https://www.febo.com/pipermail/time-nuts/2017-March/104374.html
I posted another HP cesium beam tube patent at
http://gonascent.com/papers/hp/hp5061/US3387130.pdf . It covers the
formation of the cesium beam and should be of interest. We now have
over a month of operation on our first HP5061B
https://www.febo.com/pipermail/time-nuts/2017-March/104374.html
The patent that I posted at
http://gonascent.com/papers/hp/US3323008.pdf has many interesting
facts about the cesium beam tube. It says that the temperature of the
cesium oven is 65° C. I researched the vapor pressure of cesium at
Things to be careful about!
"I think the final vacuum improvement can be achieved more
quickly if the cesium oven is on."
The outgassing in a tube that has been off for an extended time
is almost all from the oven filaments and the ionizer filament.
The metal surfaces are virtual "sponges" and
I am listening and learning. Great point on the HV supplies.
When I look at the supplies I believe they are soldered cans. Is that
correct?
Further had not thought about the ionizer. Skips pictures of the CBT tube
clearly shows a popped ionizer.
There are two oven controllers AC or the DC
John makes a good point about the ionizer filament has anyone done a 'slow
start' system for the Ionizer filament?
I.e. Limit the inrush current as is done for expensive high power transmitting
tubes?
> On Mar 21, 2017, at 9:44 PM, Donald E. Pauly wrote:
>
> It
It looks like that there is about 10% hysteresis on the cesium trip
off/on. That may not be enough to prevent cycling on and off. I may
not have made it clear but instability in the +3,500 voltage makes a
big difference in the threshold ion current required for activation.
If it fades it can
If you cannot get the ion current below 50ua or so after a week at 5kV
then you are out of luck.
Most likely you have resistive deposits on the ion pump insulator.
If you can get the tube to give a decent SN at those levels then you can
run with the alarm circuit bypassed.
I have run tubes a
https://www.febo.com/pipermail/time-nuts/2017-March/104374.html
I could not figure out why the A15 Ion Current monitor would not let
the cesium oven turn on with ion current under 25 uA. At the time we
had no way of measuring the +3,500 ion pump supply. We bought some
200 Meg resistors and made
trojancow...@gmail.com said:
> We will discuss our findings if there is interest.
This is time-nuts. Of course there is interest.
What fraction of the old/dead tubes are "gassy"?
--
These are my opinions. I hate spam.
___
time-nuts mailing
Donald welcome to the group. If a units been off a long time and it sure
sounds like thats the case it may take quite a while like a month or so for
the unit to remove all of the "Stuff" that has out gassed. So be patient
and let the pump do its job. After it does lower and my fingers are
crossed.
Do a search of the list and a few other sites. there is a LOT of
information on what challenges one can have.
And a lot of the information is practical.
As for as pumping a tube down, it took me pretty much a day + a whole
weekend, but it did go down to what was
scribbled on the door, and now
This is my first post. I just bought a non-working HP5061B on eBay
for $350. My old boss KB7APQ in Salt Lake City and I are working on
it. It came in from the Philipines in October of 2015 and was
diagnosed with a bad physics package in March 2016 by AllTest in New
Jersey. We initially found