Re: RE: error in peak positions that is linear in theta

2009-01-13 Thread prasunsc
Dear Mr. Peter
I also faced the same problem as Kurt faced, our diffractometer is the 
D8-advance model from Bruker... I took the scan using a PSD detector(Lynxeye) 
and got an increase of the error in peak position with increasing 2theta.. 
Again, when I extracted the dat file from the raw one i observed the 
rounded up values of 2theta(up to 2 decimal pt.) in the file. Now will the 
problem be optimized if I generate 2Theta values using a step which is rounded 
up upto a higher number of decimal point and replace them in place of my 2theta 
column? thanking you..

prasun



- Original Message -
From: Peter Y. Zavalij pzava...@umd.edu
Date: Friday, January 9, 2009 7:17 pm
Subject: RE: error in peak positions that is linear in theta

 Kurt,
 
 It looks like incorrect step size (difference between actual step 
 size and
 one that is written in the file) or incorrect distance to the 
 detector. The
 later may incorrect step size if detector is PSD.
 
 I have similar situation with much smaller difference in peak 
 position but
 the shift (difference between actual and written step size) varies 
 frommeasurement to measurement. Bruker is working on this but so 
 far no
 solution/cause is known.
 
 Your case may be simpler if the difference is constant. Perhaps 
 calibrationof detector or its distance to the sample is needed.
 
 Best,
 
 
 
 Peter Zavalij 
 
 
 
 X-ray Crystallographic Center
 University of Maryland
 
 College Park, MD
 
 
 
 Office: (301)405-1861
 Lab: (301)405-3230
 
 Fax: (301)314-9121
 
 
 
 From: Kurt Leinenweber [ku...@asu.edu] 
 Sent: Thursday, January 08, 2009 6:09 PM
 To: Rietveld_l@ill.fr
 Subject: error in peak positions that is linear in theta
 
 
 
 Dear Rietvelders,
 
 
 
 (I hope this list is appropriate for this question).  I am trying 
 to set up
 a new D8 diffractometer from Bruker, and have it set up with a Vario
 monochromator focusing on a sample in a capillary.  Of course I 
 hope to do
 Rietveld refinement on the samples eventually.  But there is an 
 error in two
 theta of the peak positions on all samples, and the error is 
 linear in two
 theta.  At 40 degrees the error (measured minus expected) is about 
 0.020degrees, at 80 degrees the error is about 0.040 degrees, 
 etcetera.
 
 
 Has anyone ever seen anything like this and do you know where it 
 could come
 from?
 
 
 
 Thank you very much,
 
 
 
- Kurt
 
 
 
 
 
 *
 
 Kurt Leinenweber
 
 Dept. of Chemistry and Biochemistry
 
 Arizona State University
 
 Tempe, AZ 85287-1604
 
 
 
 phone (480)-965-8853
 
 fax (480)-965-2747
 
 
 
 ***
 
 
 
 



RE: error in peak positions that is linear in theta

2009-01-12 Thread Yokochi, Alexandre
Hi all,

Does the error maximize at 2-theta = 90 then decrease again?  Back when
I was an actual crystallographer and used an INEL system, this was
observed when the capillary position was not coaxial with the detector
(in fact, when it was behind the proper position).

Alex Y


Dr. Alexandre (Alex) F. T. Yokochi
Assistant Professor
Department of Chemical Engineering
Oregon State University
Corvallis, OR 97331-2702

Voice:  (541) 737-9357
Fax:  (541) 737-4600
http://oregonstate.edu/~yokochia

 -Original Message-
  *From:* Kurt Leinenweber [mailto:ku...@asu.edu]
  *Sent:* Thursday, January 08, 2009 6:09 PM
  *To:* Rietveld_l@ill.fr
  *Subject:* error in peak positions that is linear in theta
 
 
 
  Dear Rietvelders,
 
 
 
  (I hope this list is appropriate for this question).  I am trying to
set
  up a new D8 diffractometer from Bruker, and have it set up with a
Vario
  monochromator focusing on a sample in a capillary.  Of course I hope
to
  do Rietveld refinement on the samples eventually.  But there is an
error
  in two theta of the peak positions on all samples, and the error is
  linear in two theta.  At 40 degrees the error (measured minus
expected)
  is about 0.020 degrees, at 80 degrees the error is about 0.040
degrees,
  etcetera.
 
 
 
  Has anyone ever seen anything like this and do you know where it
could
  come from?
 
 
 
  Thank you very much,
 
 
 
  - Kurt
 
 
 
 
 
  *
 
  Kurt Leinenweber
 
  Dept. of Chemistry and Biochemistry
 
  Arizona State University
 
  Tempe, AZ 85287-1604
 
 
 
  phone (480)-965-8853
 
  fax (480)-965-2747
 
 
 
  ***
 
 
 
 




RE: error in peak positions that is linear in theta

2009-01-09 Thread Peter Y. Zavalij
Kurt,

It looks like incorrect step size (difference between actual step size and
one that is written in the file) or incorrect distance to the detector. The
later may incorrect step size if detector is PSD.

I have similar situation with much smaller difference in peak position but
the shift (difference between actual and written step size) varies from
measurement to measurement. Bruker is working on this but so far no
solution/cause is known.

Your case may be simpler if the difference is constant. Perhaps calibration
of detector or its distance to the sample is needed.

Best,

 

Peter Zavalij 



X-ray Crystallographic Center
University of Maryland

College Park, MD



Office: (301)405-1861
Lab: (301)405-3230

Fax: (301)314-9121



From: Kurt Leinenweber [mailto:ku...@asu.edu] 
Sent: Thursday, January 08, 2009 6:09 PM
To: Rietveld_l@ill.fr
Subject: error in peak positions that is linear in theta

 

Dear Rietvelders,

 

(I hope this list is appropriate for this question).  I am trying to set up
a new D8 diffractometer from Bruker, and have it set up with a Vario
monochromator focusing on a sample in a capillary.  Of course I hope to do
Rietveld refinement on the samples eventually.  But there is an error in two
theta of the peak positions on all samples, and the error is linear in two
theta.  At 40 degrees the error (measured minus expected) is about 0.020
degrees, at 80 degrees the error is about 0.040 degrees, etcetera.

 

Has anyone ever seen anything like this and do you know where it could come
from?

 

Thank you very much,

 

- Kurt

 

 

*

Kurt Leinenweber

Dept. of Chemistry and Biochemistry

Arizona State University

Tempe, AZ 85287-1604

 

phone (480)-965-8853

fax (480)-965-2747

 

***

 



Re: error in peak positions that is linear in theta

2009-01-08 Thread Leonid Solovyov
Dear Kurt,

It seems that the problem is in wrong sample-to-detector distance which results 
in the nearly linear angular dependence of the peak displacement.

Regards,
Leonid

*
 Leonid A. Solovyov
 Institute of Chemistry and Chemical Technology
 660049, K. Marx 42, Krasnoyarsk, Russia
 www.icct.ru/eng/content/persons/Sol_LA
 www.geocities.com/l_solovyov
*

--- On Thu, 1/8/09, Kurt Leinenweber ku...@asu.edu wrote:

 (I hope this list is appropriate for this question).  I am
 trying to set
 up a new D8 diffractometer from Bruker, and have it set up
 with a Vario
 monochromator focusing on a sample in a capillary.  Of
 course I hope to
 do Rietveld refinement on the samples eventually.  But
 there is an error
 in two theta of the peak positions on all samples, and the
 error is
 linear in two theta.  At 40 degrees the error (measured
 minus expected)
 is about 0.020 degrees, at 80 degrees the error is about
 0.040 degrees,
 etcetera.
 
  
 
 Has anyone ever seen anything like this and do you know
 where it could
 come from?
 
  
 
 Thank you very much,
 
  
 

 - Kurt
 
  
 
  
 
 *
 
 Kurt Leinenweber
 
 Dept. of Chemistry and Biochemistry
 
 Arizona State University
 
 Tempe, AZ 85287-1604
 
  
 
 phone (480)-965-8853
 
 fax (480)-965-2747
 
  
 
 ***