[time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread Frederick Bray
I have what would probably be considered to be a newbie question, so 
please excuse my likely ignorance.


I have a HP 5385A frequency counter, which I am using with an external 
standard.  If I try to compare one 10 MHz standard against another, 
using one as the external standard and the other as the device being 
measured, the readout fluctuates between 10.000 and 9.999.  This 
happens when I am comparing a Z3801A and the Trimble look-a-like using 
the same GPS antenna and identical cables (including length) between the 
GPS antenna splitter and the two units.  I have also used identical 
cables between the 10 MHz outputs and the counters.   Hence, cable 
variations shouldn't be the issue.


The same thing happens comparing the Timesource 2700 against either of 
the GPS standards.


Is this attributable to the different standards or the frequency 
counter?  I haven't yet set up a dual trace scope to compare the various 
sources directly, but that will be my next step.


Thanks for any comments, etc.

Fred Bray
W6WAW

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread Robin Kimberley
Any frequency counter will have an given accuracy  +/- the LSD (least
significant digit), so alternating between  10.000 and 9.999 looks
perfectly normal to me.

Rob Kimberley

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Frederick Bray
Sent: 24 November 2011 17:56
To: time-nuts@febo.com
Subject: [time-nuts] Differences / Jitter Between Frequency Standard Sources

I have what would probably be considered to be a newbie question, so
please excuse my likely ignorance.

I have a HP 5385A frequency counter, which I am using with an external
standard.  If I try to compare one 10 MHz standard against another, using
one as the external standard and the other as the device being measured, the
readout fluctuates between 10.000 and 9.999.  This happens when I am
comparing a Z3801A and the Trimble look-a-like using the same GPS antenna
and identical cables (including length) between the GPS antenna splitter and
the two units.  I have also used identical 
cables between the 10 MHz outputs and the counters.   Hence, cable 
variations shouldn't be the issue.

The same thing happens comparing the Timesource 2700 against either of the
GPS standards.

Is this attributable to the different standards or the frequency counter?  I
haven't yet set up a dual trace scope to compare the various sources
directly, but that will be my next step.

Thanks for any comments, etc.

Fred Bray
W6WAW

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread Randy D. Hunt

On 11/24/2011 10:01 AM, Robin Kimberley wrote:

Any frequency counter will have an given accuracy  +/- the LSD (least
significant digit), so alternating between  10.000 and 9.999 looks
perfectly normal to me.

Rob Kimberley

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Frederick Bray
Sent: 24 November 2011 17:56
To: time-nuts@febo.com
Subject: [time-nuts] Differences / Jitter Between Frequency Standard Sources

I have what would probably be considered to be a newbie question, so
please excuse my likely ignorance.

I have a HP 5385A frequency counter, which I am using with an external
standard.  If I try to compare one 10 MHz standard against another, using
one as the external standard and the other as the device being measured, the
readout fluctuates between 10.000 and 9.999.  This happens when I am
comparing a Z3801A and the Trimble look-a-like using the same GPS antenna
and identical cables (including length) between the GPS antenna splitter and
the two units.  I have also used identical
cables between the 10 MHz outputs and the counters.   Hence, cable
variations shouldn't be the issue.

The same thing happens comparing the Timesource 2700 against either of the
GPS standards.

Is this attributable to the different standards or the frequency counter?  I
haven't yet set up a dual trace scope to compare the various sources
directly, but that will be my next step.

Thanks for any comments, etc.

Fred Bray
W6WAW

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

actually, any digital display system would be +-1 count in the least 
significant position. counters, voltmeters, etc.


randy
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread David Bobbett
Just as an aside, you /can /get rid of the 'dithering digit' problem if 
you use a counter which employs a reciprocal counting technique.


I use a Thandar Thirlby TF830 locked to a 10MHz reference derived from a 
Tbolt, but I'm sure there must be others.



David Bobbett, G4IRQ



On 24/11/2011 18:01, Robin Kimberley wrote:

Any frequency counter will have an given accuracy  +/- the LSD (least
significant digit), so alternating between  10.000 and 9.999 looks
perfectly normal to me.

Rob Kimberley

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Frederick Bray
Sent: 24 November 2011 17:56
To: time-nuts@febo.com
Subject: [time-nuts] Differences / Jitter Between Frequency Standard Sources

I have what would probably be considered to be a newbie question, so
please excuse my likely ignorance.

I have a HP 5385A frequency counter, which I am using with an external
standard.  If I try to compare one 10 MHz standard against another, using
one as the external standard and the other as the device being measured, the
readout fluctuates between 10.000 and 9.999.  This happens when I am
comparing a Z3801A and the Trimble look-a-like using the same GPS antenna
and identical cables (including length) between the GPS antenna splitter and
the two units.  I have also used identical
cables between the 10 MHz outputs and the counters.   Hence, cable
variations shouldn't be the issue.

The same thing happens comparing the Timesource 2700 against either of the
GPS standards.

Is this attributable to the different standards or the frequency counter?  I
haven't yet set up a dual trace scope to compare the various sources
directly, but that will be my next step.

Thanks for any comments, etc.

Fred Bray
W6WAW

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread Azelio Boriani
Or modern counters where internally every type of measurement is reduced to
a time interval measurement like the HP53132 (150pS resolution).

On Thu, Nov 24, 2011 at 8:04 PM, David Bobbett d.bobb...@tiscali.co.ukwrote:

 Just as an aside, you /can /get rid of the 'dithering digit' problem if
 you use a counter which employs a reciprocal counting technique.

 I use a Thandar Thirlby TF830 locked to a 10MHz reference derived from a
 Tbolt, but I'm sure there must be others.


 David Bobbett, G4IRQ




 On 24/11/2011 18:01, Robin Kimberley wrote:

 Any frequency counter will have an given accuracy  +/- the LSD (least
 significant digit), so alternating between  10.000 and 9.999 looks
 perfectly normal to me.

 Rob Kimberley

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Frederick Bray
 Sent: 24 November 2011 17:56
 To: time-nuts@febo.com
 Subject: [time-nuts] Differences / Jitter Between Frequency Standard
 Sources

 I have what would probably be considered to be a newbie question, so
 please excuse my likely ignorance.

 I have a HP 5385A frequency counter, which I am using with an external
 standard.  If I try to compare one 10 MHz standard against another, using
 one as the external standard and the other as the device being measured,
 the
 readout fluctuates between 10.000 and 9.999.  This happens when I
 am
 comparing a Z3801A and the Trimble look-a-like using the same GPS antenna
 and identical cables (including length) between the GPS antenna splitter
 and
 the two units.  I have also used identical
 cables between the 10 MHz outputs and the counters.   Hence, cable
 variations shouldn't be the issue.

 The same thing happens comparing the Timesource 2700 against either of the
 GPS standards.

 Is this attributable to the different standards or the frequency counter?
  I
 haven't yet set up a dual trace scope to compare the various sources
 directly, but that will be my next step.

 Thanks for any comments, etc.

 Fred Bray
 W6WAW

 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.





 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

  ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread Frederick Bray
Thanks to all who replied.  I thought that it was a counter artifact, 
but I didn't know why.


Any suggestions for a reasonably priced good frequency counter that uses 
the reciprocal method?  I saw there were a couple TF830's on ebay, but 
they were outside the U.S.


The HP 5340A looks like it might do the job, but I have heard that there 
are repair issues if it fails.


Thanks.

Fred Bray
W6WAW

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread cfo
On Thu, 24 Nov 2011 12:54:12 -0800, Frederick Bray wrote:

 Thanks to all who replied.  I thought that it was a counter artifact,
 but I didn't know why.
 
 Any suggestions for a reasonably priced good frequency counter that uses
 the reciprocal method?  I saw there were a couple TF830's on ebay, but
 they were outside the U.S.
 
 The HP 5340A looks like it might do the job, but I have heard that there
 are repair issues if it fails.
 
 Thanks.
 
 Fred Bray
 W6WAW

Isn't your 5385A counter a reciprocal type ?

Quote The Agilent 5385A frequency counter utilizes interpolar-enhanced 
reciprocal counting

http://www.home.agilent.com/agilent/product.jspx?
nid=-536902484.536879843.00lc=engcc=US

CFO



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread Frederick Bray

Looks like my counter is reciprocal.

Thanks for the info.

Fred Bray
W6WAW

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Differences / Jitter Between Frequency Standard Sources

2011-11-24 Thread Tom Van Baak
I have a HP 5385A frequency counter, which I am using with an external 
standard.  If I try to compare one 10 MHz standard against another, 
using one as the external standard and the other as the device being 
measured, the readout fluctuates between 10.000 and 9.999.  This 
happens when I am comparing a Z3801A and the Trimble look-a-like using 
the same GPS antenna and identical cables (including length) between the 
GPS antenna splitter and the two units.  I have also used identical 
cables between the 10 MHz outputs and the counters.   Hence, cable 
variations shouldn't be the issue.


The same thing happens comparing the Timesource 2700 against either of 
the GPS standards.


Is this attributable to the different standards or the frequency 
counter?  I haven't yet set up a dual trace scope to compare the various 
sources directly, but that will be my next step.


Fred,

This is normal (for both gated and reciprocal counters). Never trust
a frequency counter that gives you the same number every time.

In fact it is very advantages. If you log hundreds or thousands of
these frequency readings and compute the mean you will get a
more precise value for your long-term average frequency. It's
unlikely you will get a 50/50 split between the two readings; the
actual ratio between them allows you to interpolate somewhat
between 9 and 0 in the last digit. Try it.

/tvb



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.