Re: [time-nuts] +/- TI button on 5370B

2013-06-12 Thread Said Jackson
Jim,

Really annoying feature of HP counters.

If you slowly drift from a positive period to a negative one, it will indicate 
negative numbers for a while.

Then almost sudden it will do the jump to 0.9 seconds.

I found that adding a phase delay (long cable) helps keep the numbers positive.

Or trigger one on the rising, and one on the falling edge to add 50ns delay ( 
for signals with good 50% duty cycle).

Bye,
Said

On Jun 11, 2013, at 21:13, Jim Palfreyman jim77...@gmail.com wrote:

 Folks,
 
 I'm struggling to understand this button and how it reports intervals. It's
 supposed to show negative when the Stop is before the Start.
 
 When I connect up two clocks sending out 1PPS and say the one connected to
 Stop is ahead then sometimes I'd get -123.45 ns (say) and sometimes it
 flips to 999.999... ms. It can't seem to make up its mind which to use.
 
 Why is this inconsistent?
 
 If I flip it to + TI and make sure the leading clock is on Start - all
 works well.
 
 Thoughts?
 
 Jim Palfreyman
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] +/- TI button on 5370B

2013-06-12 Thread Bob Camp
Hi

Another simple solution - 

If you are using a GPSDO as a PPS source, use the cable delay to offset the pps 
you are using as a reference  by a microsecond. That's worked on every GPSDO 
I've tried it on. 

No it really doesn't solve the problem, it just covers it up. Post processing 
the data is the only real way to straighten things out in the general case. 

Bob

On Jun 12, 2013, at 2:48 AM, Said Jackson saidj...@aol.com wrote:

 Jim,
 
 Really annoying feature of HP counters.
 
 If you slowly drift from a positive period to a negative one, it will 
 indicate negative numbers for a while.
 
 Then almost sudden it will do the jump to 0.9 seconds.
 
 I found that adding a phase delay (long cable) helps keep the numbers 
 positive.
 
 Or trigger one on the rising, and one on the falling edge to add 50ns delay ( 
 for signals with good 50% duty cycle).
 
 Bye,
 Said
 
 On Jun 11, 2013, at 21:13, Jim Palfreyman jim77...@gmail.com wrote:
 
 Folks,
 
 I'm struggling to understand this button and how it reports intervals. It's
 supposed to show negative when the Stop is before the Start.
 
 When I connect up two clocks sending out 1PPS and say the one connected to
 Stop is ahead then sometimes I'd get -123.45 ns (say) and sometimes it
 flips to 999.999... ms. It can't seem to make up its mind which to use.
 
 Why is this inconsistent?
 
 If I flip it to + TI and make sure the leading clock is on Start - all
 works well.
 
 Thoughts?
 
 Jim Palfreyman
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Phase noise measurement with a scope

2013-06-12 Thread Stefan Heinzmann

Hi all,

given that digital scopes have a multichannel ADC for acquisition, which 
is similar to what a cross-correlating phase noise measurement 
instrument has, it occurred to me that phase noise measurement might 
also be possible with a standard digital scope and some post-processing 
software. The scope usually will have only 8 bits of resolution, but it 
will have a rather high sampling rate. With oversampling math, one may 
be able to trade one for the other, at least if the scope's analog 
frontend is not too bad.


Has anyone investigated or tried this? Is it a silly idea to start with?

Cheers
Stefan Heinzmann

attachment: stefan_heinzmann.vcf___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] +/- TI button on 5370B

2013-06-12 Thread Azelio Boriani
Not only HP counters. I have never seen a TI counter that outputs
negative values. I use the cable delay or user delay feature of GPSes
to delay one PPS to the other so that the result is always positive. I
have seen that only oscilloscopes can handle negative time interval
values. Maybe that the Wavecrest counters can handle negative TI, I
don't know.

On Wed, Jun 12, 2013 at 8:48 AM, Said Jackson saidj...@aol.com wrote:
 Jim,

 Really annoying feature of HP counters.

 If you slowly drift from a positive period to a negative one, it will 
 indicate negative numbers for a while.

 Then almost sudden it will do the jump to 0.9 seconds.

 I found that adding a phase delay (long cable) helps keep the numbers 
 positive.

 Or trigger one on the rising, and one on the falling edge to add 50ns delay ( 
 for signals with good 50% duty cycle).

 Bye,
 Said

 On Jun 11, 2013, at 21:13, Jim Palfreyman jim77...@gmail.com wrote:

 Folks,

 I'm struggling to understand this button and how it reports intervals. It's
 supposed to show negative when the Stop is before the Start.

 When I connect up two clocks sending out 1PPS and say the one connected to
 Stop is ahead then sometimes I'd get -123.45 ns (say) and sometimes it
 flips to 999.999... ms. It can't seem to make up its mind which to use.

 Why is this inconsistent?

 If I flip it to + TI and make sure the leading clock is on Start - all
 works well.

 Thoughts?

 Jim Palfreyman
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Phase noise measurement with a scope

2013-06-12 Thread Marek Peca

Hello,

given that digital scopes have a multichannel ADC for acquisition, which is 
similar to what a cross-correlating phase noise measurement instrument has, 
it occurred to me that phase noise measurement might also be possible with a 
standard digital scope and some post-processing software. The scope usually 
will have only 8 bits of resolution, but it will have a rather high sampling 
rate. With oversampling math, one may be able to trade one for the other, at 
least if the scope's analog frontend is not too bad.


Has anyone investigated or tried this? Is it a silly idea to start with?


yes, did it last week. I think it may have a sense with 1Gsps scope with 
good quality guts (should check with LC584AL at work).


I have tried it with a very cheap one, Rigol 2-channel, originally 50MHz, 
reflashed to 100MHz. 2 signals, refmeasured, into Ch1, Ch2. Waveforms 
(2x500Msps) acquired, sinc() interpolated. Results: short-term single-shot 
jitter around 100ps RMS. Long-term was of no interest for my purpose now, 
so no observations here.


Therefore, it is almost of no use at all for higher precision needs.


Regards,
Marek
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Ovenaire OSC 49-61C

2013-06-12 Thread Mark C. Stephens
Anybody have service, schematics or experience for one of these OCXO?
They are commonly seen in early HP 8566A spectrum analysers.

It is out of adjustable range.
Fortunately I had a spare OSC 49-61C floating about and replaced this off 
frequency reference with it.
However the replacement tends to wander about a fair bit.

By removing the 4 bottom screws and the 2 top screws, applying a little heat 
with a heat gun around the bottom to melt the glue, the oven came apart with 
little protest.
I was able to carefully access the top 2 boards and there is no sign of 
component failure there so it maybe something inside the oven itself.

I have stopped here until I get further information.


-marki

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Phase noise measurement with a scope

2013-06-12 Thread Didier Juges
I would think that considering the amount of time it takes to get the data out 
of the scope (particularly on the cheap scopes) would be a major impediment to 
that method regardless of the cleanliness of whatever data you eventually get, 
since you will only be able to analyze a small fraction of the available data 
(small aperture window).

Regarding the quality of the data, scopes are not particularly low noise. As 
long as the noise is not objectionable on screen, it is good enough for a 
scope, probably not time-nut standard.

Didier


Marek Peca ma...@duch.cz wrote:

Hello,

 given that digital scopes have a multichannel ADC for acquisition,
which is 
 similar to what a cross-correlating phase noise measurement
instrument has, 
 it occurred to me that phase noise measurement might also be possible
with a 
 standard digital scope and some post-processing software. The scope
usually 
 will have only 8 bits of resolution, but it will have a rather high
sampling 
 rate. With oversampling math, one may be able to trade one for the
other, at 
 least if the scope's analog frontend is not too bad.

 Has anyone investigated or tried this? Is it a silly idea to start
with?

yes, did it last week. I think it may have a sense with 1Gsps scope
with 
good quality guts (should check with LC584AL at work).

I have tried it with a very cheap one, Rigol 2-channel, originally
50MHz, 
reflashed to 100MHz. 2 signals, refmeasured, into Ch1, Ch2. Waveforms 
(2x500Msps) acquired, sinc() interpolated. Results: short-term
single-shot 
jitter around 100ps RMS. Long-term was of no interest for my purpose
now, 
so no observations here.

Therefore, it is almost of no use at all for higher precision needs.


Regards,
Marek
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

-- 
Sent from my Motorola Droid Razr 4G LTE wireless tracker while I do other 
things.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Phase noise measurement with a scope

2013-06-12 Thread Stefan Heinzmann

Marek Peca wrote:

Hello,

given that digital scopes have a multichannel ADC for acquisition, 
which is similar to what a cross-correlating phase noise measurement 
instrument has, it occurred to me that phase noise measurement might 
also be possible with a standard digital scope and some 
post-processing software. The scope usually will have only 8 bits of 
resolution, but it will have a rather high sampling rate. With 
oversampling math, one may be able to trade one for the other, at 
least if the scope's analog frontend is not too bad.


Has anyone investigated or tried this? Is it a silly idea to start with?


yes, did it last week. I think it may have a sense with 1Gsps scope 
with good quality guts (should check with LC584AL at work).


I have tried it with a very cheap one, Rigol 2-channel, originally 
50MHz, reflashed to 100MHz. 2 signals, refmeasured, into Ch1, Ch2. 
Waveforms (2x500Msps) acquired, sinc() interpolated. Results: 
short-term single-shot jitter around 100ps RMS. Long-term was of no 
interest for my purpose now, so no observations here.


Therefore, it is almost of no use at all for higher precision needs.


I was thinking about using a 4-channel scope with cross-spectrum 
averaging. Look at the Timepod by John Miles for an example of the 
method. I'm trying to guesstimate if the RS RTO scope, perhaps with the 
aid of the I/Q option, is capable of doing such measurements, and with 
what kind of performance.


Cheers
Stefan

attachment: stefan_heinzmann.vcf___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Ovenaire OSC 49-61C

2013-06-12 Thread Frederick Bray

Hi Mark,

I am not sure about the design of your OCXO but recently I repaired an 
Ovenaire from a Cushman 5510 service monitor.


I found that there was a bad trimmer pot in the oven control circuit.  
This wasn't apparent until I started monitoring the current drawn by the 
unit while trying to adjust the trimmer pot. I found that if I barely 
touched the adjustment screw, the current would vary greatly.  The 
temperature swing of the crystal made it impossible to adjust the 
oscillator trimmer cap properly.  These problems disappeared after I 
replaced the trimmer pot.


To get the new pot to a reasonably close starting, I measured the 
setting of the old one after I removed it and duplicated that on the new 
one before installing it.  I had to make an extra hole in the outer case 
to access the pot adjustment.  Although there was a hole in the internal 
foam to access the pot, there was no corresponding one in the outer 
can.  I adjusted the pot for the minimal variation in oven current while 
monitoring the output frequency for minimal frequency swing.


Now, based on observations conducted over a few days, the OCXO barely 
wanders and I would guess is within what Cushman required at the time 
the 5510 was built.


Good luck with yours.

Fred

On 6/12/2013 7:48 AM, Mark C. Stephens wrote:

Anybody have service, schematics or experience for one of these OCXO?
They are commonly seen in early HP 8566A spectrum analysers.

It is out of adjustable range.
Fortunately I had a spare OSC 49-61C floating about and replaced this off 
frequency reference with it.
However the replacement tends to wander about a fair bit.

By removing the 4 bottom screws and the 2 top screws, applying a little heat 
with a heat gun around the bottom to melt the glue, the oven came apart with 
little protest.
I was able to carefully access the top 2 boards and there is no sign of 
component failure there so it maybe something inside the oven itself.

I have stopped here until I get further information.


-marki

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Ovenaire OSC 49-61C

2013-06-12 Thread Mark C. Stephens
Hi Fred, Also, I neglected to ask, you wouldn't have the OXCO pin outs would 
you?

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Frederick Bray
Sent: Thursday, 13 June 2013 2:01 AM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Ovenaire OSC 49-61C

Hi Mark,

I am not sure about the design of your OCXO but recently I repaired an Ovenaire 
from a Cushman 5510 service monitor.

I found that there was a bad trimmer pot in the oven control circuit.  
This wasn't apparent until I started monitoring the current drawn by the unit 
while trying to adjust the trimmer pot. I found that if I barely touched the 
adjustment screw, the current would vary greatly.  The temperature swing of the 
crystal made it impossible to adjust the oscillator trimmer cap properly.  
These problems disappeared after I replaced the trimmer pot.

To get the new pot to a reasonably close starting, I measured the setting of 
the old one after I removed it and duplicated that on the new one before 
installing it.  I had to make an extra hole in the outer case to access the pot 
adjustment.  Although there was a hole in the internal foam to access the pot, 
there was no corresponding one in the outer can.  I adjusted the pot for the 
minimal variation in oven current while monitoring the output frequency for 
minimal frequency swing.

Now, based on observations conducted over a few days, the OCXO barely wanders 
and I would guess is within what Cushman required at the time the 5510 was 
built.

Good luck with yours.

Fred

On 6/12/2013 7:48 AM, Mark C. Stephens wrote:
 Anybody have service, schematics or experience for one of these OCXO?
 They are commonly seen in early HP 8566A spectrum analysers.

 It is out of adjustable range.
 Fortunately I had a spare OSC 49-61C floating about and replaced this off 
 frequency reference with it.
 However the replacement tends to wander about a fair bit.

 By removing the 4 bottom screws and the 2 top screws, applying a little heat 
 with a heat gun around the bottom to melt the glue, the oven came apart with 
 little protest.
 I was able to carefully access the top 2 boards and there is no sign of 
 component failure there so it maybe something inside the oven itself.

 I have stopped here until I get further information.


 -marki

 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Phase noise measurement with a scope

2013-06-12 Thread Marek Peca
(..) 
I have tried it with a very cheap one, Rigol 2-channel, originally 50MHz, 
reflashed to 100MHz. 2 signals, refmeasured, into Ch1, Ch2. Waveforms 
(2x500Msps) acquired, sinc() interpolated. Results: short-term single-shot 
jitter around 100ps RMS. Long-term was of no interest for my purpose now, 
so no observations here.


Therefore, it is almost of no use at all for higher precision needs.


I was thinking about using a 4-channel scope with cross-spectrum averaging. 
Look at the Timepod by John Miles for an example of the method. I'm trying to 
guesstimate if the RS RTO scope, perhaps with the aid of the I/Q option, is 
capable of doing such measurements, and with what kind of performance.


My point was, that DSO is basically an ADC. Therefore, there is some 
amount of noise, nonlinearity and drift, limiting the jitter measurement. 
Do you think any method can dig more information from given data than 
sinc() interpolation and zero-crossing computation?


Regards,
Marek
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Unit tests for time calculations

2013-06-12 Thread Doug Calvert
On Sun, Jun 9, 2013 at 11:29 PM, Bill Hawkins b...@iaxs.net wrote:

 Hi, Thanks in advance.

 Since this is a list for precise things, could you make your questions
 more precise?

   What sort of test cases?

   What sort of calculations?  Do you mean conversions?

   What do you mean by catching an error - where would you catch it?

   What do T1 and T2 have to do with it?

   What do you mean by timescale? Is a timescale a neighborhood?

 The work involved in clarifying those questions may make the answers
 clear to you.

 Best regards,


Are you familiar with the concept of a unit test?

I can clariy those questions but I am not sure how that is going to make
the answer clear to me. I am writing some software that reads input
from users and a number of disparate systems and I have no control over
the incoming time formats.  I would like to make sure that my time
calculatoins/conversions work and continue to work after i make changes
during development.


T1 and T2 are example time date strings.

Calculations:

T1(UTC) - T2(EST) = Time delta in seconds

12/31/1998 00:00:00 EST + 5184001 seconds = Time Date in UTC


Conversions:

01/01/1991 11:01 AM EST = XX/XX/ XX:XX UTC



As far as what do you mean by catching an error - where would you catch
it I cant tell if you are being snide. By catching an error I mean
identifying that the calculation my program made was incorrect. Maybe
you are unfamiliar with unit tests? What was unclear about what catch
errors? What else came to mind when you read the sentence.


Has anyone ever used time scale to describe a neighborhood?
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Phase noise measurement with a scope

2013-06-12 Thread Stefan Heinzmann

Marek Peca wrote:
(..) I have tried it with a very cheap one, Rigol 2-channel, 
originally 50MHz, reflashed to 100MHz. 2 signals, refmeasured, into 
Ch1, Ch2. Waveforms (2x500Msps) acquired, sinc() interpolated. 
Results: short-term single-shot jitter around 100ps RMS. Long-term 
was of no interest for my purpose now, so no observations here.


Therefore, it is almost of no use at all for higher precision needs.


I was thinking about using a 4-channel scope with cross-spectrum 
averaging. Look at the Timepod by John Miles for an example of the 
method. I'm trying to guesstimate if the RS RTO scope, perhaps with 
the aid of the I/Q option, is capable of doing such measurements, and 
with what kind of performance.


My point was, that DSO is basically an ADC. Therefore, there is some 
amount of noise, nonlinearity and drift, limiting the jitter 
measurement. Do you think any method can dig more information from 
given data than sinc() interpolation and zero-crossing computation?
The cross-spectrum averaging does indeed do just that, relying on two 
ADCs to produce uncorrelated noise, which can be averaged out.


Or am I misunderstanding your point?

Cheers
Stefan

attachment: stefan_heinzmann.vcf___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

[time-nuts] Spectracom 8170 Time of Day grief...

2013-06-12 Thread Burt I. Weiner
My good old Spectracom 8170 is not setting time.  I don't use it for 
frequency, just as a clock for my Hazetorium.  I live about 20 
minutes north of downtown Los Angles in Glendale.  The antenna I'm 
using is the Ferrite Rod loop in PVC that came with the 8170.  It's 
located on my back porch just laying on the floor with its maximum 
pickup direction towards Boulder, Colorado.  It has worked reliably 
for many years in that location.


The symptoms are...  One day about a month ago I noticed that it 
seemed to be in the start mode, that is, the far left digit was 
flashing between 0, 1, 2, and a lot of 4's, and the left most digits 
were counting time since an apparent restart.  Zeros and ones are the 
logical data states, two is a place holder and four indicates data 
errors.  I'm not aware of it having lost primary power, but it's not 
on a UPS.  In a week's time it did not give me time sync, but shows a 
green locked light most of the time.  This happened several years ago 
and a fellow at Spectracom told me to change all the electrolytics on 
the one board that has only 4 electrolytics on it.  I did and it 
started to work just peachy-keen.   The caps I put in then were of 
good quality, but I decided to change them again.  This did not solve 
the problem.


The power supply voltages are all well within limits and they look 
clean.  I seem to have plenty of, but not too much, signal.  By that 
I mean it doesn't appear to be oscillating.  The antenna is about 30' 
away from the receiver on a piece of coax that appears to look 
ok.  What I do notice is that the signal looks like it's going 
through quite a bit of turbulence.  I do not see anything that looks 
like interference.  I've looked at the output of the filter amp 
inside the 8170 and what I see is the signal dithering in amplitude 
quite rapidly and sometimes squaring off for a moment - sort of what 
you would expect to see during the Diurnal Shift periods of the 
day.  I can see the 10 dB drops, but they're not very clean due to 
the rapid dithering.  I've checked the antenna connections and tried 
different azimuth headings.  The location and azimuth it has been for 
years still seems to be the best.  Looking at the output of what I'm 
guessing is a Schmitt Trigger, I see the ST's output jumping from 0 
to +5, but erratically, which considering the received signal, makes sense.


I'm wondering if I'm just in a period of time that's receiving a poor 
quality signal?  It doesn't seem like WWVB's new phase-modulated 
format should be causing this kind of a problem.


All observations, opinions and suggestions are welcome.

Thanks,

Burt, K6OQK



Burt I. Weiner Associates
Broadcast Technical Services
Glendale, California  U.S.A.
b...@att.net
www.biwa.cc
K6OQK 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Phase noise measurement with a scope

2013-06-12 Thread Marek Peca
My point was, that DSO is basically an ADC. Therefore, there is some amount 
of noise, nonlinearity and drift, limiting the jitter measurement. Do you 
think any method can dig more information from given data than sinc() 
interpolation and zero-crossing computation?


The cross-spectrum averaging does indeed do just that, relying on two 
ADCs to produce uncorrelated noise, which can be averaged out.


Or am I misunderstanding your point?


Nothing against that. It depends on what noise level after averaging you 
require. I only posted my experience with a very low-quality DSO, which 
has 100psRMS single-shot. Using sinc() interpolation, but my point was, 
that I suppose there is no way to obtain better single-shot performance 
than this. To average out 100psRMS to, say, 1psRMS, it would require 10^4 
edges (under the assumption, that the 100psRMS is well behaved noise).


What performance it could yield with a better scope? I hope I'll try 
LC584AL some day, I guess it might give sth like 10psRMS single-shot...



Regards,
Marek
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Spectracom 8170 Time of Day grief...

2013-06-12 Thread paul swed
Burt
There is nothing wrong with your 8170. WWVB no longer allows it to work
correctly because of the phase modulation. They went to all psk about 1
month ago. They had been reverting back twice a day for things like the
8170.
So the ole 8170 is dead.
You need to build something like the d-psk-r to get it going though I have
not had time to actually add the am phase flipper precisely for the likes
of a 8170. I have one also.
Sorry
Paul
WB8TSL








On Wed, Jun 12, 2013 at 5:35 PM, Burt I. Weiner b...@att.net wrote:

 My good old Spectracom 8170 is not setting time.  I don't use it for
 frequency, just as a clock for my Hazetorium.  I live about 20 minutes
 north of downtown Los Angles in Glendale.  The antenna I'm using is the
 Ferrite Rod loop in PVC that came with the 8170.  It's located on my back
 porch just laying on the floor with its maximum pickup direction towards
 Boulder, Colorado.  It has worked reliably for many years in that location.

 The symptoms are...  One day about a month ago I noticed that it seemed to
 be in the start mode, that is, the far left digit was flashing between 0,
 1, 2, and a lot of 4's, and the left most digits were counting time since
 an apparent restart.  Zeros and ones are the logical data states, two is a
 place holder and four indicates data errors.  I'm not aware of it having
 lost primary power, but it's not on a UPS.  In a week's time it did not
 give me time sync, but shows a green locked light most of the time.  This
 happened several years ago and a fellow at Spectracom told me to change all
 the electrolytics on the one board that has only 4 electrolytics on it.  I
 did and it started to work just peachy-keen.   The caps I put in then were
 of good quality, but I decided to change them again.  This did not solve
 the problem.

 The power supply voltages are all well within limits and they look clean.
  I seem to have plenty of, but not too much, signal.  By that I mean it
 doesn't appear to be oscillating.  The antenna is about 30' away from the
 receiver on a piece of coax that appears to look ok.  What I do notice is
 that the signal looks like it's going through quite a bit of turbulence.  I
 do not see anything that looks like interference.  I've looked at the
 output of the filter amp inside the 8170 and what I see is the signal
 dithering in amplitude quite rapidly and sometimes squaring off for a
 moment - sort of what you would expect to see during the Diurnal Shift
 periods of the day.  I can see the 10 dB drops, but they're not very clean
 due to the rapid dithering.  I've checked the antenna connections and tried
 different azimuth headings.  The location and azimuth it has been for years
 still seems to be the best.  Looking at the output of what I'm guessing is
 a Schmitt Trigger, I see the ST's output jumping from 0 to +5, but
 erratically, which considering the received signal, makes sense.

 I'm wondering if I'm just in a period of time that's receiving a poor
 quality signal?  It doesn't seem like WWVB's new phase-modulated format
 should be causing this kind of a problem.

 All observations, opinions and suggestions are welcome.

 Thanks,

 Burt, K6OQK



 Burt I. Weiner Associates
 Broadcast Technical Services
 Glendale, California  U.S.A.
 b...@att.net
 www.biwa.cc
 K6OQK
 __**_
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/**
 mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Spectracom 8170 Time of Day grief...

2013-06-12 Thread paul swed
Burt
One other tidbit the phase mod will exactly cause what you see.
Spectracom used phase tracking to demodulate the AM time signal. Thats why
its nuts.


On Wed, Jun 12, 2013 at 5:49 PM, paul swed paulsw...@gmail.com wrote:

 Burt
 There is nothing wrong with your 8170. WWVB no longer allows it to work
 correctly because of the phase modulation. They went to all psk about 1
 month ago. They had been reverting back twice a day for things like the
 8170.
 So the ole 8170 is dead.
 You need to build something like the d-psk-r to get it going though I have
 not had time to actually add the am phase flipper precisely for the likes
 of a 8170. I have one also.
 Sorry
 Paul
 WB8TSL








 On Wed, Jun 12, 2013 at 5:35 PM, Burt I. Weiner b...@att.net wrote:

 My good old Spectracom 8170 is not setting time.  I don't use it for
 frequency, just as a clock for my Hazetorium.  I live about 20 minutes
 north of downtown Los Angles in Glendale.  The antenna I'm using is the
 Ferrite Rod loop in PVC that came with the 8170.  It's located on my back
 porch just laying on the floor with its maximum pickup direction towards
 Boulder, Colorado.  It has worked reliably for many years in that location.

 The symptoms are...  One day about a month ago I noticed that it seemed
 to be in the start mode, that is, the far left digit was flashing between
 0, 1, 2, and a lot of 4's, and the left most digits were counting time
 since an apparent restart.  Zeros and ones are the logical data states, two
 is a place holder and four indicates data errors.  I'm not aware of it
 having lost primary power, but it's not on a UPS.  In a week's time it did
 not give me time sync, but shows a green locked light most of the time.
  This happened several years ago and a fellow at Spectracom told me to
 change all the electrolytics on the one board that has only 4 electrolytics
 on it.  I did and it started to work just peachy-keen.   The caps I put in
 then were of good quality, but I decided to change them again.  This did
 not solve the problem.

 The power supply voltages are all well within limits and they look clean.
  I seem to have plenty of, but not too much, signal.  By that I mean it
 doesn't appear to be oscillating.  The antenna is about 30' away from the
 receiver on a piece of coax that appears to look ok.  What I do notice is
 that the signal looks like it's going through quite a bit of turbulence.  I
 do not see anything that looks like interference.  I've looked at the
 output of the filter amp inside the 8170 and what I see is the signal
 dithering in amplitude quite rapidly and sometimes squaring off for a
 moment - sort of what you would expect to see during the Diurnal Shift
 periods of the day.  I can see the 10 dB drops, but they're not very clean
 due to the rapid dithering.  I've checked the antenna connections and tried
 different azimuth headings.  The location and azimuth it has been for years
 still seems to be the best.  Looking at the output of what I'm guessing is
 a Schmitt Trigger, I see the ST's output jumping from 0 to +5, but
 erratically, which considering the received signal, makes sense.

 I'm wondering if I'm just in a period of time that's receiving a poor
 quality signal?  It doesn't seem like WWVB's new phase-modulated format
 should be causing this kind of a problem.

 All observations, opinions and suggestions are welcome.

 Thanks,

 Burt, K6OQK



 Burt I. Weiner Associates
 Broadcast Technical Services
 Glendale, California  U.S.A.
 b...@att.net
 www.biwa.cc
 K6OQK
 __**_
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/**
 mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Update on the d-psk-r schematic costas loop

2013-06-12 Thread paul swed
Good catch by Russ of time-nuttery fame. Wrong ground on the 74hc390.
Updated schematic attached.
Thanks Rudd
Regards
Paul
WB8TSL


WWVB dpskr Costas_Div nu 12Mhz 06122013.sch
Description: Binary data
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Unit tests for time calculations

2013-06-12 Thread Tom Van Baak
 12/31/1998 00:00:00 EST + 5184001 seconds = Time Date in UTC

 01/01/1991 11:01 AM EST = XX/XX/ XX:XX UTC

Doug,

This is difficult (or impossible) to do right; at a minimum you need a table of 
all past, current, and future national timezone definitions, DST rules by 
locale, and leap seconds. Start by defining your problem very explicitly, down 
to each character, digit, integer, and allowable range.

I think there is a timezone mailing list that would better be able to answer 
your question. They may point you to a number of well tested libraries that 
already do this sort of thing.

/tvb

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] New to list and GPSDO questions

2013-06-12 Thread Bob Stewart
Hello to the list.  I saw on K3PGP's site a mention that the UT-41 GPS receiver 
had a 10KHz signal on-board so I decided, why not build a GPSDO for my new HP 
5334B?  Unfortunately the one I bought doesn't have a 10KHz point, and the 
board doesn't even pull out the 1PPS signal from the chip.  So I've had to 
switch gears and will go with VE2ZAZ's board.  I've got a Motorola UT+ on 
order, but I was wondering about using the board's LED output as a sort of 
low-rent source of 1PPS.  Would the short term accuracy be too bad to even 
bother with?  Has anyone done any tests to see if the LED and 1PPS signals are 
essentially the same signal on one of these cheapo boards, or at least to find 
out how often the LED output is corrected?  The one I got has a Prolific 
PL-6313 chip.  

Bob - AE6RV
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Unit tests for time calculations

2013-06-12 Thread Chris Albertson
I think it depends on you definition of Unit Test.  Some people do a
sanity check time test just to verify the function works.  Those who want
a better test will use a code coverage tool and will add test cases untill
every path through the function is exercised.


One thing to remember is NO TEST can prove a function correct, you can only
prove it wrong.  No matter how many tests you do yo only gain confedance
that the function is likely not bad.

A famous proof goes like this:  Can I divide 60 by 1?  Yes of course, by
2? yes, by 3, by 4, by 5 , by 6 all yes but this is going to slow so let's
try 10, 20 ,30.  Ok end of proof: 60 is divisible by all integers.   Sady,
software unit tests can only be as good as this proof.  So just use enough
tet cases to exercise every code path and then you will know there is
nothing stupid wrong with it.


On Wed, Jun 12, 2013 at 8:02 PM, Tom Van Baak t...@leapsecond.com wrote:

  12/31/1998 00:00:00 EST + 5184001 seconds = Time Date in UTC

  01/01/1991 11:01 AM EST = XX/XX/ XX:XX UTC

 Doug,

 This is difficult (or impossible) to do right; at a minimum you need a
 table of all past, current, and future national timezone definitions, DST
 rules by locale, and leap seconds. Start by defining your problem very
 explicitly, down to each character, digit, integer, and allowable range.

 I think there is a timezone mailing list that would better be able to
 answer your question. They may point you to a number of well tested
 libraries that already do this sort of thing.

 /tvb

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




-- 

Chris Albertson
Redondo Beach, California
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Spectracom 8170 Time of Day grief...

2013-06-12 Thread Burt I. Weiner

Paul,

U... What a revolting development this is.  My 9150-52054 have a 
smaller readout out that I can't see without climbing up on my bench, 
so I suppose I will no longer know what time it is.


I've never really had to delve into the 8170 so I wasn't up to speed 
on how they derived the data from the carrier prior to the new PSK 
format.  I was under the misunderstanding that it was strictly from 
amplitude variation.  I guess it is, except for the manner in which 
they determine the changes.


Please tell me about the d-psk-r you mentioned.  Also do you know 
the times of day WWVB reverts to the old method and for how 
long?  I'll have to go to their site and see.


Burt, K6OQK

At 04:46 PM 6/12/2013, time-nuts-requ...@febo.com wrote

Burt
One other tidbit the phase mod will exactly cause what you see.
Spectracom used phase tracking to demodulate the AM time signal. Thats why
its nuts.


On Wed, Jun 12, 2013 at 5:49 PM, paul swed paulsw...@gmail.com wrote:

 Burt
 There is nothing wrong with your 8170. WWVB no longer allows it to work
 correctly because of the phase modulation. They went to all psk about 1
 month ago. They had been reverting back twice a day for things like the
 8170.
 So the ole 8170 is dead.
 You need to build something like the d-psk-r to get it going though I have
 not had time to actually add the am phase flipper precisely for the likes
 of a 8170. I have one also.
 Sorry
 Paul
 WB8TSL

 On Wed, Jun 12, 2013 at 5:35 PM, Burt I. Weiner b...@att.net wrote:

 My good old Spectracom 8170 is not setting time.  I don't use it for
 frequency, just as a clock for my Hazetorium.  I live about 20 minutes
 north of downtown Los Angles in Glendale.  The antenna I'm using is the
 Ferrite Rod loop in PVC that came with the 8170.  It's located on my back
 porch just laying on the floor with its maximum pickup direction towards
 Boulder, Colorado.  It has worked reliably for many years in 
that location.


 The symptoms are...


Burt I. Weiner Associates
Broadcast Technical Services
Glendale, California  U.S.A.
b...@att.net
www.biwa.cc
K6OQK  


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Phase noise measurement with a scope

2013-06-12 Thread Chris Albertson
My dim memory says there is some analog way to multiply the phase noise.
 What does that?   Then it might be easier to measure.


On Wed, Jun 12, 2013 at 2:36 PM, Marek Peca ma...@duch.cz wrote:

 My point was, that DSO is basically an ADC. Therefore, there is some
 amount of noise, nonlinearity and drift, limiting the jitter measurement.
 Do you think any method can dig more information from given data than
 sinc() interpolation and zero-crossing computation?


  The cross-spectrum averaging does indeed do just that, relying on two
 ADCs to produce uncorrelated noise, which can be averaged out.

 Or am I misunderstanding your point?


 Nothing against that. It depends on what noise level after averaging you
 require. I only posted my experience with a very low-quality DSO, which has
 100psRMS single-shot. Using sinc() interpolation, but my point was, that I
 suppose there is no way to obtain better single-shot performance than this.
 To average out 100psRMS to, say, 1psRMS, it would require 10^4 edges (under
 the assumption, that the 100psRMS is well behaved noise).

 What performance it could yield with a better scope? I hope I'll try
 LC584AL some day, I guess it might give sth like 10psRMS single-shot...


 Regards,
 Marek
 __**_
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/**
 mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




-- 

Chris Albertson
Redondo Beach, California
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts