus Box 1169
Saint Louis, MO 63130
USA
http://aqgeochem.wustl.edu/

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit