Manichandra

in tghe seisan manual you find a detailed description on how to use poles and zeros. In the documentation you sent there is also a constat. if you generate the responsee using the constat and the poles and zereos, you should get a response curve similar to the one shown in the figure . i suspect that the sesmometer gain of 1500 v/m/s is not included so the gain should be 1. in addtion you have o add the gain of the digitizer which you find somewhere in the documentation, it depends on the actual setup of you digitizer. you can disregard the digitizer filters. The digitizer gain should be in the range 100 000 to 1 000 000 counts per volt.

alterantively you can make a simplified response assuming the sts2 to be a standard seismometer, using the following:


free period 120 s

damping 0.7

generator constant 1500 v/m/s

gain of digitizer, somtehing like 200 000 counts/V, check you spec sheet

jens

On Sat, 12 Feb 2011, Manichandra Sanoujam wrote:

Dear Sir,
Recently we have been using Q330 with STS-2 for our observatories located in
the Subduction zone of Indo-Myanmar region. We have some difficulties in
preparing the Seisan response files for these system. I am attaching the
information given by Q330 group to me related with the response file. They
have sent the poles-zeros information to me however I could not get the
value Ao. The STS-2 that we are using is having 9 poles and 4 zeros.

Please help me in this regard.

With warm regards,

Yours sincerely
Dr. Manichandra Sanoujam
Seismological Observatory
Manipur University, Imphal
INDIA