Re: [volt-nuts] 3420B chopper drive problem

2017-11-14 Thread David C. Partridge
November 2017 14:08 To: 'Discussion of precise voltage measurement' Subject: Re: [volt-nuts] 3420B chopper drive problem Interestingly the PCB is marked to suggest the schematic is correct! And while the connections to the neons have been reworked, it would appear the diodes never have. Reversing

Re: [volt-nuts] 3420B chopper drive problem

2017-11-07 Thread David C. Partridge
: [volt-nuts] 3420B chopper drive problem Hello Dave, in the schematic, the polarity of the diodes in series with the neon bulbs is definitely wrong. If you compare with the 3420A on page 7-7 (29), but also with the Fluke 845 which is similar, both neons have to be switched at 180°. Maybe someone

[volt-nuts] 3420B chopper drive problem

2017-11-07 Thread Dr. Frank
Another probable error in the manual, on page 5-8 (80), diagram, which shows the monitoring of the neon currents: That diagram is ok for the A-version, as both neon currents are measured with the same polarity in its oscillator circuit. The B-version anyhow should look different, as the

[volt-nuts] 3420B chopper drive problem

2017-11-07 Thread Dr. Frank
Hello Dave, in the schematic, the polarity of the diodes in series with the neon bulbs is definitely wrong. If you compare with the 3420A on page 7-7 (29), but also with the Fluke 845 which is similar, both neons have to be switched at 180°. Maybe someone else has erroneously reversed one

[volt-nuts] 3420B chopper drive problem

2017-11-07 Thread David C. Partridge
I'm trying to resurrect a 3420B. I'm looking at the waveform at TP1 on the A2 board. This is supposed to be a square(ish) wave, but within a 6mS period, there should be two cycles of the signal, one with an amplitude of (say) 150mV, and the next cycle only 100mV. I think that the larger