Re: [time-nuts] Another use for a Trimble Thunderbolt

2014-12-26 Thread WarrenS via time-nuts

Authur wrote:
Take a look at the plot where I adjust a rubidium standard and see what you 
think.


I think it is a great idea, and shows that LadyHeather and a modified TBolt 
can make a great high end, stand alone Time-Nut tester without the need of a 
reference osc or offset osc, or any other special counters.
(your plots show a great antenna and position setup, an important first step 
to get the low phase noise desired)


To speed up the initial frequency tuning, what I've found helps is to set 
LadyHeather's display filter to around 100 sec, then the freq offset plot 
can be used to course tune the Rb's freq to under 1e-11 in just a few 
minutes.
After the course freq adj, if you want better than that, the phase plot can 
be used to measure freq, tempco,  and/or aging of any 10 MHz osc with better 
than 1e-13 / day resolution or 1e-14 using a week long test run.
That should be useful for any time nut's Oscillator, even those with Cs and 
Hydrogen Masers.


Attached is a 13 day lady Heather test plot of a LPRO Rb that I did a few 
years back using a Tbolt modified to use an external Osc.
The plot shows a best case undisciplined drift rate of 20 ns change over 3.5 
days, 6ns /day,  which equals to a 0.67e-13 freq error  during days 9-11,

Then right after that at day 12, 5ns /hr for a freq error of ~1e-12.

ws

**


Date: Thu, 25 Dec 2014 20:11:23 -0500
From: Arthur Dent
Subject: [time-nuts] Another use for a Trimble Thunderbolt


’d say that the plot is telling the truth. It also seems to be giving
you information fast enough that thermal drift and barometric pressure
is not to big an issue. If you had to wait a day or three for the same
data, drift would be a much bigger issue. Yes, when you get to the
“close enough” trace, drift may be an issue.  (yes close enough is
indeed close enough …).


Keep in mind that I'm talking about using a GPS signal from a Thunderbolt
to adjust a common rubidium standard that would be used in a telco or
other piece of general test equipment and thermal drift and barometric
pressure effects are never an issue for me.


I suspect that if you try the trick with something way far off frequency
(many 10’s of ppm), the GPS may not play nice. At any normal tune range
on an Rb, it should be fine.


Actually it does play nice-very nice over any range I'm interested in. 
Keep

in mind that I wanted a simple method that would work with a 10 Mhz
frequency
standard to give me closer readings than I could get by watching the scope
or
the counter. I can easily use just the counter to check the frequency of a
less than stellar oscillator so what I'm describing would be used with a
fairly close 10 Mhz frequency standard and not one that isn't even close.
The Pendulum CNT-81 frequency counter I have can display a 10 Mhz error to
5
decimal places in 10 seconds using the math function and an external time
base.

Anyone who has used a WWVB comparator remembers the plot zipping back to
the
zero position when the plotted frequency difference would exceed the
chart's
maximum deflection. The Thunderbolt's display on Lady Heather works 
exactly

the same way. If you look at the plots in the link that follows you will
see that the 10 Mhz appears very stable but it is actually set by a
synthesizer
to be 10,000,000.025000 hz in the upper trace and so to keep it in the
vertical
center position on the graph I have an oscillator offset of -2500 PPT in
Lady
Heather. In the lower trace the synthesizer frequency is set to
10,000,000.010 hz
and the offset is -1000 PPT to keep the 10 Mhz trace centered. The
reference for
the synthesizer and the Thunderbolt is the GPS signal from the Tbolt so 
the

same
reference is used for everything.

http://s906.photobucket.com/user/rjb1998/media/tboltplots_zpsd20a083b.jpg.html

-Arthur




On Dec 24, 2014, at 8:28 PM, Arthur Dent golgarfrinc...@gmail.com wrote:

Those of you who know I had hacked the RFTG-u REF 1 GPS years
ago and had one running for 4 years before other time nuts
discovered these units probably won't be too surprised that
I have tried another hack that may have limited interest but
works for me.

Having owned a large number of Thunderbolts, I ran across a
few that needed repairs of various sorts. One of these had
a defective oscillator so I removed the OXCO and brought the
EFC and 10Mhz connections out through the side of the case with
SMA connectors so I could test various oscillators, as others
have done before. Then I got to thinking that if I connected the
Thunderbolt up to run and output to Lady Heather but connected
a free running oscillator to the 10Mhz input, ignoring the EFC
connection, it might work as a comparator to plot the drift of
the free running oscillator. I have a few Efratom/Datum Rubidium
standards I'm adjusting and I can watch drift on my scope at 5
ns/cm or the 10 Mhz output to the 5th decimal place on my Pendulum
CNT-81 counter and try to determine which way it's drifting but
that gets old

[time-nuts] Another use for a Trimble Thunderbolt

2014-12-25 Thread Arthur Dent
’d say that the plot is telling the truth. It also seems to be giving
you information fast enough that thermal drift and barometric pressure
is not to big an issue. If you had to wait a day or three for the same
data, drift would be a much bigger issue. Yes, when you get to the
“close enough” trace, drift may be an issue.  (yes close enough is
indeed close enough …).

Keep in mind that I'm talking about using a GPS signal from a Thunderbolt
to adjust a common rubidium standard that would be used in a telco or
other piece of general test equipment and thermal drift and barometric
pressure effects are never an issue for me.

I suspect that if you try the trick with something way far off frequency
(many 10’s of ppm), the GPS may not play nice. At any normal tune range
on an Rb, it should be fine.

Actually it does play nice-very nice over any range I'm interested in. Keep
in mind that I wanted a simple method that would work with a 10 Mhz
frequency
standard to give me closer readings than I could get by watching the scope
or
the counter. I can easily use just the counter to check the frequency of a
less than stellar oscillator so what I'm describing would be used with a
fairly close 10 Mhz frequency standard and not one that isn't even close.
The Pendulum CNT-81 frequency counter I have can display a 10 Mhz error to
5
decimal places in 10 seconds using the math function and an external time
base.

Anyone who has used a WWVB comparator remembers the plot zipping back to
the
zero position when the plotted frequency difference would exceed the
chart's
maximum deflection. The Thunderbolt's display on Lady Heather works exactly
the same way. If you look at the plots in the link that follows you will
see that the 10 Mhz appears very stable but it is actually set by a
synthesizer
to be 10,000,000.025000 hz in the upper trace and so to keep it in the
vertical
center position on the graph I have an oscillator offset of -2500 PPT in
Lady
Heather. In the lower trace the synthesizer frequency is set to
10,000,000.010 hz
and the offset is -1000 PPT to keep the 10 Mhz trace centered. The
reference for
the synthesizer and the Thunderbolt is the GPS signal from the Tbolt so the
same
reference is used for everything.

http://s906.photobucket.com/user/rjb1998/media/tboltplots_zpsd20a083b.jpg.html

-Arthur
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Another use for a Trimble Thunderbolt

2014-12-25 Thread Bob Camp

 On Dec 25, 2014, at 8:11 PM, Arthur Dent golgarfrinc...@gmail.com wrote:
 
 ’d say that the plot is telling the truth. It also seems to be giving
 you information fast enough that thermal drift and barometric pressure
 is not to big an issue. If you had to wait a day or three for the same
 data, drift would be a much bigger issue. Yes, when you get to the
 “close enough” trace, drift may be an issue.  (yes close enough is
 indeed close enough …).
 
 Keep in mind that I'm talking about using a GPS signal from a Thunderbolt
 to adjust a common rubidium standard that would be used in a telco or
 other piece of general test equipment and thermal drift and barometric
 pressure effects are never an issue for me.

Your  15 ns / hour limit comes out to ~ 4.12 x10^-12. 
Your “close enough” comes out (eyeball @ 1/4 slope)  to  1x10^-12.

If your Rb has a tempco of +/- 2x10^-10 over 100 C that *might* be 4x10^-12 / 
C. If your room moves +/-2 C per hour (as some do) temperature could be an 
issue. You would be seeing 1.6x10^-11 cycles under those conditions. In order 
to reasonably “see” your 1x10^-12 limit, the setup you are running would have 
to be at least 10X better. At that point temperature rather than set point 
would dominate the plot. In order to measure set point, you would need 100X 
better. 

Are all Rb’s “worst case?”, of course not. Are they all polite and straight 
line linear over the entire range? - not on the ones I’ve seen. Throw in things 
like a draft on the heat sink when the HVAC fires up ….

—

This is *not* in any way a knock on the approach. It seems to work very well 
for what you are trying to do. It’s well though out and functional. It’s simply 
a caution that drift can be an issue doing this sort of thing to the 1x10^-12 
level on Telco Rb’s. 

Bob


 
 I suspect that if you try the trick with something way far off frequency
 (many 10’s of ppm), the GPS may not play nice. At any normal tune range
 on an Rb, it should be fine.
 
 Actually it does play nice-very nice over any range I'm interested in. Keep
 in mind that I wanted a simple method that would work with a 10 Mhz
 frequency
 standard to give me closer readings than I could get by watching the scope
 or
 the counter. I can easily use just the counter to check the frequency of a
 less than stellar oscillator so what I'm describing would be used with a
 fairly close 10 Mhz frequency standard and not one that isn't even close.
 The Pendulum CNT-81 frequency counter I have can display a 10 Mhz error to
 5
 decimal places in 10 seconds using the math function and an external time
 base.
 
 Anyone who has used a WWVB comparator remembers the plot zipping back to
 the
 zero position when the plotted frequency difference would exceed the
 chart's
 maximum deflection. The Thunderbolt's display on Lady Heather works exactly
 the same way. If you look at the plots in the link that follows you will
 see that the 10 Mhz appears very stable but it is actually set by a
 synthesizer
 to be 10,000,000.025000 hz in the upper trace and so to keep it in the
 vertical
 center position on the graph I have an oscillator offset of -2500 PPT in
 Lady
 Heather. In the lower trace the synthesizer frequency is set to
 10,000,000.010 hz
 and the offset is -1000 PPT to keep the 10 Mhz trace centered. The
 reference for
 the synthesizer and the Thunderbolt is the GPS signal from the Tbolt so the
 same
 reference is used for everything.
 
 http://s906.photobucket.com/user/rjb1998/media/tboltplots_zpsd20a083b.jpg.html
 
 -Arthur
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Another use for a Trimble Thunderbolt

2014-12-24 Thread Arthur Dent
Those of you who know I had hacked the RFTG-u REF 1 GPS years
ago and had one running for 4 years before other time nuts
discovered these units probably won't be too surprised that
I have tried another hack that may have limited interest but
works for me.

Having owned a large number of Thunderbolts, I ran across a
few that needed repairs of various sorts. One of these had
a defective oscillator so I removed the OXCO and brought the
EFC and 10Mhz connections out through the side of the case with
SMA connectors so I could test various oscillators, as others
have done before. Then I got to thinking that if I connected the
Thunderbolt up to run and output to Lady Heather but connected
a free running oscillator to the 10Mhz input, ignoring the EFC
connection, it might work as a comparator to plot the drift of
the free running oscillator. I have a few Efratom/Datum Rubidium
standards I'm adjusting and I can watch drift on my scope at 5
ns/cm or the 10 Mhz output to the 5th decimal place on my Pendulum
CNT-81 counter and try to determine which way it's drifting but
that gets old pretty fast.

The 10 Mhz output from Lady Heather appears to be an instantaneous
reading so that always looks very good but the PPS output appears
to be the cumulative signed difference between the GPS and the free
running oscillator. The link is to a plot from Lady Heather with
just the 10 Mhz and PPS signals on the screen. The EFC is still
trying to control the oscillator but seeing it isn't connected
the readings could range from a meaningless 0-5 volts and I don't
care about the temp plot either. I also know that there are other
ways of doing this but the definition of yankee inginuity' is
doing things the hard way. I could also check an RFTG-u REF 1 with
the antenna off to see how well it keeps to the correct frequency
on holdover. I suspect that like the Z3801 it tries to predict and
adjust the output when the GPS signal goes away. Take a look at the
plot where I adjust a rubidium standard and see what you think.

http://s906.photobucket.com/user/rjb1998/media/TboltRbadjF2_zps3a1a9922.jpg.html

-Arthur
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Another use for a Trimble Thunderbolt

2014-12-24 Thread Bob Camp
Hi


 On Dec 24, 2014, at 8:28 PM, Arthur Dent golgarfrinc...@gmail.com wrote:
 
 Those of you who know I had hacked the RFTG-u REF 1 GPS years
 ago and had one running for 4 years before other time nuts
 discovered these units probably won't be too surprised that
 I have tried another hack that may have limited interest but
 works for me.
 
 Having owned a large number of Thunderbolts, I ran across a
 few that needed repairs of various sorts. One of these had
 a defective oscillator so I removed the OXCO and brought the
 EFC and 10Mhz connections out through the side of the case with
 SMA connectors so I could test various oscillators, as others
 have done before. Then I got to thinking that if I connected the
 Thunderbolt up to run and output to Lady Heather but connected
 a free running oscillator to the 10Mhz input, ignoring the EFC
 connection, it might work as a comparator to plot the drift of
 the free running oscillator. I have a few Efratom/Datum Rubidium
 standards I'm adjusting and I can watch drift on my scope at 5
 ns/cm or the 10 Mhz output to the 5th decimal place on my Pendulum
 CNT-81 counter and try to determine which way it's drifting but
 that gets old pretty fast.
 
 The 10 Mhz output from Lady Heather appears to be an instantaneous
 reading so that always looks very good but the PPS output appears
 to be the cumulative signed difference between the GPS and the free
 running oscillator. The link is to a plot from Lady Heather with
 just the 10 Mhz and PPS signals on the screen. The EFC is still
 trying to control the oscillator but seeing it isn't connected
 the readings could range from a meaningless 0-5 volts and I don't
 care about the temp plot either. I also know that there are other
 ways of doing this but the definition of yankee inginuity' is
 doing things the hard way. I could also check an RFTG-u REF 1 with
 the antenna off to see how well it keeps to the correct frequency
 on holdover. I suspect that like the Z3801 it tries to predict and
 adjust the output when the GPS signal goes away. Take a look at the
 plot where I adjust a rubidium standard and see what you think.

’d say that the plot is telling the truth. It also seems to be giving you 
information fast enough that thermal drift and barometric pressure is not to 
big an issue. If you had to wait a day or three for the same data, drift would 
be a much bigger issue. Yes, when you get to the  “close enough” trace, drift 
may be an issue.  (yes close enough is indeed close enough …).

I suspect that if you try the trick with something way far off frequency (many 
10’s of ppm), the GPS may not play nice. At any normal tune range on an Rb, it 
should be fine. 

Bob

 
 http://s906.photobucket.com/user/rjb1998/media/TboltRbadjF2_zps3a1a9922.jpg.html
 
 -Arthur
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.