RE: Negative Uiso in GSAS

2010-03-04 Thread Alan Hewat
Michael Glazer said:
> Rietveld refinement (as opposed to single-crystal refinement) is in fact
> refinement of degraded data (it is one-dimensional instead of
> three-dimensional) and so the errors will be more significant.

While I agree with Mike's list, I can't let him get away with such a
sweeping statement implying that single crystal refinements necessarily
give better structures, simply because "the data is three dimensional".

Errors in crystal refinement arise from all kinds of effects apart from
problems due to overlapping peaks. Extinction is a cause for error that is
usually more serious for single crystals than for powders - especially
with strain associated with a structural transition. Texture is a problem
associated mainly with X-ray powder diffraction etc...

With todays high resolution powder diffractometers and Rietveld
refinement, I don't even believe that overlapping reflections from "one
dimensional data" is the main cause of error. Its true that if peaks are
strongly overlapping at high angles then refining the background may
result in it being overestimated, leading to higher temperature factors.
Refinement of absorption will also strongly correlate with temperature
factors.

That is why you should not keep adding more parameters such as background,
absorption etc simply to obtain a lower R-factor and expect that to equate
to a better structure determination. You should try to actually measure
these other effects and correct for them, not simply refine them away.
__
Dr Alan Hewat, NeutronOptics, Grenoble, FRANCE
 +33.476.98.41.68
  http://www.NeutronOptics.com/hewat
__



Re: Negative Uiso in GSAS

2010-03-03 Thread Reinhard Kleeberg
To add some more simple reasons for apparently overmuch high intensity 
at high angles causing meaningless negative temperature factors:
- the low angle peak intensities are suppressed by beam overflow 
(typically fixed slit Bragg Brentano geometry)
- one is using automatic slit data instead of fixed slit ones, and the 
program used is unable to account for that
- if one has used converted ADS data to fixed slit data: The conversion 
failed because of any wrong background correction, or the primary ADS 
data are wrong because of misprogramming or misalignement of the ADS system.


And one addition to Jon's important point 5 below: This is often related 
to physically meaningless peak shape models, running out of meaningful 
peak width values, just by becoming broader and broader and "taking 
intensity" from the background into the high angle peak intensities. One 
should use a physically based and more rigid peak shape model.


Regards

Reinhard


Jon Wright schrieb:

I can't resist adding one more to Mike's excellent list:

5. When peaks overlap strongly it becomes difficult to determine the 
background level. Negative Uiso is a consequence of the background 
refining to a value which is too low, especially where the peaks are 
most dense in the pattern (shorter d-spacings or higher angles).


All the best,

Jon

Michael Glazer wrote:

Negative U's in Rietveld can arise from several causes, so that there is
not one single answer. Some of the reasons are
1. The structural model is simply incorrect.
2. High absorption means that the low-angle data are weaker than they
should be, or conversely that the high-angle data appear stronger than
they should be. Abnormally strong high-angle data give rise to a
decrease in U's, even making them appear negative
3. Correlation between the refinement parameters. For instance U's will
tend to be highly correlated with site occupation parameters, often
making it difficult to separate them.
4. In general, many of the errors that one encounters tend to end up in
the refined U's, and this is why their precise values have to be treated
with caution.
Rietveld refinement (as opposed to single-crystal refinement) is in fact
refinement of degraded data (it is one-dimensional instead of
three-dimensional) and so the errors will be more significant.

Mike Glazer


-Original Message-
From: carolina.zip...@fi.isc.cnr.it
[mailto:carolina.zip...@fi.isc.cnr.it] Sent: 03 March 2010 16:08
To: rietveld_l@ill.fr
Subject: Negative Uiso in GSAS

Dear all,

could someone explain to me the meaning of obtaining a negative Uiso in
GSAS?
I thought it was always positive...(p. 123 manual)

thanks

Carolina


_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-


  Dr. Carolina Ziparo

   Istituto dei Sistemi Complessi - sezione di Firenze,
   C.N.R. - Consiglio Nazionale delle Ricerche

   via Madonna del Piano, 10
   I-50019 Sesto Fiorentino Italy


   tel.:   +39 055 5226693
   fax:+39 055 5226683
   e-mail: carolina.zip...@fi.isc.cnr.it



  






Re: Negative Uiso in GSAS

2010-03-03 Thread Jon Wright

I can't resist adding one more to Mike's excellent list:

5. When peaks overlap strongly it becomes difficult to determine the 
background level. Negative Uiso is a consequence of the background 
refining to a value which is too low, especially where the peaks are 
most dense in the pattern (shorter d-spacings or higher angles).


All the best,

Jon

Michael Glazer wrote:

Negative U's in Rietveld can arise from several causes, so that there is
not one single answer. Some of the reasons are
1. The structural model is simply incorrect.
2. High absorption means that the low-angle data are weaker than they
should be, or conversely that the high-angle data appear stronger than
they should be. Abnormally strong high-angle data give rise to a
decrease in U's, even making them appear negative
3. Correlation between the refinement parameters. For instance U's will
tend to be highly correlated with site occupation parameters, often
making it difficult to separate them.
4. In general, many of the errors that one encounters tend to end up in
the refined U's, and this is why their precise values have to be treated
with caution.
Rietveld refinement (as opposed to single-crystal refinement) is in fact
refinement of degraded data (it is one-dimensional instead of
three-dimensional) and so the errors will be more significant.

Mike Glazer


-Original Message-
From: carolina.zip...@fi.isc.cnr.it
[mailto:carolina.zip...@fi.isc.cnr.it] 
Sent: 03 March 2010 16:08

To: rietveld_l@ill.fr
Subject: Negative Uiso in GSAS

Dear all,

could someone explain to me the meaning of obtaining a negative Uiso in
GSAS?
I thought it was always positive...(p. 123 manual)

thanks

Carolina


_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-


  Dr. Carolina Ziparo

   Istituto dei Sistemi Complessi - sezione di Firenze,
   C.N.R. - Consiglio Nazionale delle Ricerche

   via Madonna del Piano, 10
   I-50019 Sesto Fiorentino Italy


   tel.:   +39 055 5226693
   fax:+39 055 5226683
   e-mail: carolina.zip...@fi.isc.cnr.it



  




RE: Negative Uiso in GSAS

2010-03-03 Thread Michael Glazer
Negative U's in Rietveld can arise from several causes, so that there is
not one single answer. Some of the reasons are
1. The structural model is simply incorrect.
2. High absorption means that the low-angle data are weaker than they
should be, or conversely that the high-angle data appear stronger than
they should be. Abnormally strong high-angle data give rise to a
decrease in U's, even making them appear negative
3. Correlation between the refinement parameters. For instance U's will
tend to be highly correlated with site occupation parameters, often
making it difficult to separate them.
4. In general, many of the errors that one encounters tend to end up in
the refined U's, and this is why their precise values have to be treated
with caution.
Rietveld refinement (as opposed to single-crystal refinement) is in fact
refinement of degraded data (it is one-dimensional instead of
three-dimensional) and so the errors will be more significant.

Mike Glazer


-Original Message-
From: carolina.zip...@fi.isc.cnr.it
[mailto:carolina.zip...@fi.isc.cnr.it] 
Sent: 03 March 2010 16:08
To: rietveld_l@ill.fr
Subject: Negative Uiso in GSAS

Dear all,

could someone explain to me the meaning of obtaining a negative Uiso in
GSAS?
I thought it was always positive...(p. 123 manual)

thanks

Carolina


_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-


  Dr. Carolina Ziparo

   Istituto dei Sistemi Complessi - sezione di Firenze,
   C.N.R. - Consiglio Nazionale delle Ricerche

   via Madonna del Piano, 10
   I-50019 Sesto Fiorentino Italy


   tel.:   +39 055 5226693
   fax:+39 055 5226683
   e-mail: carolina.zip...@fi.isc.cnr.it





Re: Negative Uiso in GSAS

2010-03-03 Thread Jean-Gabriel ROQUEFERE



  Dear Carolina,

  I have the answer. It's about the electron cloud of your elements.  
I'll tell you further tomorrow because I'm at home right now. I speak  
about this feature in one of my publications. You will have the answer  
tomorrow, if nobody replies.


  Best regards.

--
Dr. Jean-Gabriel ROQUEFERE
Institut de Recherche sur l'Hydrogène
Département Physique
(1)-819-701-3603
www.irhcanada.com

carolina.zip...@fi.isc.cnr.it a écrit :


Dear all,

could someone explain to me the meaning of obtaining a negative Uiso in GSAS?
I thought it was always positive...(p. 123 manual)

thanks

Carolina


_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-


  Dr. Carolina Ziparo

   Istituto dei Sistemi Complessi - sezione di Firenze,
   C.N.R. - Consiglio Nazionale delle Ricerche

   via Madonna del Piano, 10
   I-50019 Sesto Fiorentino Italy


   tel.:   +39 055 5226693
   fax:    +39 055 5226683
   e-mail: carolina.zip...@fi.isc.cnr.it




-
Courriel expédié via https://courriel.uqtr.ca



Re: Negative Uiso in GSAS

2010-03-03 Thread Goutenoire

As mentioned recently,

Their is no physical meaning to have negative thermal agitation factor.
It means that at high angle the observed data at upon the calculated.

It means that you should put more electron (if you are doing X-Ray 
diffraction) or more electron or nuclear density in your 
crystallographic model.


One of the usual problem could be:
- You have used some automatic variable slit !! and you try to refined 
such data without correction.

- Check your crystallographic model

But if you have done such last things it should come from some 
absorption problem : if you are using X-ray with high absorption 
compound in reflection geometry this problem come quite often.


For example the simple crystallographic compound metal Platinum simple 
cubic Fm-3m (N°225) Pt in 0,0,0


The absorption is big for copper radiation 4300cm-1.

I find -0.6ang2 with X'Pert HighScore Plus.

Normally we can make some correction for such absorption see :


 /Référence/

/*20.- W. Pitschke, N. Mattern and H. Hermann, Powder Diffraction 8(4),*/

/* 223-228 (1993).*/

/ (In this paper the authors refine simultaneously all occupation/

/ numbers. This practice should be avoided. If the calculation were/

/ exact that gives rise to a singular matrix when the scale factor/

/ is also refined)/

/ See also/

/* W. Pitschke, H. Hermann and N. Mattern, Powder Diffraction 8(2),*/

/* **74-83 (1993).*/

/ /

/ Dans HighScore on trouve la reference :/

//H. HERMAN & M. ERMRICH///, Microabsorption of X-ray intensity in 
randomly packed powder specimens, Acta Cryst. (1987), *A43*, 401 -- 405./


But I did not find so easy to apply such correction at least with 
HighScore or Fullprof.


The data of the platinum or the molybdenum could be found in :



http://perso.univ-lemans.fr/~fgouten/RX_data

Best Whishes
François Goutenoire


Le 03/03/2010 17:08, carolina.zip...@fi.isc.cnr.it a écrit :

Dear all,

could someone explain to me the meaning of obtaining a negative Uiso 
in GSAS?

I thought it was always positive...(p. 123 manual)

thanks

Carolina


_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-_-


 Dr. Carolina Ziparo

  Istituto dei Sistemi Complessi - sezione di Firenze,
  C.N.R. - Consiglio Nazionale delle Ricerche

  via Madonna del Piano, 10
  I-50019 Sesto Fiorentino Italy


  tel.:   +39 055 5226693
  fax:+39 055 5226683
  e-mail: carolina.zip...@fi.isc.cnr.it





--
Dr. François Goutenoire
Laboratoire des Oxydes et Fluorures UMR6010
tel: 02.43.83.33.53