Bob,
to me your ADEV plot looks pretty exactly like a measurement system
noise floor plot with its typical 10/decade shape.
I would assume that you connected the same signal to both counter
inputs, or, both sources have been phase locked.
Adrian
Bob Stewart schrieb:
Hi Bob,
I take
-f.html
I'm just not sure what temperature rating the original fuse has.
Maybe someone knows?
Adrian
Dan Rae schrieb:
I'm sure this has been discussed to death before, but does anyone know
of a source of thermal fuses small enough to fit in the -hp- 10811? I
have found plenty of larger ones, none
Magnus,
here is at least some information on the B-5400:
http://tf.boulder.nist.gov/general/pdf/26.pdf
That is the paper mentioned in the 3200 manual.
Btw. do you have any information on the B-1325?
Cheers,
Adrian
Magnus Danielson schrieb:
Bert,
I hope this will satisfy your curiosity
resolution of 50 ps is insufficient because up to 100
sec, to the most part of it, the actual ADEV of the oscillators can be
below the measurement limit of the counter.
Adrian
Karen Tadevosyan schrieb:
Thanks again for your explanations and advices.
I raised this question as a ADEV measurement
I got a nice old Varian R-20 rubidium frequency standard that would be
nice to get working.
To begin with, there is at least a problem with the voltage regulator
circuit.
Does anyone have a manual? I couldn't locate anything on the web.
Regards,
Adrian
John,
thank you for your post on the time-nuts list.
Unfortunately, I couldn't locate the article you are refering to. All I
could find are just two associated files, a BOM and an Excel sheet.
Could you please send me a link or a copy of the article?
Thanks,
Adrian
This article discusses timing errors due to mismatch and multiple
reflections in transmissin lines:
http://www.dtic.mil/dtic/tr/fulltext/u2/a508044.pdf
Adrian
Edgardo Molina schrieb:
Dear Group,
Good evening. I just arrived home after the first day of conferences at the
Electrical Metrology
what has happened or
what to look for first?
Regards,
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
W max. at turn on (mine draws some
9 W), and Steady state power drops to approximately 2 W at 25°C in still
air at 20 V (mine draws some 1.9 W at 12 V without powering the outer oven).
Adrian
Azelio Boriani schrieb:
And, why 12V for the oven? The 10811A manual states from 20V to 30V
and capturing noise.
Btw. the Sprague 6800 uF / 40 V from my RS NGA power supply has indeed
died, but independently of the osc. problem.
Adrian
gandal...@aol.com schrieb:
Perhaps a silly question, but as I get the impression that both seem to
have failed simultaneously with the same fault
to find an original diode (even though
I could leave it as is).
As by the manual, the original is a '0122-0244 DIODE-VVC 100PF 5%
C4/C25-MIN=2 BVR-30V Mfr. 28480'
Mfr. 28480 is listed as HP, but the diode is marked 'M' for Motorola and
'0244' and '312'.
Adrian
circuitry left, mostly the oscillator stage itself, the
tunig circuit and power supply.
Adrian
John Miles schrieb:
That's extremely interesting, Adrian. I've never heard of a noisy varactor,
but then I've never looked for one, either. It'd be great if some of the
problems that have been blamed
noise has gone away since I replaced the diode.
The noise was clearly there, and it was also there using a test adapter
and two lab power supplies.
Adrian
Rick Karlquist schrieb:
I wonder if the glass case got a crack in it or if the kovar seal
was failing. Maybe a failed solder joint (which gets
stopped working.
Is there any information about the Sulzer desiagn available? Anyone who
repaired one of these oscillators?
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time
John,
yes, I have a scan. It's the one from the Agilent site.
There are just some pages mixed up. Nothing appears to be missing.
The pages are in this sequence: 3-5, 3-6, 3-9, 3-10, 3-7, 3-8, 3-11,
3-12 etc.
Adrian
John Ackermann N8UR schrieb:
Does anyone have a copy of the 5061B ops manual
to Sulzer's brilliant idea to include an independent inner oven
temperature indicator, I was able to verify that nothing has changed.
Now I'm waiting for the frequency to stabilize again, which I'm
expecting to happen within the next 2...4 weeks.
Adrian
Magnus Danielson schrieb:
Fellow time
, but the oscillator stuff
appears to be exactly the same.
Adrian
paul swed schrieb:
Thanks for the pointer I have a pretty dirty pdf from some place the links
very clean.
Paul
WB8TSL
On Thu, Nov 1, 2012 at 7:24 PM, Magnus Danielson mag...@rubidium.dyndns.org
wrote:
On 11/01/2012 10:27 PM, Attila Kinali
Anyone have a Wavecrest DTS 2075/77/79 service manual?
I've recently got a DTS 2077 that however turned out to be defective.
There was a power supply failure that I could fix, but there is still a
power-up selftest error message and calibratin fails.
Adrian
and / or your reference, or simply the noise floor of
your test setup. This can be the hardest part. You might want to compare
your test results with some trustworthy reference measurements like
this: http://www.leapsecond.com/museum/manyadev.gif
Adrian
John Miles schrieb:
Hi, Edgardo
understand that I was no longer motivated to investigate
further, especially because the '131 showed much less 'improvement' as a
'132 which I saw a plot of but don't have access to.
Cheers,
Adrian
Magnus Danielson schrieb:
Adrian,
On 15/11/12 03:59, Adrian wrote:
Hi Edgardo,
however, the 53132A
as door stops of land fill
once defective... I've never been confronted with such a poor customer
sevice attitude before.
Nevertheless... is there anyone who has acquired some experience with
the DTS-series innards who would be willing to share his knowledge?
Adrian
Adrian schrieb:
Anyone
('Test Interval Quantizer
Board').
It seems like the failure is IQM board related. When I run calibration
with the IQM board pulled, my unit still behaves the same way.
Adrian
Henk schrieb:
Hi,
The Wavecrest DTS 207X has a number of proms and also battery back-up
devices. I think it will be wise
count me in.
Adrian
cdel...@juno.com schrieb:
As Bert mentioned once the amount if interest is established purchase
details will post.
We also will post a FAQ for this project.
Thanks,
Corby
The New #34;Skinny#34; Fruit
How
+23 dBm signals at hands, but I don't see that the ZRPD-1
can compete with the 11848A PD or an external +23 dBm mixer at input
levels of +17 dBm and above.
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https
follow me there! These can
be challenging to fix if not impossible. PAL's and other unobtainable
'field programmables' appear to die faster than anything else in that
boxes. Finally, I needed three of them to get one working unit...)
Adrian
Bruce Griffiths schrieb:
Adrian wrote:
Bruce Griffiths schrieb:
Demian Martin wrote:
I asked Wenzel about mixers for phase noise measurement and they
directed me
to Marki Microwave as what they use:
http://www.markimicrowave.com/2770/Mixers.aspx I have not obtained or
tested any myself
into digital signal processing before the
3562A came...
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions
business to size.
I only get a HP power cord listed.
Adrian
cdel...@juno.com schrieb:
I'm selling a few of these cards. See eBay item 321033088669
I also have a quantity of the bath tub style from HP 5061 Cesium
standards. (If I remember 5, 8, and 3uf available)
Built specially for long time constant
I just found an interesting new Symmetricom product
http://www.symmetricom.com/products/test-and-measurement/phase-noise-allan-deviation-test-sets/3120A-Phase-Noise-Test-Probe/
It looks familiar, doesn't it?
Adrian
___
time-nuts mailing list
For phase noise the frequency range is 1MHz to 8/26.5/50GHz
The spectrum analyzer works from 20Hz to max.
Adrian
Azelio Boriani schrieb:
Isn't the FSUP a 110K euros equipment 20Hz-50GHz capable? 125KHz shouldn't
be a problem. I had an FSUP for 25 seconds to play with... really
impressive
a RAY-3. It goes from 70kHz to 200MHz.
The IF must be specified from DC, which for the above is the case.
Between mixer and baseband input a lowpass filter is required to
suppress the sum signal (2x f_input) sufficiently.
Adrian
Azelio Boriani schrieb:
Yes, I have taken a look and the FSUP
hours, but it would be nice to know how to handle phase wraps properly.
In this case, both, the reference and input frequency are normalized to
1 Hz, so the actual input frequency makes no difference.
Adrian
John Miles schrieb:
Hi, Luciano --
The sampling interval would be whatever rate
, use TimeLab and a TI counter!
Regards,
Adrian
Chris Howard schrieb:
On 12/29/2012 7:17 PM, Magnus Danielson wrote:
I would recommend to use a TIC and TimeLab to assist you on that task.
It is recommended to look at the phase plot (P), frequency plot (F) and
modified Allan deviation plot (M
Hi Tom,
Bill has actually published detailed schematics etc here:
http://www.stable32.com/A%2010%20MHz%20OCVCXO%20and%20PLL%20Module.pdf
Btw. what do you think about his small DMTD system?
http://www.wriley.com/A%20Small%20DMTD%20System.pdf
Adrian
Tom Van Baak schrieb:
Hi Bob,
The TPLL
about ultra low
PS design:
http://waltjung.org/Library.html
Adrian
Richard (Rick) Karlquist schrieb:
I know this topic has been discussed in the past on the list, but
a colleague is asking if there are any off the shelf low
noise power supplies for testing oscillators. Something
a cut above
. There is an insertion loss of 6 dB
to be entered in the 3048A software when using the DC block in
conjunction with the 11848A.
Obviously, any FFT analyzer plus LNA and coupling cap can be used for
that method. The major advantage of the 3048A system is the automatic
generation of plots.
Adrian
John
battery noise measurement.
Adrian
Charles P. Steinmetz schrieb:
tvb wrote:
I'll answer the question with anther question -- how does one
properly measure power supply noise? Does it boil down to a single
number, a couple of key numbers, or is it a plot, or several plots?
There are a number
Hello,
I'm forwarding the message below on behalf of Luciano, because there
appears to be a problem with the time-nusts list that does no longer
transmit any of his messages.
Adrian
+
Hi all,
A new paper, an Ultra Low Phase Noise 5 to 10 MHz Frequency Doubler
John's fine TimeLab of course!
http://www.ke5fx.com/timelab/readme.htm
Just don't ask me which Racal counter is or isn't supported.
Adrian
Dan Rae schrieb:
I am working on 400 - 500 MHz DDS clocks for an amateur radio project
and would like to record over gpib and plot the relative frequency
%20Resonator
Adrian
Bob Camp schrieb:
Hi
That's not one of the good ones. If you look at it closely - It's mounted in
springs, not clips. The same seller does have some interesting stuff though:
http://www.ebay.com/itm/ULTRA-STABLE-BLILEY-QUARTZ-BG11-GLASS-ENVELOPE-WARNER-RESONATOR-HUGE-BLANK
with these units?
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
to get the
late A3 version pages.
I have to say that Corby's modification in its simplicity and achieved
improvement is absolutely ingenious.
Adrian
Robert Darby schrieb:
I have a question for the list that I'm sure Corby can address but
perhaps someone else can explain as well.
Corby
.
Adrian
Joseph Gray schrieb:
Thanks to all for the advice. I agree with Said that a 8560E would be
much nicer. However, I may be able to get this 8590A very
inexpensively. Considering the price difference between this unit and
a newer/better SA, I'm inclined to go with this one. I'm thinking
as soon as I
have some spare time..
Adrian
gandal...@aol.com schrieb:
In a message dated 07/03/2010 10:17:54 GMT Standard Time, f6...@wanadoo.fr
writes:
Could not locate any OM manual on the web for this equipement
can somebody help ?
-
Once again Didier saves the day
caused by load
changes.
After some 1 1/2 years of continous operation of my 5065A, I could't
measure any difference to my 5061A, even after averaging over 1 hour of
counter readings.
The FRK however suffered from massive runaway after plugging it in the
counter...
Adrian
John Miles schrieb
If you're not looking for DIY / surplus stuff, check Schomandl or Hopf
in Lüdenscheid
http://www.schomandl.de/en/
http://www.hopf.com/en/index.html
Eugen Leitl schrieb:
Hi -- a couple somewhat lunatic questions. Figured this would
be the best place to ask.
Anyone aware of a time standard
...+23 dBm mixer.
Replacing the 11729C PD with a ultra high level mixer should get the
noise floor close to 11848A specs. It would just require to feed L and R
directly into the mixer rather than using the instrument inputs.
Any thoughts / experiences referring to this?
Adrian
specified. I wish I had some reference
measurements at hands...
Adrian
John Miles schrieb:
It's not really the mixer's fault -- there are a couple of other things to
keep in mind with the 11729. The most important is that you can't measure
10 MHz sources with it unless you install an LPF
.
Finally, I checked the filter response, and it looks ok to me, being
within +/- 0.4 dB up to 750 kHz, and +1 dB at 1 MHz, measured with the
beat note.
Adrian
John Miles schrieb:
John,
I'd say you nailed it.
After some more testing, I can confirm that the limiter amp and the LPF
before...
And, I have a noise floor of some -165 to -169 dBc/Hz at +7.5 dBm input,
and I'm seeing the 10811A's some 5 dB above that.
Adrian
John Miles schrieb:
John,
I'd say you nailed it.
After some more testing, I can confirm that the limiter amp and the LPF
are the culprit.
I opened
, but we're now refering to a carrier level of +22.5 dBm instead
of the above +7.5 dBm (+15 dBm versus 0 dBm at the mixer output).
That's why we can now measure down to -170 dBm/Hz -(+15 dBm) = -185 dBc/Hz.
Adrian
John Miles schrieb:
That sounds about right to me. I was guessing you meant 40
Minus the 6 dB L(f) conversion factor,
so we get theoretical L(f) noise floors of -176 dBc/Hz and -191 dBc/Hz.
Adrian
Adrian schrieb:
Thanks to Rick's great hint, I'm now awaiting some AM123 amps to arrive.
Why is it that I just don't get better than -165...-170 dBc/Hz?
Let me try some
tried it yesterday again, it worked for about 2 hours before the
shift appeared. Oddly enough, the green 'Locked' light was still on.
This happens every 2...20 hours.
Any ideas what could cause that failure?
Adrian
___
time-nuts mailing list -- time
old notes I see that the 'failure mode' 5 Mhz output was
5,000,000.500,32 Hz. It was always stable at +1E-07 off.
Well, it would be much easier to have a spare 1000B, but I don't.
Anyone got a 1000B that I could swap in?
Adrian
John Miles schrieb:
Sounds like the OCXO has a jumping problem
and down with a sharp maximum at every third scan
when the tuning voltage has reached its maximum.
Thus, it seems like the OCXO is the culprit.
Adrian
ewkeh...@aol.com schrieb:
If it is always stable at 5x500.32 and has the lock light on than it
has to be in the PLL.
Bert Kehren
In a message
it, the tuning voltage is
almost exactly at center scale!), and we shall see if it behaves or not.
Adrian
John Miles schrieb:
Sounds like the OCXO has a jumping problem, going too far out of range for
the CS loop to correct. Have you watched it for a long period of time with
the loop turned off
cover,
but stabilized quickly after again.
Now I'm in need of more counter resolution than the 1E-11 last digit of
my 53131A
Adrian
ewkeh...@aol.com schrieb:
Adrian, maybe I missed something which oscillator does your unit have? Bert
-Original Message-
From: Adrianrfn
Stop Bits 1
Parity Odd
Flow Control None
Hope that helps.
Adrian
Robert Benward schrieb:
To All,
An update to my Z3801A saga.
Summary: As you know, I purchased this Z3801A at the Dayton Hamfest. The
person who sold it to me said it was modified for RS-232. I powered the
unit up and the only
It's in the same menu where you start / stop communication, just below
the 'Port Open' command.
CommPort Settings Baud Rate
Adrian
Robert Benward schrieb:
The minimum voltage is +/-3 volts, I'm getting about +/-6 right now.
How does one change the serial port speed, I don't see anything
on and no signs of instability.
Adrian
Magnus Danielson schrieb:
On 06/14/2010 05:45 AM, Glenn Little WB4UIV wrote:
This is known as gold embrittlement.
The gold has to be removed before a good solder connection is made.
To remove the gold, tin the gold plated area.
Desolder the tinning.
Retin
be established, and the main form
should begin to periodically update every few seconds.
Shoot me a PM if you don't have the manual.
Btw. I've now installed Ulrich's superb Z38XX that is much more
versatile than the basic SatStat.
Adrian
Robert Benward schrieb:
Adrian,
I thought
So you're having something like that in mind?
http://www.polytron.de/Produkte/G07.pdf
4...2400 MHz / 1000V
Yuri Ostry schrieb:
Hello,
Thursday, June 17, 2010, 2:44:43, Didier Juges wrote:
D It is not clear to me why the antenna should be isolated from the
D receiver, I would prefer to keep
Thanks to your assistance that inspired me to to take the 1000B OCXO
apart, the FTS 4060 is now living happily in my frequency/time rack.
Call this a good day!
Adrian
Stanley Reynolds schrieb:
Pictures of 1000b dissembled here:
http://www.n4iqt.com/fts1000b/pictures/
Believe it is similar
Bob,
guess you know that detailed documentaion is available here:
http://www.prc68.com/I/HPE1938.shtml
http://www.leapsecond.com/museum/e1938a/
Open the can and report back!
Adrian
Robert Benward schrieb:
All,
Anybody out there know how to fix an E1938 oscillator? Something in the
metal can
a coupling cap to remove the DC component.
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Brent,
aside of the small and inexpensive module boxes, how 'bout a HP 5065A or
a Rohde Schwarz XSRM?
Both provide a lot more fun :)
Adrian
Brent Gordon schrieb:
I'm considering buying a rubidium standard, not that I really need
one, just because I'm a Time Nut. I have some questions
as the VCXO alone.
Also keep the power supply contribution into account that can be
surprisingly high.
And, the PN of most frequency standards is significantly lower than what
you can measure with any spectrum analyzer with PN measurement software
(except for the RS FSUP of course).
Adrian
needs it has way too much
gain and +5 dBm 1 dB compression makes it completely useless for PN
measurement purposes.
Adrian
Rick Karlquist schrieb:
It is good that you asked this FAQ. Basically, what is magic
about the AMC-123 is that it has certifiably low phase noise,
guaranteed by design
+1 on the 5315 / 5316A/B.
5316's come with HPIB.
Both can be surprisingly cheap, and, they have no fan.
Adrian
Robert Atkinson schrieb:
Hi Stephen,
I assume from the callsign that you are in the UK. You could do worse than to buy a
used HP 5316A for £75 from Stewart of Reading
http
It's still being made by Tyco / MaCOM:
http://www.macomtech.com/datasheets/AM-123_AMC-123.pdf
Adrian
Mark J. Blair schrieb:
On Aug 20, 2010, at 3:41 PM, Rick Karlquist wrote:
It might still be available through Tyco/MAcom. They have continued to
make selected Anzac components
and 432 MHz.
Adrian
Mark J. Blair schrieb:
On Aug 21, 2010, at 12:10 PM, Adrian wrote:
It's still being made by Tyco / MaCOM:
http://www.macomtech.com/datasheets/AM-123_AMC-123.pdf
Thanks! According to Avnet (the only one of their US distributors where I found
a price posted
, it boils down to the one question:
What is your motivation to start a new board?
Adrian
Stanley Reynolds schrieb:
Started setting up a time-nuts board at:
http://forums.n4iqt.com/modules.php?name=Forums
Need ideas for topics.
Stanley
___
time-nuts
to sync them through a reliable and reproduceable
link.
Just a quick shot though.
Adrian
Ralph Smith schrieb:
We have a requirement for approximately ten radio sites to be synchronized
to within 30 ns of each other. Ordinarily you could throw in an
appropriate GPSDO and be done with it. However, we
Martyn,
check this:
http://www.wriley.com/
Regards,
Adrian
Martyn Smith schrieb:
Hello,
Can anyone recommend software to calulate the Allan varaince from a
set of data?
I have taken many measurements with the SR620 using it's 500 second
gate and downloaded the results into Excel.
I
picture), Source is the left terminal,
drain the upper one.
Also check voltages at the output MMIC (6 pin device). Check input
(lower left) and output (upper right) as well as the supply pin (lower
right), plus the actual voltage that is coming in from the output coax
cable, and report back.
Adrian
John,
I don't have mine at hands to cross-check if I'm right, but you probably
have a Ball Efratom / Symmetricom FRS.
Adrian
John Simmons schrieb:
I have what I believe to be a rubidium frequency reference made by
Efratom. Research with Google has not helped me find info on using
a single
device from a supplyer...
...or try it here:
http://www.msc-ge.com/frame/d/wir/inhalt_hig.html
... they are also in Madrid and in Barcelona :-)
http://www.msc-ge.com/frame/d/wir/inhalt_hig.html
How sounds that?
Adrian
Jose Manuel schrieb:
Thank you very much Bruce and all who replied my
are or aren't
useful for your task. It should also show most of the functionality of
the trim pots. That way, without having turned any pots, you'll have
learned a lot about how the unit works, and that should be quite helpful
for troubleshooting.
Adrian
Mark Sims schrieb:
The rubidium lamp
sounds like an interesting parts unit... ;)
Jim Palfreyman schrieb:
Only this group will understand and suffer with this news...
Our observatory has recently had three Russian hydrogen masers arrive (along
with some Russians to install them).
It turns out that one of the masers had a rough
the ripple on the
unregulated side. You may use a smaller capacitor, just large enough for
not getting below the minimum regulator input voltage at the lowest
momentary voltage across the cap when your wall voltage is at its low
spec side...
Adrian
Bob Stewart schrieb:
The transformer gives me
How about using the 10811 from your broken Z3801A?
Adrian
Jim Sanford schrieb:
Bob:
Thanks.
You correctly anticipated that I will use this to lock a 100 MHz
Oscillator, and either multiply or lock from there. (I'm considering
a VCXO at 1GHz as well, it would either be locked to 100 MHz
diagram.
Regards,
Adrian
Graham / KE9H schrieb:
On 7/5/2013 12:23 AM, Perry Sandeen wrote:
List,
I was looking on Ebay for some HP E1938A
oscillators
What I found listed were:
HP E1938A 10 MHz Quartz Oscillator with EFC on PC
board. $100 Fluke.l
HP E1938-60201 Ovenized Crystal
sounds quite typical for a dried out electrolytic cap.
Look for the unregulated -10V that goes into the regulator circuit.
Most likely C3 on the power supply motherboard is dead.
Adrian
Mark C. Stephens schrieb:
Checked the PSU voltages on the PSU card next to the 10811.
The -5.2 is reading
for the cap to
loosen... ;)
Adrian
Mark C. Stephens schrieb:
Hi Nigel,
The only screw type electro can find is 29000uf@10V. it's the same dimensions.
Should I risk the strain on the rectifiers (another 10Kuf is rather a lot)?
Without this timer I am dead in the water so I need to do the right
Ed,
thanks for posting!
I'm still looking for a MUX board for my faulty 2077.
Is there anyone using a DTS-207x with TimeLab?
Adrian
Ed Palmer schrieb:
FYI, I did a teardown on my Wavecrest DTS-2077. It's posted here:
http://www.eevblog.com/forum/testgear/wavecrest-dts-2077-teardown
Ed
To be more precise, I have a working DTS-2077 and tried it a few times
with TimeLab.
TimeLab was always running fine for a few seconds but then it stopped
and displayed a timeout error.
So, what's wrong?
Adrian
Adrian schrieb:
Ed,
thanks for posting!
I'm still looking for a MUX board
John,
I'm using a NI GBIP-USB-B.
The monitor feature does not time out, just the acquisition itself.
Adrian
John Miles schrieb:
What GPIB adapter are you using? Does the 'Monitor' feature also time out,
or just the acquisition itself?
-- john, KE5FX
Miles Design LLC
-Original Message
worse.
Sometimes it times already out while establishing GPIB connection. Best
case it runs up to tau = some 3 seconds.
I've never observed the monitor function timing out.
Adrian
John Miles schrieb:
Hmm. Both the monitor window and acquisition loop use 5-second GPIB
timeouts, so they should
David,
it isn't even secret, just a stash of A03.02 bits 'n bytes.
Shoot me a PM.
Adrian
David Hooke schrieb:
Hi Folks,
This is from the June 1990 version of the HP 3048A System Ops manual:
Does anyone have a secret stash of the RMB A.03.0x version?
BTW, I now have a fully operational
modulation index.
So far, everything seems to work properly except for the missing 2nd
harmonic and the unusual beam curren behavior.
Any help would be much appreciated.
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https
operation outputs on the rear, and a A5 time delay divider.
There is no clock.
Thanks,
Adrian
cdel...@juno.com schrieb:
Adrian,
Most used High stability tubes are DOA but you might get lucky!
You say the low frequency test worked.
What was the amplitude of the peak in nanoamps and was the peak well
I see you already have the user manual, but here it is:
http://www.to-way.com/tf/4060man.pdf
And Brooke's 4060 site:
http://www.prc68.com/I/FTS4060.shtml
Adrian
J. L. Trantham schrieb:
Suspicious for a dead tube.
Joe
-Original Message-
From: time-nuts-boun...@febo.com [mailto:time
parts of the
control loop.
Adrian
cdel...@juno.com schrieb:
Adrian,
Most used High stability tubes are DOA but you might get lucky!
You say the low frequency test worked.
What was the amplitude of the peak in nanoamps and was the peak well
above the noise or baseline level?
Once you were
, the control voltage shows
only a small jump and goes slowly back to the initial value. That's not
exactly what I expect from an integrator.
Adrian
cdel...@juno.com schrieb:
Adrian,
You say it does not lock.
When you put in operate and push the logic reset button does the green
light come
, the response to fine adjust changes was very slow.
So, at least the basics are there and I should be able to get it running
if I find the correct OPamp.
Adrian
cdel...@juno.com schrieb:
Adrian,
You say it does not lock.
When you put in operate and push the logic reset button does the green
light
do know
how to operate the FSSM100 manually, but there is an interface connector
on the rear that might be useful.
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
Hi Bruce,
no. Same color code here.
However, certain carbon composition resistors from the 60's/70's are
notoriously unreliable. The common effect is drift to significantly
higher values. Besides that, they can get pretty noisy.
Adrian
brucekar...@aol.com schrieb:
While tracing out a PC
-publications.html
Best regards,
Adrian
Wolfgang Wallner schrieb:
Hello Time-Nuts community,
I'm interested in the simulation of oscillator noise (especially in discrete
event simulators).
I came across this topic as part of the literature research for my master's
thesis, and have to admit
and DDS together with a HP 53131A in TI
mode with TimeLab.
There are a couple of questions that are probably easy to answer by
someone who has built that system.
Any hints would be much appreciated.
Adrian
___
time-nuts mailing list -- time-nuts
?
Adrian
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
1 - 100 of 327 matches