Re: [time-nuts] SR620 Problem

2016-08-29 Thread Alexander Pummer
how fare are you familiar with electronic equipment troubleshooting? 
could you find components, test points if somebody would ask you for 
that? do you have some electronic test equipment e.g. multimeter, 
oscilloscope? --last year I helped somebody to fix one HP high-frequency 
generator just per e-mail back and fort and he was able to fix the 
equipment, if you are interested let me know


Regards
Alex

On 8/29/2016 4:41 PM, Don@True-Cal wrote:

Sure, my original post ask for anyone in the US as I live in Springfield 
Missouri but I will be happy to ship anywhere in the states. I have the 
complete schematics for the unit. Was hoping to find someone specializing in 
SRS equipment or already having repair experience with the SR620.

Regards,
Don

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Tom Miller
Sent: Monday, August 29, 2016 4:37 PM
To: Discussion of precise time and frequency measurement <time-nuts@febo.com>
Subject: Re: [time-nuts] SR620 Problem

You might say where you are. Maybe someone lives close that can assist you.

Regards,
Tom

- Original Message -
From: "Don@True-Cal" <truecalservi...@gmail.com>
To: "'Time Nuts Group'" <time-nuts@febo.com>
Sent: Monday, August 29, 2016 5:17 PM
Subject: [time-nuts] SR620 Problem



Hello All,



Sorry if this is a repeat but not showing up in my sent mail. Changed the
list address slightly.



I'm looking for someone that has repair experience on the SR620 TIC. I
have
a failed unit that has previously worked perfectly for many years while
running continuously on my lab bench. The failure occurred after a very
dirty mains power failure during a storm. There was no other evidence of a
severe lightning strike so I suspect just a major power line glitch. I
have
a second unit so time to repair is not critical. I have checked and
confirmed all of the power supplies and eliminated all of the easy fix
scenarios already. The major symptom is absolutely no display of
indicators
or readout (unit looks to be off), -DROPOUT form the PS is ok. I believe
the
loss of LED and display is from the processor stopped U131B flop. I don't
have sufficient confidence, let alone spare IC to delve into the processor
chain. Is there someone in the US with this experience that I can send
this
unit to for repair. Being retired and all this just being a major hobby
makes repair cost a big concern.



Regards.

Don J.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


-
No virus found in this message.
Checked by AVG - www.avg.com
Version: 2016.0.7752 / Virus Database: 4649/12903 - Release Date: 08/29/16


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] SR620 Problem

2016-08-29 Thread Charles Steinmetz

Sorry if this is a repeat but not showing up in my sent mail. Changed the
list address slightly.


Email clients lie all the time.

When there is a question of whether a mail to the list was distributed, 
it is very useful to go to the list web page (address is at the bottom 
of every list message), select the list archive, and see if the message 
is in the archive.  If it is, it was distributed, whether or not you can 
find your copy of the sent or received message.


This simple step can very often save one from needlessly re-sending a 
message that everyone else already received.


Best regards,

Charles


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] SR620 Problem

2016-08-29 Thread Don@True-Cal
Sure, my original post ask for anyone in the US as I live in Springfield 
Missouri but I will be happy to ship anywhere in the states. I have the 
complete schematics for the unit. Was hoping to find someone specializing in 
SRS equipment or already having repair experience with the SR620. 

Regards,
Don

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Tom Miller
Sent: Monday, August 29, 2016 4:37 PM
To: Discussion of precise time and frequency measurement <time-nuts@febo.com>
Subject: Re: [time-nuts] SR620 Problem

You might say where you are. Maybe someone lives close that can assist you.

Regards,
Tom

- Original Message - 
From: "Don@True-Cal" <truecalservi...@gmail.com>
To: "'Time Nuts Group'" <time-nuts@febo.com>
Sent: Monday, August 29, 2016 5:17 PM
Subject: [time-nuts] SR620 Problem


> Hello All,
>
>
>
> Sorry if this is a repeat but not showing up in my sent mail. Changed the
> list address slightly.
>
>
>
> I'm looking for someone that has repair experience on the SR620 TIC. I 
> have
> a failed unit that has previously worked perfectly for many years while
> running continuously on my lab bench. The failure occurred after a very
> dirty mains power failure during a storm. There was no other evidence of a
> severe lightning strike so I suspect just a major power line glitch. I 
> have
> a second unit so time to repair is not critical. I have checked and
> confirmed all of the power supplies and eliminated all of the easy fix
> scenarios already. The major symptom is absolutely no display of 
> indicators
> or readout (unit looks to be off), -DROPOUT form the PS is ok. I believe 
> the
> loss of LED and display is from the processor stopped U131B flop. I don't
> have sufficient confidence, let alone spare IC to delve into the processor
> chain. Is there someone in the US with this experience that I can send 
> this
> unit to for repair. Being retired and all this just being a major hobby
> makes repair cost a big concern.
>
>
>
> Regards.
>
> Don J.
>
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there. 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] SR620 Problem

2016-08-29 Thread Alexander Pummer

Hi Tom, do you have circuit diagram for the unit?

73

Alex


On 8/29/2016 2:37 PM, Tom Miller wrote:
You might say where you are. Maybe someone lives close that can assist 
you.


Regards,
Tom

- Original Message - From: "Don@True-Cal" 
<truecalservi...@gmail.com>

To: "'Time Nuts Group'" <time-nuts@febo.com>
Sent: Monday, August 29, 2016 5:17 PM
Subject: [time-nuts] SR620 Problem



Hello All,



Sorry if this is a repeat but not showing up in my sent mail. Changed 
the

list address slightly.



I'm looking for someone that has repair experience on the SR620 TIC. 
I have

a failed unit that has previously worked perfectly for many years while
running continuously on my lab bench. The failure occurred after a very
dirty mains power failure during a storm. There was no other evidence 
of a
severe lightning strike so I suspect just a major power line glitch. 
I have

a second unit so time to repair is not critical. I have checked and
confirmed all of the power supplies and eliminated all of the easy fix
scenarios already. The major symptom is absolutely no display of 
indicators
or readout (unit looks to be off), -DROPOUT form the PS is ok. I 
believe the
loss of LED and display is from the processor stopped U131B flop. I 
don't
have sufficient confidence, let alone spare IC to delve into the 
processor
chain. Is there someone in the US with this experience that I can 
send this

unit to for repair. Being retired and all this just being a major hobby
makes repair cost a big concern.



Regards.

Don J.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there. 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


-
No virus found in this message.
Checked by AVG - www.avg.com
Version: 2016.0.7752 / Virus Database: 4649/12903 - Release Date: 
08/29/16


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] SR620 Problem

2016-08-29 Thread Tom Miller

You might say where you are. Maybe someone lives close that can assist you.

Regards,
Tom

- Original Message - 
From: "Don@True-Cal" <truecalservi...@gmail.com>

To: "'Time Nuts Group'" <time-nuts@febo.com>
Sent: Monday, August 29, 2016 5:17 PM
Subject: [time-nuts] SR620 Problem



Hello All,



Sorry if this is a repeat but not showing up in my sent mail. Changed the
list address slightly.



I'm looking for someone that has repair experience on the SR620 TIC. I 
have

a failed unit that has previously worked perfectly for many years while
running continuously on my lab bench. The failure occurred after a very
dirty mains power failure during a storm. There was no other evidence of a
severe lightning strike so I suspect just a major power line glitch. I 
have

a second unit so time to repair is not critical. I have checked and
confirmed all of the power supplies and eliminated all of the easy fix
scenarios already. The major symptom is absolutely no display of 
indicators
or readout (unit looks to be off), -DROPOUT form the PS is ok. I believe 
the

loss of LED and display is from the processor stopped U131B flop. I don't
have sufficient confidence, let alone spare IC to delve into the processor
chain. Is there someone in the US with this experience that I can send 
this

unit to for repair. Being retired and all this just being a major hobby
makes repair cost a big concern.



Regards.

Don J.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there. 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] SR620 Problem

2016-08-29 Thread Don@True-Cal
Hello All,

 

Sorry if this is a repeat but not showing up in my sent mail. Changed the
list address slightly.

 

I'm looking for someone that has repair experience on the SR620 TIC. I have
a failed unit that has previously worked perfectly for many years while
running continuously on my lab bench. The failure occurred after a very
dirty mains power failure during a storm. There was no other evidence of a
severe lightning strike so I suspect just a major power line glitch. I have
a second unit so time to repair is not critical. I have checked and
confirmed all of the power supplies and eliminated all of the easy fix
scenarios already. The major symptom is absolutely no display of indicators
or readout (unit looks to be off), -DROPOUT form the PS is ok. I believe the
loss of LED and display is from the processor stopped U131B flop. I don't
have sufficient confidence, let alone spare IC to delve into the processor
chain. Is there someone in the US with this experience that I can send this
unit to for repair. Being retired and all this just being a major hobby
makes repair cost a big concern.

 

Regards.

Don J.

 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.