Re: [Ifeffit] Fit in R or k

2021-10-12 Thread Kelly, Shelly D.
Hi Will:

The relationship between the data in k-space and R-space is described in this 
tutorial:
https://docs.xrayabsorption.org/tutorials/Basics_of_XAFS_to_chi_2009.pdf
Take a look at what a Fourier transform does.  In few words I'd say it doesn't 
matter which representation of the data (k-space or R-space) you model. Both 
have their good and not so good points.
R-space:  You can filter out noise, and higher and/or lower frequencies that 
you might not be interested in.  You can also weight the lower or higher part 
of the spectra you are fitting by using k-weighting.
k-space: You are working with less processed data and can directly see the 
noise and high/lower frequency components "on top of" your signal of interest.

If it still isn't clear I'd suggest you start with some of the material at this 
web site:
https://xrayabsorption.org/tutorials/

There is also this very readable book by Scott Calvin:
https://www.amazon.com/XAFS-Everyone-Scott-Calvin/dp/1439878633/ref=asc_df_1439878633/?tag=hyprod-20=df0=312091458201==g=7369607741759776259c===9021703=pla-455399119345=1===63669393113===312091458201==g=7369607741759776259==c===9021703=pla-455399119345

I also have an EXAFS chapter that I can send you a copy of, if you send me a 
request directly.


Kind regards,
Shelly



From: Ifeffit  On Behalf Of LIMA DA 
SILVA, WILL (PGR)
Sent: Tuesday, October 12, 2021 4:10 PM
To: ifeffit@millenia.cars.aps.anl.gov
Subject: [Ifeffit] Fit in R or k

Dear IFEFFIT members,

I have been fitting some EXAFS data, and my supervisor and I are having doubts 
if it's better to fit in k or R space. What are the main differences between 
the two approaches?
I have learnt that R space is the right approach, but I am not sure how to 
explain it.

Much appreciated,

Will Silva
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit


Re: [Ifeffit] S02 selection from reviewer

2021-10-02 Thread Kelly, Shelly D.
Hi Peng:

It might be helpful to understanding some of Matt's points regarding S02 
transferability, Ei and energy resolution by looking at this paper. 

Comparison of EXAFS foil spectra from around the world 
DOI: 10.1088/1742-6596/190/1/012032 

Kind regards,
Shelly

-Original Message-
From: Ifeffit  On Behalf Of Matthew 
Marcus
Sent: Saturday, October 2, 2021 9:04 AM
To: ifeffit@millenia.cars.aps.anl.gov
Subject: Re: [Ifeffit] S02 selection from reviewer

Since S02 is a parameter in the description of EXAFS and not of the experiment, 
it's independent of technique.  Overabsorption (misnamed
'self-absorption') can reduce the *measured* amplitude, an effect which can be 
fudged in analysis by reducing S02.  If the sample is truly homogeneous (on the 
scale of the absorption length), then you can calculate the amount of 
overabsorption to see if it's significant. 
However, many kinds of samples, such as concentrated powders mixed with a 
diluent such as BN, this condition is not met.  If the particles are large 
enough for each to have significant absorption edge jumps, then diluting them 
in BN doesn't fix the problem.
mam

On 10/2/2021 12:49 AM, Peng Liu wrote:
> Dear IFEFFIT members,
> 
> I am sorry to bother you again. I asked about S02 selection for the 
> first major revision. I just received the second revision. The 
> reviewer is not satisfied with one S02 value for all our samples.
> "
> 
> 1. I am still not satisfied with selected SO2 value (it is set to 0.85). 
> SO2 is not transferable between different samples and detection methods. 
> It is not possible to use a value obtained from different compound 
> using transmission measurement mode to completely different other 
> compound measured using fluorescence mode. One method to fix SO2 value 
> is to measure diluted solution (to avoid self-absorption) of reference 
> material in fluorescence mode. Other is to use multiple spectra 
> fitting for all samples of interest (e.g. with Sb(V)) measured using 
> fluorescence mode where SO2 parameter is the same for all samples.
> 
> At the same time I am confident that CN values 5.6, 7.1 and 6.9 
> correspond to CN(Sb-O)=6. I suggest reconsidering the SO2 value for 
> measurements in fluorescence mode.
> 
> "
> 
> We do get the S02 from a similar reference material measured in 
> transmission mode, and our samples were all measured in fluorescence 
> mode. It is not possible to measure the diluted reference material in 
> fluorescence mode in one or two months. If you could give me some 
> suggestions, that would be great.
> 
> 
> --
> Best Regards,
> 
> Peng Liu
> 
> ___
> Ifeffit mailing list
> Ifeffit@millenia.cars.aps.anl.gov
> http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
> Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit
> 
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit


[Ifeffit] Beamline Scientist position

2021-07-01 Thread Kelly, Shelly D.
Hello XAFS community:

I'm excited to let you know that we have an opening for beamline scientist 
within the Spectroscopy group at the Advanced Photon Source. Please help spread 
the word.
https://bit.ly/3iN7Ehe

Kind Regards,
Shelly Kelly
Spectroscopy Group Leader
Advanced Photon Source

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
Unsubscribe: http://millenia.cars.aps.anl.gov/mailman/options/ifeffit


Re: [Ifeffit] quotation of figures

2008-07-16 Thread Kelly, Shelly D.
I'd like to add my 2cents worth about the mean square displacement term.  I 
wrote it up on the EXAFS wiki, since I like to say it over and over. see 
http://www.xafs.org/Common_Mistakes
 
Shelly



From: [EMAIL PROTECTED] on behalf of Scott Calvin
Sent: Wed 7/16/2008 3:54 PM
To: XAFS Analysis using Ifeffit
Subject: Re: [Ifeffit] quotation of figures



Hi Rich,

Mean-square radial disorder (MSRD) is one good alternative.

I'll also point out another bit of ambiguity with Debye-Waller 
factor aside from the confusion with the XRD term. Some EXAFS 
practitioners use Debye-Waller factor to indicate all disorder-type 
modifications to the EXAFS equation, including higher cumulants. Most 
use it only for the Gaussian part; i.e. the second cumulant. Another 
good reason to avoid the term.

--Scott Calvin
Sarah Lawrence College

On Jul 16, 2008, at 4:41 PM, Richard Mayes wrote:





 I also dislike the phrase Debye-Waller factor for the exafs 
 disorder
 term.  In crystallography, the Debye-Waller factor refers to disorder
 of atoms about their lattice positions.  In exafs, the disorder is
 about the path length of the N-body configuration -- clearly not the
 same thing.  I am, apparently, in the minority on this topic --
 Debye-Waller factor is in wide use in the exafs literature.  But, 
 as
 everyone here knows, I often like to stand up on my soapbox and yell
 into the crowd. ;-)


 Bruce,

 While you're on your soapbox, if we drop Debye-Waller factor to 
 describe
 the EXAFS disorder term, would you suggest using only disorder or 
 do you
 have a better phrase to use?

 Rich

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit



___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] quotation of figures

2008-07-16 Thread Kelly, Shelly D.
Hi Eckhard,
 
The first slide also has problems.  The first path is a single scattering path. 
 All the others are double scattering paths.  The word single, double and 
triple are used to describe the number of atoms that scatter.  The number of 
legs is used to tell the difference between the different kinds of double 
scattering.  So the first (A-B-A) single scattering path has 2 legs (nlegs=2). 
The next one (A-C-B-A) is a double triangle scattering path with 3 legs. The 
next one (A-B-C-A) is equivalent to the previous one so that the CN for that 
path is twice the number of atoms for that path. The last one (A-C-B-C-A) is a 
double scattering path with 4 legs.
 
HTH
Shelly



From: [EMAIL PROTECTED] on behalf of [EMAIL PROTECTED]
Sent: Wed 7/16/2008 11:04 AM
To: ifeffit@millenia.cars.aps.anl.gov
Subject: [Ifeffit] quotation of figures



Hi,
i have a print out of a talk (about XANES, XAFS and multiple
scattering...) and i want to cite some figures (attached). Can anybody
tell me the author etc...? I searched everywhere on the web (on several
sites of XAFS-Summer Schools...). Unfortunately i do not have the talk as
pdf-file only the print out. Maybe the talk is from Matt Newville? Maybe
anybody recalls on this talk/slides or recognizes them?

Thanks for your help.

Have a nice day,
Eckhard


-
Eckhard Bosman
[EMAIL PROTECTED]
+49 (0)551-39-14441
Raum: E0.104
Institut für Röntgenphysik
Friedrich-Hund-Platz 1
37077 Göttingen
Germany 


___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] quotation of figures

2008-07-16 Thread Kelly, Shelly D.
Yah know,,,I'm wrong The last one is a triple scattering path with just two 
scattering atoms. the leg part is right...
 
oops
SK



From: [EMAIL PROTECTED] on behalf of Kelly, Shelly D.
Sent: Wed 7/16/2008 11:45 PM
To: XAFS Analysis using Ifeffit
Subject: Re: [Ifeffit] quotation of figures



Hi Eckhard,

The first slide also has problems.  The first path is a single scattering path. 
 All the others are double scattering paths.  The word single, double and 
triple are used to describe the number of atoms that scatter.  The number of 
legs is used to tell the difference between the different kinds of double 
scattering.  So the first (A-B-A) single scattering path has 2 legs (nlegs=2). 
The next one (A-C-B-A) is a double triangle scattering path with 3 legs. The 
next one (A-B-C-A) is equivalent to the previous one so that the CN for that 
path is twice the number of atoms for that path. The last one (A-C-B-C-A) is a 
double scattering path with 4 legs.

HTH
Shelly



From: [EMAIL PROTECTED] on behalf of [EMAIL PROTECTED]
Sent: Wed 7/16/2008 11:04 AM
To: ifeffit@millenia.cars.aps.anl.gov
Subject: [Ifeffit] quotation of figures



Hi,
i have a print out of a talk (about XANES, XAFS and multiple
scattering...) and i want to cite some figures (attached). Can anybody
tell me the author etc...? I searched everywhere on the web (on several
sites of XAFS-Summer Schools...). Unfortunately i do not have the talk as
pdf-file only the print out. Maybe the talk is from Matt Newville? Maybe
anybody recalls on this talk/slides or recognizes them?

Thanks for your help.

Have a nice day,
Eckhard


-
Eckhard Bosman
[EMAIL PROTECTED]
+49 (0)551-39-14441
Raum: E0.104
Institut für Röntgenphysik
Friedrich-Hund-Platz 1
37077 Göttingen
Germany


___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit



___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Athena: k-weight=3 in background removal problem

2008-07-08 Thread Kelly, Shelly D.
Hi Joseph,

I have a book chapter that explains how the different k-weighting affect
the chi(k) data produced in the background removal.  I'll send it to you
directly, so as not to fill up everyone's mail box.

Here is the reference for others that might be interested. 

Kelly, S.D., Hesterberg, D., and Ravel, B., in Analysis of Soils and
Minerals Using X-ray Absorption Spectroscopy, (A.L. Ulery and L.R.
Drees, Eds.) p. 367. Soil Science Society of America, Madison, WI, USA,
2008.

Shelly

-Original Message-
From: [EMAIL PROTECTED]
[mailto:[EMAIL PROTECTED] On Behalf Of Joseph
Washington
Sent: Tuesday, July 08, 2008 6:03 AM
To: ifeffit@millenia.cars.aps.anl.gov
Subject: Re: [Ifeffit] Athena: k-weight=3 in background removal problem

Hi Carlo,
Since much of our data is located at high k values, it would be an 
improvement to be using a high k weight in the background removal. 
Whether or not this will be a vast improvement might be questionable. My

main concern is why this occurs for all our Ge edge data  across the 
board, and if this is indicative of an underlying issue in data 
collection at the beamline or preprocessing.

Please see an attached project file containing some scans and a merge.

I appreciate the assistance,

Joseph

Carlo Segre wrote:
 
 Hi Joseph:
 
 On Mon, 7 Jul 2008, Joseph Washington wrote:
 
 Greetings all,
 I am an APS user and a graduate of XAFS Summer School - class of 2007
:)
   We have been doing EXAFS analysis on Ge[2]Sb[2]Te[x] thin films. I
 noticed that at the Ge edge in particular, the k-weight value used in
 the background removal cannot equal 3, or else the background
function
 no longer follows the data and oscillates wildly and/or it shoots up
at
 the edge energy dominating the plot window. Values of 1 and 2 work
fine.
 Even more puzzling is that this problem is inherent for the Ge edge
of
 data taken over numerous runs, and for various samples (GST-224,
 GST-225, 226, 227).
 
 COuld you tell us why you would like to use a k-weight of 3 in the 
 background removal process?  I find that 1 (or 2 sometimes) is 
 sufficient to get a good result.  I just checked on a recent data set 
 and I agree that k-weight 1 nd 2 give nearly identical results but
that 
 k-weight 3 for the same parameters is slightly different.
 
 It would help a lot if you oculd send a project file for us to look
at.
 
 Carlo
 

-- 
Joseph Washington
Research Assistant
Department of Physics
North Carolina State University
431 Riddick Hall
Raleigh, NC 27695-8202
email: [EMAIL PROTECTED]

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Dodgy Edge Steps

2008-04-28 Thread Kelly, Shelly D.
Hi Lachlan,

 

The reason that you want to keep the total absorption less than 3 is because 
the number of transmitted x-rays becomes small.  I=I0 exp(-mu x).  So for mu x 
=3 the number of transmitted x-rays are 5% of the incident x-ray intensity.  If 
you have a lot of x-rays then you can measure spectra with greater total 
absorption.  At most second generation sources we have around 10^6 x-rays 
/second.  That means that only 5 x10^4 x-rays are measured in It for a sample 
that has a total absorption of 3.  Since the statistical noise goes like the 
square root of N, you will have 5% noise by measuring only 1 second per point, 
so you will have to measure about 25 sec per point to get usable EXAFS data.  
As the signal in It becomes smaller, you will find that the oscillatory part of 
the absorption spectra becomes damped because you can not accurately measure 
the fluctuations in the transmitted signal.  

 

If you have a lot of x-rays and the total absorption is dominated by the 
element that you measured, your data might be salvageable.  Look for a 
dampening affect.  If you see that when comparing to your other data then you 
are in trouble.  I don't know how to reliably fix it.  You can use Hephaestus 
to calculate the number of x-rays per second in It from the voltage and 
detector gain.

 

Cheers,

Shelly

 



From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On Behalf Of Stefan Mangold
Sent: Monday, April 28, 2008 11:46 AM
To: XAFS Analysis using Ifeffit
Subject: Re: [Ifeffit] Dodgy Edge Steps

 

Hi,

 

the edge step should be less than 1; the edge step at white line should be not 
higher than 1.5 and die maximum absorption should not be higher than 2.7 (max. 
3) (reason is the signal to noise ratio). The reason for limiting the edge jump 
are none linearities of your detector systems. So I would not use the data for 
further analysis.

 

best regards 

 

Stefan 

--
Dr. Stefan Mangold
Institut für Synchrotronstrahlung
Forschungszentrum Karlsruhe

 

  

 

Am 28.04.2008 um 18:27 schrieb LACHLAN MACLEAN:





Hi All

Long-term lurker, first time poster.

I've recently collected some EXAFS data with a large edge step (~3.4) and for 
some reason didn't pick up on it until it was too late.  I've been told that 
ideally we would want an edge step of 1 and anything above ~1.5-2 is too high.  
I'm wondering if there is a way to salvage the dataset in order to compare it 
with EXAFS data that I collected from two other samples (all three are supposed 
to be synthetic goethite).  The first two samples that I want to compare have 
edge steps of about 0.35.  

I know Athena allows one to adjust the edge step but wonder how appropriate 
this is and how much change this would cause the EXAFS data?  Trying to adjust 
the edge step from 3.4 down to ~1 or so seems to be quite a jump.  I'm curious 
whether there is a way to salvage this data rather than wait until the next 
beamrun.

Any help would be appreciated.

Cheers
Lachlan





___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit





 

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] How to generate the input information for Atom

2007-11-06 Thread Kelly, Shelly D.
Hi Fenglong,

You can do the next part of the analysis in Artemis.  I have a powerpoint 
tutorial showing the different parts of Artemis on the xafs.org wiki.  
http://www.xafs.org/Tutorials
It is called Basics_of_XAFS_analysis.pdf  It is a bit dated now, but the main 
points are all the same.

I recommend Matt's tutorials on that page.

Bruce has documentation for Artemis at 
http://cars9.uchicago.edu/~ravel/software/docs.html 

And Matt has a data base of crystal structure input files that are already in 
the correct format for Artemis at
http://cars9.uchicago.edu/~newville/adb/ 

Artemis comes with several working examples that you can find in the 
distribution package.

These links should get you going.
Cheers,
Shelly


From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On Behalf Of Fenglong Sun
Sent: Monday, November 05, 2007 6:00 PM
To: ifeffit@millenia.cars.aps.anl.gov
Subject: [Ifeffit] How to generate the input information for Atom

Hi, Bruce,
   I am working on an experiment of arsenic adsorption on pyrite particles. I 
have the data of mu(E)~E. I use Anthena to do the Fourier transform. Now I want 
to know the cordination information of arsenic atom. How can I get the input 
file for Atom if I want to do the fitting job by using Iffefit package? Thanks!

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


[Ifeffit] Re xanes feff8 sulfur

2007-10-15 Thread Kelly, Shelly D.
Hi Eckard,

 

I realize that you have posted a message every month for about 3 months
in a row without any reply:  Shame on us... I will try to answer some of
your questions and maybe that will help to get some discussion going.
The energy scale in feff8 is NOT defined as you would expect it to be
for determining the absolute energy of an edge.  The energy scale in
FEFF8 is related to the difference between the energy level of the
intersist (potential level inbetween the muffin tins that represent
atoms) and the threshold energy level for the excitation of the
electron.

 

I realize that this doesn't solve your problem, but hopefully it will
spark some discussion.

 

Cheers,

Shelly

 

 
Mon Aug 13 06:00:19 CDT 2007 
Does anybody know a person, who has experience respective the topics
xanes+feff8 and sulfur K-Edges?
I would be very thankful for an answer :-)
 
sincerely yours,
Eckhard Bosman
 

 

Tue Jul 10 10:07:38 CDT 2007 

Hello,
 
i have problems with feff-xanes calculations related to pyrite. I try to
get the k-edge of sulfur (as absorber) in FeS2. I get the edge at ~2472
eV
and i think this is wrong because i get the same energy with gypsum and
i
think there should be an energyshift related to 2472 eV. My proceeding
is
as follows:
 
calculate the atomic positions with the asymetric unit from the
http://rruff.geo.arizona.edu/AMS/amcsd.php database (Pyrite spacegroup
Pa3) respectively
http://rruff.geo.arizona.edu/AMS/xtal_data/CIFfiles/00596.cif
with powdercell 2.4
 
feffrun with the feff.inp-file like in the attachment
(thanks to Josh Kas for his feff-hints)
 
I started the feff-run on the one hand with 12 atoms (unitcell) and on
the
other hand with 22 atoms (14x Fe and 8x S in the cube)
 
After this i plotted from xmu.dat the absorbtion mu (column 4) as a
function of the energy (column 1). The result is in the attachment too.
 
I don't know why i get everytime the Sulfur-k-edge at the same position
(for pyrite or gypsum) respectively without energyshift related to 2472
eV.
 
many thanks for your help :-)
Ciao,
Eckhard

 

Fri Jun 15 08:48:21 CDT 2007

Hallo @all,
 
(first at all, sorry for my bad english)
 
i'm a student working on my diploma thesis (nexafs/xanes) and my job is
to
simulate with FEFF 8.4 spectra, for example absorption as a function of
energy (mu_(E)). I'm a newbie in this topic and i tried to get the
k-edge
of Sulfur for the substance gypsum (CaSO4 * 2(H2O)) and pyrite (FeS2).
I'm
interested in the XANES-range of the spectra not the EXAFS. I get
everytime my peaks at ~2472 eV for Sulfur as absorber whatever substance
i
use (gypsum, pyrite,...). There are no differences with the peak
position
on the energy-scale or shifts. I'm really confused :-( I think, there
should be differences in the spectra... I send you two feff.inp files
from
my feff-runs. My source is the mineralogy database
(http://rruff.geo.arizona.edu/AMS/amcsd.php) where i get the data for
the
unitcell. Then i put the data in the XtalDraw program to get the
positions
of the Atoms in Angstrom, by the way i get i nice 3D-View of the
unitcell
:-). Then i put the values of the positions into feff...
I simulated pyrite with 12 atoms and gypsum with 48.
I'm appreciative for every help :-)
 
ciao,
Eckhard

 

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] question about Eo and simultaneous fitting

2007-02-14 Thread Kelly, Shelly D.
Hi Tony,

The approach that I like to take, is to vary the choice of E0 in the
background subtraction step so that the chi(k) spectra are well aligned
with each other and the theory.  Then only one Ezero parameter is
needed.

Cheers,
Shelly


 -Original Message-
 From: [EMAIL PROTECTED] [mailto:ifeffit-
 [EMAIL PROTECTED] On Behalf Of Scott Calvin
 Sent: Wednesday, February 14, 2007 7:52 AM
 To: XAFS Analysis using Ifeffit
 Subject: Re: [Ifeffit] question about Eo and simultaneous fitting
 
 Hi Tony,
 
 A quick answer before I try to fight my way through an ice storm to
 get to work. :)
 
 Fitting the three spectra simultaneously makes sense, but if
 oxidation state is expected to be different, then you really need
 different E0's.
 
 As far as doping, it depends on the details. In some cases, it may
 require a different E0's, in others, not. Of course, you can try it
 both ways, and see what the fits think.
 
 --Scott Calvin
 Sarah Lawrence College
 
 At 05:04 AM 2/14/2007, you wrote:
 Hi,
 
 first I would like to thank all experienced  people answering in
 this forum!! I have learned really a lot from you.
 
 And now my question:
 Should I fit EXAFS spectra of a doped material and the same
 material, but oxidized and reduced (three spectra) simultaneously?
 In my opinion: yes, as I am looking for small differences. Is
 varying one Eo for the three samples the right approach?
 I am also wondering, should I vary one Eo for samples with different
 doping concentrations?
 
 
 
 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] question about Eo and simultaneous fitting

2007-02-14 Thread Kelly, Shelly D.

Hi Scott,

The change in frequency is a bigger affect at high k than the energy
shift parameter.  The energy shift parameter affects the data more
strongly at low k.  So I align the data to each other at low k, and then
they become out of phase with each other at high k.  A similar effect is
seen for pressure dependent data where the oxidation state doesn't
change but the bond length changes.  I also look at how well the data
are both aligned to the theory used to describe the data sets
independently.  The energy shift values should overlap within the
uncertainty to justify using a single parameter for them.  I suppose
that my method might not work for all systems.  But if the data are
similar enough so that a multiple data set fit has significant overlap
in the models used to describe the spectra, then it has worked.  

Shelly

 -Original Message-
 From: [EMAIL PROTECTED] [mailto:ifeffit-
 [EMAIL PROTECTED] On Behalf Of Scott Calvin
 Sent: Wednesday, February 14, 2007 9:55 AM
 To: XAFS Analysis using Ifeffit
 Subject: Re: [Ifeffit] question about Eo and simultaneous fitting
 
 Hi Shelly,
 
 Interesting--I've never tried that method with samples that I expect
 to differ in oxidation state. If it's differing in oxidation state,
 the local environment is probably different too...particularly,
 nearest-neighbor distances should be changed significantly. So how do
 you align the chi(k) data with each other, when they oscillate at a
 different frequency?
 
 --Scott Calvin
 Sarah Lawrence College
 
 At 10:38 AM 2/14/2007, you wrote:
 
 The approach that I like to take, is to vary the choice of E0 in the
 background subtraction step so that the chi(k) spectra are well
aligned
 with each other and the theory.  Then only one Ezero parameter is
 needed.
 
 Cheers,
 Shelly
 
 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


RE: [Ifeffit] Projekt with the flatten-problem

2007-01-11 Thread Kelly, Shelly D.
Hi Jens,

I believe the problem is related to the structure in the data before E0.
If you choose e0 to be at the beginning of the structure then the
normalization is handled better by Athena.  Take a look at the attached
project file.  The first data set with e0 on the edge as you would
choose for EXAFS analysis (e1) shows the error that Jens reported.  The
second data set with a lower value for e0 at the beginning of the
structure (e2) gets the job done.

Cheers,
Shelly

---
Shelly Kelly
Argonne National Laboratory
Bldg 203, RM E113
9700 S. Cass Ave
Argonne, IL 60439

Molecular Environmental Science Group
www.mesg.anl.gov
[EMAIL PROTECTED]
phone: 630-252-7376


 -Original Message-
 From: [EMAIL PROTECTED] [mailto:ifeffit-
 [EMAIL PROTECTED] On Behalf Of Jens Kruse
 Sent: Thursday, January 11, 2007 2:38 AM
 To: Athena mailing List
 Subject: [Ifeffit] Projekt with the flatten-problem
 
 Hi,
 
 find attached an example with the flatten-problem.
 thanks a lot,
 
 jens
 
 
 
 
 Hi,
 
 I have a normalization problem with Athena. If I turn on the flatten
 option in the background removal section, Athena creates in some of
the
 normalized spectra a step in the pre-edge range.  How can I  get rid
of
 it?
 
 Thanks,
 
 jens


Flatten.prj
Description: Flatten.prj
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


RE: [Ifeffit] Projekt with the flatten-problem

2007-01-11 Thread Kelly, Shelly D.
Good going Dave.  I think that you are correct.

Shelly

 -Original Message-
 From: [EMAIL PROTECTED] [mailto:ifeffit-
 [EMAIL PROTECTED] On Behalf Of Barton, David (DG)
 Sent: Thursday, January 11, 2007 10:49 AM
 To: 'XAFS Analysis using Ifeffit'
 Subject: RE: [Ifeffit] Projekt with the flatten-problem
 
 Jens,
 
 It appears to me that the E-shift is causing the problems with the
step in
 the pre-edge data. Shelly's data only appears to remove the step
because
 the pre-edge and the extrapolated post-edge plots are nearly parallel.
If
 you adjust the Normalization range in Shelly's data you will see the
step
 appear again.
 
 However, if you set the E-shift to zero and reset Eo, pre-edge range,
and
 normalization range (to your liking) you will see that the step is
goes
 away.  So the work around would be to subtract the E-shift from your
Raw
 data, then import into Athena and do not use any E-shift.
 
 Bruce or Matt may have to comment, but it appears to me that the
 normalization/flatten routine may not be dealing with the E-shift
 appropriately.
 
 Dave
 
 -Original Message-
 From: [EMAIL PROTECTED]
 [mailto:[EMAIL PROTECTED] On Behalf
 Of Kelly, Shelly D.
 Sent: Thursday, January 11, 2007 11:11 AM
 To: XAFS Analysis using Ifeffit
 Subject: RE: [Ifeffit] Projekt with the flatten-problem
 
 Hi Jens,
 
 I believe the problem is related to the structure in the data
 before E0.
 If you choose e0 to be at the beginning of the structure then
 the normalization is handled better by Athena.  Take a look at
 the attached project file.  The first data set with e0 on the
 edge as you would choose for EXAFS analysis (e1) shows the
 error that Jens reported.  The second data set with a lower
 value for e0 at the beginning of the structure (e2) gets the job
done.
 
 Cheers,
 Shelly
 
 ---
 Shelly Kelly
 Argonne National Laboratory
 Bldg 203, RM E113
 9700 S. Cass Ave
 Argonne, IL 60439
 
 Molecular Environmental Science Group
 www.mesg.anl.gov
 [EMAIL PROTECTED]
 phone: 630-252-7376
 
 
  -Original Message-
  From: [EMAIL PROTECTED] [mailto:ifeffit-
  [EMAIL PROTECTED] On Behalf Of Jens Kruse
  Sent: Thursday, January 11, 2007 2:38 AM
  To: Athena mailing List
  Subject: [Ifeffit] Projekt with the flatten-problem
 
  Hi,
 
  find attached an example with the flatten-problem.
  thanks a lot,
 
  jens
 
 
 
 
  Hi,
 
  I have a normalization problem with Athena. If I turn on the
 flatten
  option in the background removal section, Athena creates in some of
 the
  normalized spectra a step in the pre-edge range.  How can I  get
rid
 of
  it?
 
  Thanks,
 
  jens
 
 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


RE: [Ifeffit] Is there a physical meaning to a negative SO2

2007-01-09 Thread Kelly, Shelly D.
Hi Todd,

One thing to try is to set s02 to a positive value and all the other
variables to reasonable values and then look at the real or imaginary
parts of the Fourier transform and also look at the chi(k) spectra.  You
should see the affect that Scott mentioned.

Shelly


 -Original Message-
 From: [EMAIL PROTECTED] [mailto:ifeffit-
 [EMAIL PROTECTED] On Behalf Of Scott Calvin
 Sent: Tuesday, January 09, 2007 10:27 AM
 To: 'XAFS Analysis using Ifeffit'
 Subject: Re: [Ifeffit] Is there a physical meaning to a negative SO2
 
 Hi Todd,
 
 Your paragraph below is correct; i.e. you do not have a physically
 realistic model. One common cause of a negative S02 is that the model
 also is off by half an oscillation. After all, the negative S02 turns
 the signal upside down, shifting it by half an oscillation can then
 cause a rough (but spurious) alignment between model and data.
 
 In any case, the model needs work; depending on your system and
 degree of prior knowledge, you might be assuming the wrong species of
 scattering atom, for example.
 
 --Scott Calvin
 Sarah Lawrence College
 
 At 10:41 AM 1/9/2007, Todd Luxton wrote:
 I am very new to XAFS analysis, I am wondering if there is a
 physical meaning to a negative value for SO2.  Based on the XAFS
 equation I can't understand what the physical significance of a
 negative SO2 value would be, therefore I am assuming that my
 modeling results using Athena are not representative of a physically
 realistic model due to the negative value calculated fro SO2 or more
 specifically the amp variable in the guess set.  Any insight would
 be greatly appreciated.
 
 
 
 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


RE: [Ifeffit] Generating Atom lists using Artemis

2007-01-08 Thread Kelly, Shelly D.


---
Shelly Kelly
Argonne National Laboratory
Bldg 203, RM E113
 
 On Saturday 06 January 2007 10:53, Kelly, Shelly D. wrote:
  Hello Artemis users,
 
  I found an interesting feature in Artemis when generating a list of
 atoms
  from the attached cif file worth sharing on the list.  In the
attached
 cif
  file for Pd2Si the position of the Si2 atoms are at x,y,z = 0.,
 0.6667,
  0 with a space group of p -6 2 m. Using this cif file and choosing
 either
  Pd1 or Pd2 as the core atom results in errors in the atom list.  For
  example for the Pd2 site there are 4 Si neighbors with the 1st and
3rd
 and
  the 2nd and 4th on top of each other.  If you edit the atomic
positions
 so
  that the Si2 atoms are correct to an additional digit x,y,z =
0.3,
  0.7,0 then the atom list is fixed with only two Pd2-Si neighbors
at
  each of the unique positions.  The interesting feature is that the
  coordinates on the atoms page need to be correct to the 5 decimal
place.
 
 
 This has been discussed many, many times on the mailing list and is
 explained in question 3 at
http://cars9.uchicago.edu/iffwiki/FAQ/RunningFeff
 
 The version of Feff that ships with Ifeffit (and, I believe, all
 others) uses 5 significant figures in the path finder.  Since Atoms
 main purpose is to interact with Feff, it interprets the 5 digit thing
 quite strictly.
 
 B
 
 

[sdk] Hi Bruce,

Well, I thought that I had skipped your here is a link describing that
response by searching the Ifeffit wiki.  I didn't run FEFF, so I missed
the description of the problem under the FEFF error message of
overlapping atoms.   

I think that it would be helpful for the ATOMS page within Artemis to
contain a note of some sort indicating that the 5th digit is
significant.  Now that I have had additional time to ponder this
problem, I realize that I have learned it several times only to forget
it again:)

Cheers,
Shelly

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


[Ifeffit] Generating Atom lists using Artemis

2007-01-06 Thread Kelly, Shelly D.
Hello Artemis users,

I found an interesting feature in Artemis when generating a list of atoms from 
the attached cif file worth sharing on the list.  In the attached cif file for 
Pd2Si the position of the Si2 atoms are at x,y,z = 0., 0.6667, 0 with a 
space group of p -6 2 m. Using this cif file and choosing either Pd1 or Pd2 as 
the core atom results in errors in the atom list.  For example for the Pd2 site 
there are 4 Si neighbors with the 1st and 3rd and the 2nd and 4th on top of 
each other.  If you edit the atomic positions so that the Si2 atoms are correct 
to an additional digit x,y,z = 0.3, 0.7,0 then the atom list is fixed 
with only two Pd2-Si neighbors at each of the unique positions.  The 
interesting feature is that the coordinates on the atoms page need to be 
correct to the 5 decimal place.

Cheers,
Shelly