Re: some uncertainties regarding the Rietveld refinement results in TOPAS
Hi Daniel The peak shape variation comes from exponential absorption of the incident and directed X-rays due to their interaction below the specimen's surface. In the case of a finite thickness sample, this exponential gets truncated. I'm pretty sure I talk about this in Rowles & Buckley. Intensity is the main thing. Peak shape is going to be dominated by crystallite size/strain, but in its absence, youll see cut off in the peak trails due to finite thickness. Matthew On Thu, 7 Nov 2024, 16:39 Daniel Chateigner, wrote: > Dear All, > > Just to be sure on two of the Q&As: > Le 07/11/2024 à 05:03, Matthew Rowles a écrit : > > Great questions! > > > > Is it worthy to use the parameter known as "Sample Thickness", > associated with the Absorption Correction in TOPAS? > > Generally, no. You only need to worry about it if the penetration depth of > your incident beam is larger than the thickness (t) of the specimen. If > that is the case, then "Sample Thickness" and "Scale Intensity" is > necessary to correct for the change in peak shape due to the finite depth, > and the change in diffracting volume. > > - Seems from your answer Matthew, that diffracting volume changes (Vd) > through the scan in BB geometry gives rise to some peak shape variations (I > assume here that "shape" includes "width"). I agree that the diffracting > volume changes with theta, and with the sample thickness, but these two > effects should not contribute the same on the peak shape: > > -- For a given sample thickness, the diffracting volume varies with Theta, > but to me the main contribution to the peak shape is due to the irradiated > surface (beam track on the sample surface). Since deeper dVd will be more > absorbed than the ones closer to the surface. You can picture this > measuring a 2Theta scan at fixed incident angle, up to 2Theta=90°, the peak > width increases notably with 2Theta because of this. > > -- For the same theta on the diffraction diagram, varying the sample > thickness will probably not modify this much the shape of the peaks, but > intensities. For the same reasons on dVd. For instance, if measuring 2 > samples with a 100 nm and 200 nm thickness respectively, the peak shape > variation will probably not be visible using lab instruments, while the > intensities will differ significantly. > > - I do not like much the term "penetration depth" (p), since the depth > reached by x-rays is always infinite. For instance, if as usual p is > defined as 1/mu (mu = absorption coefficient of the sample, the one we > never fully know exactly), using a Theta-2Theta scan in BB on a t=infinity > sample, the measured diffracted intensity is larger than the one coming > from p only. And, on a thin specimen, both absorption and Vd are varying > during the scans, each of them bringing a contribution to the intensity. On > the contrary, for t=infinite specimen, they compensate each-other and their > contribution is the same on the diffracted intensity throughout the diagram > (but the intensity does not come from the same irradiated volume). I was > then wondering if Topas was simply corrected from volume variations or also > including the absorption variations in the case of "thin enough" samples ? > > > > > > Is the negative intensity in the difference curve... > > It is showing that there is some mismatch between your model and the data. > The intensities could be off because the peaks positions are wrong, the > peak-shape is wrong, the atomic displacement parameters are wrong (don't > refine these from normal QPA data), the LP factor correction is wrong, the > site occupancies are wrong, the atomic scattering factors are wrong. Could > also be that you have forgotten to refine a parameter that you should be. > > What about preferred orientations ? Anisotropic crystallite shapes and > microstrains ? Improperly defined film thickness can give rise to > variations of the ADPs also. > > > My 2€cts ! > > Daniel > > > > > > > > > > > > On Wed, 6 Nov 2024 at 19:12, Łukasz Kruszewski > wrote: > >> Dear Rietvelders, >> >> after some years of being a happy user of TOPAS & Rietveld method (after >> working-out a good approach, thanks to a great help from users of this >> mailing list for which I am really thanful, and having it checked by >> attendance in Reynolds Cup) I was recently confronted with few questions >> I couldn't answer. I thus have a kind request of a help here: >> >> - does the Absorption correction also influences microabsorption? This >> question arose due to somewhat lowered expected wt.% of pyrite (19 >> instead of 25), FeS2, even though all the used parameters (Absorption, >> CrySize, Strain - for all the sample components) were checked and proven >> to give value > error, i.e., being physically meaningful for the >> refinement model. I was wondering how this supposed non-precision for >> pyrite could be addressed... >> >> - Is it worthy to use the parameter known as "Sample Thickness", >> associated with the
Re: [External] Re: some uncertainties regarding the Rietveld refinement results in TOPAS
I agree with everything Dave said. I was a little terse in my response. Physically, altering a unit cell's contents is going to alter its size and the intensities. On Thu, 7 Nov 2024, 20:49 Bish, David L, wrote: > Matthew has given a nice summary for you, but I'd like to add one thing. > Whenever the site occupancies change, the unit-cell parameters will/should > change (unless the substituting ions have identical radii). You can > actually use the refined unit-cell parameters to estimate compositions in > many cases. You should always refine the uc parameters (with the exception > of a certified standard), and if your materials' compositions change, you > can expect the uc parameters to vary. Even if they change by only a small > amount, they can have a significant effect on the fit. I would not fix the > uc parameters in a refinement unless you are working with materials of a > known, fixed composition, e.g., synthetic, pure materials. Having said > this, if the substitutions are very minor, e.g., ppm level, you will likely > not see any effect. But, as you said, reflection positions and thus uc > parameters vary even for the same materials listed in the PDF and the COD. > > Also, keep in mind that the extent of the effect of site occupancy > variations on intensities will depend on the scattering factors of the > substituting ions. Si-Al substitutions, for example, in minerals such as > feldspars will have a negligible effect on intensities but a larger effect > on uc parameters. On the other hand, Mg-Fe substitution in the olivine > minerals will affect both intensities and uc parameters, and you can use > either or both of these to estimate composition. > > Best of luck with your refinements. > > Dave > -- > *From:* [email protected] on behalf > of Matthew Rowles > *Sent:* Wednesday, November 6, 2024 11:03 PM > *To:* Łukasz Kruszewski > *Cc:* Rietveld > *Subject:* [External] Re: some uncertainties regarding the Rietveld > refinement results in TOPAS > > > This message was sent from a non-IU address. Please exercise caution when > clicking links or opening attachments from external sources. > > Great questions! > > > does the Absorption correction also influences microabsorption? > > No. The absorption correction in TOPAS is only a peak-shape correction to > account for the transparency of the specimen. There is no real way to > correct for microabsorption beyond grinding to a fine particle size, > changing your wavelength, or moving to neutrons. > > > > Is it worthy to use the parameter known as "Sample Thickness", > associated with the Absorption Correction in TOPAS? > > Generally, no. You only need to worry about it if the penetration depth of > your incident beam is larger than the thickness of the specimen. If that is > the case, then "Sample Thickness" and "Scale Intensity" is necessary to > correct for the change in peak shape due to the finite depth, and the > change in diffracting volume. > > > >Is there any difference, to be expected, in the calculated unit cell > parameters, between approximating a phase using (1) a chemically pure > STRUCTURE model, and (2) a STRUCTURE with Occupancy information updated > with the known content of impurities? > > Altering site occupancies should only affect peak intensities, and so unit > cell parameters should only be a second-order effect. (disregarding the > physicality of various unit cell contents vs dimension pairings) > > > > Is the negative intensity in the difference curve... > > It is showing that there is some mismatch between your model and the data. > The intensities could be off because the peaks positions are wrong, the > peak-shape is wrong, the atomic displacement parameters are wrong (don't > refine these from normal QPA data), the LP factor correction is wrong, the > site occupancies are wrong, the atomic scattering factors are wrong. Could > also be that you have forgotten to refine a parameter that you should be. > > > > We have been working using few computers with, supposedly the very same > version of the software > > That does seem concerning. There can be differences between different > versions of TOPAS. You would also need to verify that all the scattering > factors, wavelengths, MACs, and space group files are the same for all > installations, and that nothing has been incorrectly edited. Also that > TOPAS.inc hasn't been touched. > > > Hope this helps. > > > Matthew > > > > > > > > > > > On Wed, 6 Nov 2024 at 19:12, Łukasz Kruszewski > wrote: > > Dear Rietvelders, > > after some yea
Re: some uncertainties regarding the Rietveld refinement results in TOPAS
Dear All, Just to be sure on two of the Q&As: Le 07/11/2024 à 05:03, Matthew Rowles a écrit : Great questions! > Is it worthy to use the parameter known as "Sample Thickness", associated with the Absorption Correction in TOPAS? Generally, no. You only need to worry about it if the penetration depth of your incident beam is larger than the thickness (t) of the specimen. If that is the case, then "Sample Thickness" and "Scale Intensity" is necessary to correct for the change in peak shape due to the finite depth, and the change in diffracting volume. - Seems from your answer Matthew, that diffracting volume changes (Vd) through the scan in BB geometry gives rise to some peak shape variations (I assume here that "shape" includes "width"). I agree that the diffracting volume changes with theta, and with the sample thickness, but these two effects should not contribute the same on the peak shape: -- For a given sample thickness, the diffracting volume varies with Theta, but to me the main contribution to the peak shape is due to the irradiated surface (beam track on the sample surface). Since deeper dVd will be more absorbed than the ones closer to the surface. You can picture this measuring a 2Theta scan at fixed incident angle, up to 2Theta=90°, the peak width increases notably with 2Theta because of this. -- For the same theta on the diffraction diagram, varying the sample thickness will probably not modify this much the shape of the peaks, but intensities. For the same reasons on dVd. For instance, if measuring 2 samples with a 100 nm and 200 nm thickness respectively, the peak shape variation will probably not be visible using lab instruments, while the intensities will differ significantly. - I do not like much the term "penetration depth" (p), since the depth reached by x-rays is always infinite. For instance, if as usual p is defined as 1/mu (mu = absorption coefficient of the sample, the one we never fully know exactly), using a Theta-2Theta scan in BB on a t=infinity sample, the measured diffracted intensity is larger than the one coming from p only. And, on a thin specimen, both absorption and Vd are varying during the scans, each of them bringing a contribution to the intensity. On the contrary, for t=infinite specimen, they compensate each-other and their contribution is the same on the diffracted intensity throughout the diagram (but the intensity does not come from the same irradiated volume). I was then wondering if Topas was simply corrected from volume variations or also including the absorption variations in the case of "thin enough" samples ? > Is the negative intensity in the difference curve... It is showing that there is some mismatch between your model and the data. The intensities could be off because the peaks positions are wrong, the peak-shape is wrong, the atomic displacement parameters are wrong (don't refine these from normal QPA data), the LP factor correction is wrong, the site occupancies are wrong, the atomic scattering factors are wrong. Could also be that you have forgotten to refine a parameter that you should be. What about preferred orientations ? Anisotropic crystallite shapes and microstrains ? Improperly defined film thickness can give rise to variations of the ADPs also. My 2€cts ! Daniel On Wed, 6 Nov 2024 at 19:12, Łukasz Kruszewski wrote: Dear Rietvelders, after some years of being a happy user of TOPAS & Rietveld method (after working-out a good approach, thanks to a great help from users of this mailing list for which I am really thanful, and having it checked by attendance in Reynolds Cup) I was recently confronted with few questions I couldn't answer. I thus have a kind request of a help here: - does the Absorption correction also influences microabsorption? This question arose due to somewhat lowered expected wt.% of pyrite (19 instead of 25), FeS2, even though all the used parameters (Absorption, CrySize, Strain - for all the sample components) were checked and proven to give value > error, i.e., being physically meaningful for the refinement model. I was wondering how this supposed non-precision for pyrite could be addressed... - Is it worthy to use the parameter known as "Sample Thickness", associated with the Absorption Correction in TOPAS? I was taught in a 5-day course of TOPAS that it's not a good idea, especially with the Bragg-Brentano geometry, but I'm not sure... - Is there any difference, to be expected, in the calculated unit cell parameters, between approximating a phase using (1) a chemically pure STRUCTURE model (i.e., taken directly from a database / CIF file), and (2) a STRUCTURE with Occupancy information updated with the known content of impurities? I always thought that it shouldn't matter, as the software simply "iteratively"
Re: some uncertainties regarding the Rietveld refinement results in TOPAS
Great questions! > does the Absorption correction also influences microabsorption? No. The absorption correction in TOPAS is only a peak-shape correction to account for the transparency of the specimen. There is no real way to correct for microabsorption beyond grinding to a fine particle size, changing your wavelength, or moving to neutrons. > Is it worthy to use the parameter known as "Sample Thickness", associated with the Absorption Correction in TOPAS? Generally, no. You only need to worry about it if the penetration depth of your incident beam is larger than the thickness of the specimen. If that is the case, then "Sample Thickness" and "Scale Intensity" is necessary to correct for the change in peak shape due to the finite depth, and the change in diffracting volume. >Is there any difference, to be expected, in the calculated unit cell parameters, between approximating a phase using (1) a chemically pure STRUCTURE model, and (2) a STRUCTURE with Occupancy information updated with the known content of impurities? Altering site occupancies should only affect peak intensities, and so unit cell parameters should only be a second-order effect. (disregarding the physicality of various unit cell contents vs dimension pairings) > Is the negative intensity in the difference curve... It is showing that there is some mismatch between your model and the data. The intensities could be off because the peaks positions are wrong, the peak-shape is wrong, the atomic displacement parameters are wrong (don't refine these from normal QPA data), the LP factor correction is wrong, the site occupancies are wrong, the atomic scattering factors are wrong. Could also be that you have forgotten to refine a parameter that you should be. > We have been working using few computers with, supposedly the very same version of the software That does seem concerning. There can be differences between different versions of TOPAS. You would also need to verify that all the scattering factors, wavelengths, MACs, and space group files are the same for all installations, and that nothing has been incorrectly edited. Also that TOPAS.inc hasn't been touched. Hope this helps. Matthew On Wed, 6 Nov 2024 at 19:12, Łukasz Kruszewski wrote: > Dear Rietvelders, > > after some years of being a happy user of TOPAS & Rietveld method (after > working-out a good approach, thanks to a great help from users of this > mailing list for which I am really thanful, and having it checked by > attendance in Reynolds Cup) I was recently confronted with few questions > I couldn't answer. I thus have a kind request of a help here: > > - does the Absorption correction also influences microabsorption? This > question arose due to somewhat lowered expected wt.% of pyrite (19 > instead of 25), FeS2, even though all the used parameters (Absorption, > CrySize, Strain - for all the sample components) were checked and proven > to give value > error, i.e., being physically meaningful for the > refinement model. I was wondering how this supposed non-precision for > pyrite could be addressed... > > - Is it worthy to use the parameter known as "Sample Thickness", > associated with the Absorption Correction in TOPAS? I was taught in a > 5-day course of TOPAS that it's not a good idea, especially with the > Bragg-Brentano geometry, but I'm not sure... > > - Is there any difference, to be expected, in the calculated unit cell > parameters, between approximating a phase using (1) a chemically pure > STRUCTURE model (i.e., taken directly from a database / CIF file), and > (2) a STRUCTURE with Occupancy information updated with the known > content of impurities? I always thought that it shouldn't matter, as the > software simply "iteratively" compares the sample' reflections with the > structural information contained in STRUCTURE models (?). I have > compared these two situations, and option "2" actually gave me worse > fit. I thus wondered: does it really matter that much which model > (approximation) of the structure - chemically pure or with added > impurity info - is used? For most substances, the reflection positions > vary even for different standards found in databases like PDF or COD... > > - Is the negative intensity in the difference curve only informing us of > underestimation of the content of a particular phase, suggesting to > replace the Structure model used with another one? > > We have been working using few computers with, supposedly the very same > version of the software. However, even though we've been using exactly > the same Structure models, results were varying depending on the > computer stand used. I was wondering if this could have to do with some > errors in the JAVA script, or maybe a non-identical installation > conditions, or something about the Windows system? > > I would be thankful for any tips. > > Best regards, > > -- > Łukasz Kruszewski, Ph.D., Associate Professor > Polish Academy of Sciences > Institute of