Re: [volt-nuts] AC calibration

2014-08-27 Thread acbern
what nist means is that a precision meter is not considered a standard. you always measure against a true standard (732a, esi sr104...). nist does not mean that as part of doing equipment calibration a 3458a cannot be used as aid. also keep in mind nist has a different approach than a cal lab,

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Randy Evans
I am doing multiple 100 measurements simply to characterize the stability of the 3458A and 732A units I just bought. After about 10 measurement sets over 2 days I am seeing a variance of about .5 uV for the 10V output, or 0.05 ppm. However, the mean varies over a range of 10 uV, or 1 ppm. Does

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Randy Evans
I forgot to mention that I reduced the number of measurements to 100 per set since I wasn't seeing much difference in the variance between 100 and 1000 measurements and the 1000 measurement per set takes too long. Randy On Wed, Aug 27, 2014 at 5:13 AM, Randy Evans randyevans2...@gmail.com

Re: [volt-nuts] AC calibration

2014-08-27 Thread Steve - Home
Fred, The 3458A is 8.5 digits, which puts it into the standards category. They are used as lab standards in many, many labs. 73, Steve WB0DBS On Aug 27, 2014, at 12:08 AM, pa4...@gmail.com wrote: Thanks, I did not new NIST has so much interesting information on their site. I

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread acbern
well, what you do is to measure the stability of the 3458a. 1ppm drift though sounds much as overall short term average variation if your temperature is stable and your 3458a is always on. probably the temp is not stable, and that is what you see, amongst potentially some other smaller drifts

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Bill Gold
Randy: Something I had never tried to measure. As I have found out in the past there is a lot of overhead going on in the meter during and after a measurement. In thinking about this I turned OFF the autozero AZERO and the time for each SMPL was cut in half to around the estimated 16.66

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Don@True-Cal
I have included a link to an ~30-hour measurement run that I consistently do that should give you some idea of expected measurement drift over time. Virtually all of the measurement drift is due to the 3458A internal temperature differences around the 7V reference caused by ambient temperature

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Randy Evans
The HP3458A and the Fluke 732A are on continuously and I do an ACAL at least every few hours, or when the room temperature changes by more than 1 degree C. The total range of measurements is 10uV so the drift is +/-5uV, or 0.5 ppm. The room temperature is not particularly stable and varies over a

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Randy Evans
Don, If I am reading your charts correctly, it looks like your system is cycling over an 7 uV range over several days in summer, similar to what I am seeing. I don't have air conditioning (Northern California) so I do get some pretty good temperature variations from day to night, but less inside

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Don@True-Cal
Randy, Sorry for the two graphs being at different scales so just be sure to readjust your reference. I like to think in PPM terms so the first graph is +/- 5 PPM for the whole gray plot area while the second is +/- 1 PPM. The most extreme outliers on the first graph is +0.3 to -0.5 PPM so

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread acbern
hi randy, the specified drift of the 3458a over your 38.1 to 40.3 (about 1k) is 1ppm +/- allone in the 10v range. thats 10uv. in other ranges its worse. unless your 732a is very bad (very unlikely), you measure mostly the 3458a temp. drift. 1000nplc and 100 readings average do not make sense in

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread M K
On 26/08/2014 16:05, Mike S wrote: After some more research, I think I've answered some of my own questions - Tellurium copper is used for binding posts, not because it has any special thermal or EMF mojo, but because it machines much better than pure copper. And, I suppose, because it

Re: [volt-nuts] 732A and Prologix received

2014-08-27 Thread Stan Katz
As a non-scholar in metrology, I tend to want to simplify the results of an academic debate to make the results of the debate useful to me. One thing is clear from my web search, copper alloyed with Tellurium, or Beryllium, still oxidizes, only at a slower rate. It appears that a big disadvantage

[volt-nuts] fluke 5200a extender board.

2014-08-27 Thread Ken Goodhew1
Hi all, Over the last 6 months I have been steadily repairing a faulty 5200A that I brought, it had numerous faults in the power supply regulator and power amplifier boards that I have managed to repair with a lot of difficulty as I do not have an extender board to operate the

[volt-nuts] Using a DVM to adjust a 752A divider

2014-08-27 Thread Mitch Van Ochten
Folks, Had some correspondence with Fluke a while back regarding adjustment of the 752A Reference Divider. They no longer sell the 845AB Null Detector but they have some suggestions on alternate meters to do the job. They also make reference to zeroing the inputs of the detector, which has