[time-nuts] HP 5065A, no 2nd harmonic.

2021-11-28 Thread cdelect
With that offset either the oven set point has changed or there has een a component value shift in the oscillator circuit. The easiest way to test is to stick in another 10811 if available. You could also install one of the original 105 style oscillators if you have one. Cheers, Corby

[time-nuts] HP 5065A, no 2nd harmonic.

2021-11-20 Thread cdelect
Jared, There is a non-hermetic electrolytic cap in the A7 (100uf C3) that if it leaks can "eat" the lead or trace off killing the gain.. Pull the A7 and take a look at that. Cheers, Corby ___ time-nuts mailing list -- time-nuts@lists.febo.com -- To

[time-nuts] HP 5065A, no 2nd harmonic.

2021-11-18 Thread Jared Cabot via time-nuts
Hi all, I have a 5065A Rubidium Standard that was working ok, but has lost the 2nd harmonic. My voltages are all ok, the physics package hasn't overheated and I have replaced the usual resistors on the lamp assembly (and it was running fine afterward for a while), the lamp illuminates, and