Re: [time-nuts] FE5680A Calibration

2012-03-15 Thread Bob Camp
Hi

Drive the GPS pps into the set input on a flip flop, drive the pps from the
FE into the reset input. Use the UC10 to measure the period of the waveform
on the Q output. Not super high resolution, but if you are patient, you can
get the job done.

Bob 

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of David
Sent: Wednesday, March 14, 2012 7:34 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] FE5680A Calibration

I picked up a gimpy Beckman UC10 universal counter not long ago for
about $10 from Ebay.  Even better, I just repaired a Tektronix 7D15
(it has a whole board full of those junk TI integrated circuit sockets
which need to be replaced) although you need to leave an entire
oscilloscope mainframe on to use it.  The advantage of the later is
adjustable slope, sensitivity, and triggering.

On Wed, 14 Mar 2012 17:50:28 -0400, Bob Camp li...@rtty.us wrote:

At least the don't mess to much with it part has sunk in. That puts you
ahead of most people at this point.

A usable counter should be a sub $100 sort of thing either at auction or
surplus. With some careful shopping it can be a sub $40 item.

On Mar 14, 2012, at 4:25 PM, Chris Stake st...@btinternet.com wrote:

 Hi Bob,
 Thanks for pointing-out the noisy output of the FE5680A. I'll probably
try
 to lock a crystal oscillator to it.
 Unfortunately I don't have a suitable counter.
 I was hoping I could use some sort of higher frequency standard but I
 confess I had not really grasped the fact that the unit may well be
within
 millihertz of the nominal frequency: too delicate to twiddle with
anything
 other than precision equipment and long timebases.
 Kind regards
 Chris
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bob Camp
 Sent: 14 March 2012 16:57
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] FE5680A Calibration
 
 Hi
 
 Do you have a dual channel counter that you can put the GPS into on the
 start and the FE into as the stop? The HP 5334, 5335, and 5345 are
all
 examples of this sort of counter.
 
 Bob
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Chris Stake
 Sent: Wednesday, March 14, 2012 12:49 PM
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] FE5680A Calibration
 
 Nice idea,
 But I haven't got a DSO and the persistence of my scope isn't good
enough
 to
 view 10Mhz sampled at 1pps.
 Chris
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of EB4APL
 Sent: 14 March 2012 15:46
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] FE5680A Calibration
 
 My fast approach would be to trigger a scope with the 1 PPS from the
GPS
 receiver and observe the how the 10 MHz output of your Rb drifts.  1
 full cycle per second is 1 e-7 so you'll need to use an stopwatch to
 time long periods when you are fine adjusting .
 Building (or buying) a GPSDO allows yo to make the comparison between
 both 10 MHz outputs without the jitter in the GPS receiver 1 PPS.
 Probably others in this list can suggest more elaborated and convenient
 approaches to this.
 
 On 14/03/2012 16:17, Chris Stake wrote:
 Hello all,
 
 I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it
 to a
 16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to
 drive
 a LED for the locked signal. I can communicate with it using the
 excellent
 Fe5680Calibrator software. As received the frequency offset is set to
 Zero
 and the unit seems to work well.
 
 I plan to put it into service as a workshop frequency reference and so
 would
 like to set it close to 10.0Mhz. I have various oscillator modules,
 signal
 generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
 test-gear. I live in a river valley in North Devon UK so radio
 reception
 is
 a bit restricted but I receive digital television via a masthead
 amplifier
 and long downlead.
 
 Could someone please suggest a way of going about this?
 
 
 
 Regards
 
 Chris Stake
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
https

[time-nuts] FE5680A Calibration

2012-03-14 Thread Chris Stake
Hello all,

I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it to a
16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to drive
a LED for the locked signal. I can communicate with it using the excellent
Fe5680Calibrator software. As received the frequency offset is set to Zero
and the unit seems to work well.

I plan to put it into service as a workshop frequency reference and so would
like to set it close to 10.0Mhz. I have various oscillator modules, signal
generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
test-gear. I live in a river valley in North Devon UK so radio reception is
a bit restricted but I receive digital television via a masthead amplifier
and long downlead.

Could someone please suggest a way of going about this?

 

Regards

Chris Stake

 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread EB4APL
My fast approach would be to trigger a scope with the 1 PPS from the GPS 
receiver and observe the how the 10 MHz output of your Rb drifts.  1 
full cycle per second is 1 e-7 so you'll need to use an stopwatch to 
time long periods when you are fine adjusting .
Building (or buying) a GPSDO allows yo to make the comparison between 
both 10 MHz outputs without the jitter in the GPS receiver 1 PPS.
Probably others in this list can suggest more elaborated and convenient 
approaches to this.


On 14/03/2012 16:17, Chris Stake wrote:

Hello all,

I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it to a
16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to drive
a LED for the locked signal. I can communicate with it using the excellent
Fe5680Calibrator software. As received the frequency offset is set to Zero
and the unit seems to work well.

I plan to put it into service as a workshop frequency reference and so would
like to set it close to 10.0Mhz. I have various oscillator modules, signal
generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
test-gear. I live in a river valley in North Devon UK so radio reception is
a bit restricted but I receive digital television via a masthead amplifier
and long downlead.

Could someone please suggest a way of going about this?



Regards

Chris Stake



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread Pete Lancashire
The first thing to consider is something to put the 5680 into. Search
the list for
temperature considerations. Basically you will need to heat sink the 5680. It
can be run without, but won't last long. The technique I'm going to
use it to heat
sink the 5680 and with a pair of small variable speed fans (one as a backup) the
inside of the box will be keep at a reasonably constant temperature with the set
point based on keeping the 5680 in a safe region.

There's been quite a bit of discussion on 10 MHz distribution. For now
I'm thinking
of using a surplus commercial video distribution amplifier. Most
studio grade DA's
are good to 15 MHz +- 3db. I also have a HP 5087A but its input is 5 MHz and
the output cards are mostly 5MHz. A future project to convert.

I have not decided if I want to put the Rb which maybe a LPRO, the Thunderbolt,
and maybe a 10811 into one box or not. One idea is with them all on
one box having
the ability to adjust the Rb and the 10811 via the Thunderbolt, and
one box can be
looked at as a temp controlled 'oven' for all three.

One could also put a DIY DA into the same box.

-pete

PS A friend and time-nut has be threatening to do a modern easy to
build DA, many on
the list need one.

On Wed, Mar 14, 2012 at 8:17 AM, Chris Stake st...@btinternet.com wrote:
 Hello all,

 I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it to a
 16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to drive
 a LED for the locked signal. I can communicate with it using the excellent
 Fe5680Calibrator software. As received the frequency offset is set to Zero
 and the unit seems to work well.

 I plan to put it into service as a workshop frequency reference and so would
 like to set it close to 10.0Mhz. I have various oscillator modules, signal
 generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
 test-gear. I live in a river valley in North Devon UK so radio reception is
 a bit restricted but I receive digital television via a masthead amplifier
 and long downlead.

 Could someone please suggest a way of going about this?



 Regards

 Chris Stake



 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread Bob Camp
Hi

Consider that the noise on the 10 MHz output of the FE5680 is pretty bad.
You may / may not want to propagate it all over your shop. At the very
least, a simple band pass filter is needed. A better solution would be to
lock a crystal oscillator up to the FE and use the output of the oscillator.


It's a reasonable bet that you unit is within +/- 1x10^-10 of the right
frequency. If you are lucky, it's  10X better than that. To observe it's
frequency adequately to set it well, you will need to average it's output
for at least a couple hundred seconds. You don't want to make it worse
trying to set it closer

If you are comparing to a normal GPS receiver, they have noise on their PPS
output as well. To get to adequate stability / resolution you will need to
average them for a while. Just how long depends on which one you have and a
few other things. 

Best approach / lazy approach / provides adequate time for a beer: Look at
the edge of the FE's pps relative to the pps out of the GPS on a scope. Note
the offset on a piece of paper along with the time. Come back in an hour and
repeat the process. Get a couple of points this way. The drift (if any) will
show up as a linear trend. The noise will likely be in the 20 to 200 ns. At
this point all you really are looking for is a noise estimate.

Say the noise is 200 ns. If you do observations at a 200 second spacing you
would get 1x10^-9. To get to 1x10^-12 you need to observe for about 1000X
longer. 200,000 seconds is a couple of days. Yes indeed there are fancy math
things try to speed things up. They don't fit under the lazy approach
(there also are a few other reasons to take your time). 

Based on the noise estimate you get, and the calculations above, come up
with a schedule. No need to get obsessive about it. All you are looking for
is: read every day / every three days / once a week sort of spacing. Start
logging your phase offset at what ever spacing makes sense. Take at least
ten readings. 

Based on your readings, take a stab at correcting the FE. Adjust it and then
go back to taking readings. Eventually it will be close enough.

So much fun...

Bob

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Chris Stake
Sent: Wednesday, March 14, 2012 11:17 AM
To: 'Discussion of precise time and frequency measurement'
Subject: [time-nuts] FE5680A Calibration

Hello all,

I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it to a
16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to drive
a LED for the locked signal. I can communicate with it using the excellent
Fe5680Calibrator software. As received the frequency offset is set to Zero
and the unit seems to work well.

I plan to put it into service as a workshop frequency reference and so would
like to set it close to 10.0Mhz. I have various oscillator modules, signal
generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
test-gear. I live in a river valley in North Devon UK so radio reception is
a bit restricted but I receive digital television via a masthead amplifier
and long downlead.

Could someone please suggest a way of going about this?

 

Regards

Chris Stake

 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread Chris Stake
Nice idea,
But I haven't got a DSO and the persistence of my scope isn't good enough to
view 10Mhz sampled at 1pps.
Chris

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of EB4APL
 Sent: 14 March 2012 15:46
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] FE5680A Calibration
 
 My fast approach would be to trigger a scope with the 1 PPS from the GPS
 receiver and observe the how the 10 MHz output of your Rb drifts.  1
 full cycle per second is 1 e-7 so you'll need to use an stopwatch to
 time long periods when you are fine adjusting .
 Building (or buying) a GPSDO allows yo to make the comparison between
 both 10 MHz outputs without the jitter in the GPS receiver 1 PPS.
 Probably others in this list can suggest more elaborated and convenient
 approaches to this.
 
 On 14/03/2012 16:17, Chris Stake wrote:
  Hello all,
 
  I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it
 to a
  16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to
 drive
  a LED for the locked signal. I can communicate with it using the
 excellent
  Fe5680Calibrator software. As received the frequency offset is set to
 Zero
  and the unit seems to work well.
 
  I plan to put it into service as a workshop frequency reference and so
 would
  like to set it close to 10.0Mhz. I have various oscillator modules,
 signal
  generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
  test-gear. I live in a river valley in North Devon UK so radio reception
 is
  a bit restricted but I receive digital television via a masthead
 amplifier
  and long downlead.
 
  Could someone please suggest a way of going about this?
 
 
 
  Regards
 
  Chris Stake
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread Bob Camp
Hi

Do you have a dual channel counter that you can put the GPS into on the
start and the FE into as the stop? The HP 5334, 5335, and 5345 are all
examples of this sort of counter.

Bob

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Chris Stake
Sent: Wednesday, March 14, 2012 12:49 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] FE5680A Calibration

Nice idea,
But I haven't got a DSO and the persistence of my scope isn't good enough to
view 10Mhz sampled at 1pps.
Chris

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of EB4APL
 Sent: 14 March 2012 15:46
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] FE5680A Calibration
 
 My fast approach would be to trigger a scope with the 1 PPS from the GPS
 receiver and observe the how the 10 MHz output of your Rb drifts.  1
 full cycle per second is 1 e-7 so you'll need to use an stopwatch to
 time long periods when you are fine adjusting .
 Building (or buying) a GPSDO allows yo to make the comparison between
 both 10 MHz outputs without the jitter in the GPS receiver 1 PPS.
 Probably others in this list can suggest more elaborated and convenient
 approaches to this.
 
 On 14/03/2012 16:17, Chris Stake wrote:
  Hello all,
 
  I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it
 to a
  16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to
 drive
  a LED for the locked signal. I can communicate with it using the
 excellent
  Fe5680Calibrator software. As received the frequency offset is set to
 Zero
  and the unit seems to work well.
 
  I plan to put it into service as a workshop frequency reference and so
 would
  like to set it close to 10.0Mhz. I have various oscillator modules,
 signal
  generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
  test-gear. I live in a river valley in North Devon UK so radio reception
 is
  a bit restricted but I receive digital television via a masthead
 amplifier
  and long downlead.
 
  Could someone please suggest a way of going about this?
 
 
 
  Regards
 
  Chris Stake
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread Chris Stake
Hi Bob,
Thanks for pointing-out the noisy output of the FE5680A. I'll probably try
to lock a crystal oscillator to it.
Unfortunately I don't have a suitable counter.
I was hoping I could use some sort of higher frequency standard but I
confess I had not really grasped the fact that the unit may well be within
millihertz of the nominal frequency: too delicate to twiddle with anything
other than precision equipment and long timebases.
Kind regards
Chris

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bob Camp
 Sent: 14 March 2012 16:57
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] FE5680A Calibration
 
 Hi
 
 Do you have a dual channel counter that you can put the GPS into on the
 start and the FE into as the stop? The HP 5334, 5335, and 5345 are all
 examples of this sort of counter.
 
 Bob
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Chris Stake
 Sent: Wednesday, March 14, 2012 12:49 PM
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] FE5680A Calibration
 
 Nice idea,
 But I haven't got a DSO and the persistence of my scope isn't good enough
 to
 view 10Mhz sampled at 1pps.
 Chris
 
  -Original Message-
  From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
  Behalf Of EB4APL
  Sent: 14 March 2012 15:46
  To: time-nuts@febo.com
  Subject: Re: [time-nuts] FE5680A Calibration
 
  My fast approach would be to trigger a scope with the 1 PPS from the GPS
  receiver and observe the how the 10 MHz output of your Rb drifts.  1
  full cycle per second is 1 e-7 so you'll need to use an stopwatch to
  time long periods when you are fine adjusting .
  Building (or buying) a GPSDO allows yo to make the comparison between
  both 10 MHz outputs without the jitter in the GPS receiver 1 PPS.
  Probably others in this list can suggest more elaborated and convenient
  approaches to this.
 
  On 14/03/2012 16:17, Chris Stake wrote:
   Hello all,
  
   I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it
  to a
   16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to
  drive
   a LED for the locked signal. I can communicate with it using the
  excellent
   Fe5680Calibrator software. As received the frequency offset is set to
  Zero
   and the unit seems to work well.
  
   I plan to put it into service as a workshop frequency reference and so
  would
   like to set it close to 10.0Mhz. I have various oscillator modules,
  signal
   generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
   test-gear. I live in a river valley in North Devon UK so radio
 reception
  is
   a bit restricted but I receive digital television via a masthead
  amplifier
   and long downlead.
  
   Could someone please suggest a way of going about this?
  
  
  
   Regards
  
   Chris Stake
  
  
  
   ___
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to https://www.febo.com/cgi-
  bin/mailman/listinfo/time-nuts
   and follow the instructions there.
  
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-
  nuts
  and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread Bob Camp
Hi

At least the don't mess to much with it part has sunk in. That puts you ahead 
of most people at this point.

A usable counter should be a sub $100 sort of thing either at auction or 
surplus. With some careful shopping it can be a sub $40 item.

Bob



On Mar 14, 2012, at 4:25 PM, Chris Stake st...@btinternet.com wrote:

 Hi Bob,
 Thanks for pointing-out the noisy output of the FE5680A. I'll probably try
 to lock a crystal oscillator to it.
 Unfortunately I don't have a suitable counter.
 I was hoping I could use some sort of higher frequency standard but I
 confess I had not really grasped the fact that the unit may well be within
 millihertz of the nominal frequency: too delicate to twiddle with anything
 other than precision equipment and long timebases.
 Kind regards
 Chris
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bob Camp
 Sent: 14 March 2012 16:57
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] FE5680A Calibration
 
 Hi
 
 Do you have a dual channel counter that you can put the GPS into on the
 start and the FE into as the stop? The HP 5334, 5335, and 5345 are all
 examples of this sort of counter.
 
 Bob
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Chris Stake
 Sent: Wednesday, March 14, 2012 12:49 PM
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] FE5680A Calibration
 
 Nice idea,
 But I haven't got a DSO and the persistence of my scope isn't good enough
 to
 view 10Mhz sampled at 1pps.
 Chris
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of EB4APL
 Sent: 14 March 2012 15:46
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] FE5680A Calibration
 
 My fast approach would be to trigger a scope with the 1 PPS from the GPS
 receiver and observe the how the 10 MHz output of your Rb drifts.  1
 full cycle per second is 1 e-7 so you'll need to use an stopwatch to
 time long periods when you are fine adjusting .
 Building (or buying) a GPSDO allows yo to make the comparison between
 both 10 MHz outputs without the jitter in the GPS receiver 1 PPS.
 Probably others in this list can suggest more elaborated and convenient
 approaches to this.
 
 On 14/03/2012 16:17, Chris Stake wrote:
 Hello all,
 
 I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it
 to a
 16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to
 drive
 a LED for the locked signal. I can communicate with it using the
 excellent
 Fe5680Calibrator software. As received the frequency offset is set to
 Zero
 and the unit seems to work well.
 
 I plan to put it into service as a workshop frequency reference and so
 would
 like to set it close to 10.0Mhz. I have various oscillator modules,
 signal
 generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
 test-gear. I live in a river valley in North Devon UK so radio
 reception
 is
 a bit restricted but I receive digital television via a masthead
 amplifier
 and long downlead.
 
 Could someone please suggest a way of going about this?
 
 
 
 Regards
 
 Chris Stake
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread Chris Albertson
On Wed, Mar 14, 2012 at 8:17 AM, Chris Stake st...@btinternet.com wrote:
 Hello all,

 I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it to a
 16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to drive
 a LED for the locked signal. I can communicate with it using the excellent
 Fe5680Calibrator software. As received the frequency offset is set to Zero
 and the unit seems to work well.

What you need to do is build a GPSDO using your FE5680A as the O.
You can use digital logic or a human in the control loop.  Either way
it works the same.

Get a 7400 series flip flop  A 74AC74 costs about 50 cents.  Wite it
so the PPS for the GPS flips it on and the next edge of the 10MHz out
from the FE5680 re-sets the signal.  A simple XOR can also work.

Now what you do is measure the time the signal is high.  The classic
way is to use the signal to charge a capacitor then measure the
voltage on the cap.  So now you can use a DMM to measure  time.

The controller adjusts the FE5680A's frequency so as to keep the time
the signal is high (and the voltage on the cap) at some arbitrary
fixed value

People have built the above using all kinds of methods, from pure
analog to micro controllers, 7400 logic and even a human in the loop.
   If you want to build a working system fast look at using an Arduino
and divide the 10Mhz signal down by at least 10 or 20 so that (1) the
speed is compatible with solderless breadboards and the flip flop's
output will stay high longer and will be easier to measure.


Chris Albertson
Redondo Beach, California

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] FE5680A Calibration

2012-03-14 Thread David
I picked up a gimpy Beckman UC10 universal counter not long ago for
about $10 from Ebay.  Even better, I just repaired a Tektronix 7D15
(it has a whole board full of those junk TI integrated circuit sockets
which need to be replaced) although you need to leave an entire
oscilloscope mainframe on to use it.  The advantage of the later is
adjustable slope, sensitivity, and triggering.

On Wed, 14 Mar 2012 17:50:28 -0400, Bob Camp li...@rtty.us wrote:

At least the don't mess to much with it part has sunk in. That puts you 
ahead of most people at this point.

A usable counter should be a sub $100 sort of thing either at auction or 
surplus. With some careful shopping it can be a sub $40 item.

On Mar 14, 2012, at 4:25 PM, Chris Stake st...@btinternet.com wrote:

 Hi Bob,
 Thanks for pointing-out the noisy output of the FE5680A. I'll probably try
 to lock a crystal oscillator to it.
 Unfortunately I don't have a suitable counter.
 I was hoping I could use some sort of higher frequency standard but I
 confess I had not really grasped the fact that the unit may well be within
 millihertz of the nominal frequency: too delicate to twiddle with anything
 other than precision equipment and long timebases.
 Kind regards
 Chris
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Bob Camp
 Sent: 14 March 2012 16:57
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] FE5680A Calibration
 
 Hi
 
 Do you have a dual channel counter that you can put the GPS into on the
 start and the FE into as the stop? The HP 5334, 5335, and 5345 are all
 examples of this sort of counter.
 
 Bob
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Chris Stake
 Sent: Wednesday, March 14, 2012 12:49 PM
 To: 'Discussion of precise time and frequency measurement'
 Subject: Re: [time-nuts] FE5680A Calibration
 
 Nice idea,
 But I haven't got a DSO and the persistence of my scope isn't good enough
 to
 view 10Mhz sampled at 1pps.
 Chris
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of EB4APL
 Sent: 14 March 2012 15:46
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] FE5680A Calibration
 
 My fast approach would be to trigger a scope with the 1 PPS from the GPS
 receiver and observe the how the 10 MHz output of your Rb drifts.  1
 full cycle per second is 1 e-7 so you'll need to use an stopwatch to
 time long periods when you are fine adjusting .
 Building (or buying) a GPSDO allows yo to make the comparison between
 both 10 MHz outputs without the jitter in the GPS receiver 1 PPS.
 Probably others in this list can suggest more elaborated and convenient
 approaches to this.
 
 On 14/03/2012 16:17, Chris Stake wrote:
 Hello all,
 
 I purchased a FE5680A from a Chinese  Ebay vendor. I have connected it
 to a
 16.5V laptop supply, added 7805-based 5V rail and a PMOS Fet switch to
 drive
 a LED for the locked signal. I can communicate with it using the
 excellent
 Fe5680Calibrator software. As received the frequency offset is set to
 Zero
 and the unit seems to work well.
 
 I plan to put it into service as a workshop frequency reference and so
 would
 like to set it close to 10.0Mhz. I have various oscillator modules,
 signal
 generators, SDR receivers, GPS receivers, a scanner, some uncalibrated
 test-gear. I live in a river valley in North Devon UK so radio
 reception
 is
 a bit restricted but I receive digital television via a masthead
 amplifier
 and long downlead.
 
 Could someone please suggest a way of going about this?
 
 
 
 Regards
 
 Chris Stake
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-
 bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-
 nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions