i had a closer look into the calibration of EMG and ECG signals to derive a formula that converts ECG output to measurements in mV. Here is what i found:

ECG:
Offset (optimal): 2048
Gain factor (data sheet): 175
ADC range: 3.0V (fullscale, -1.5V to +1.5V)
Measurement_in_mV = [ ( ADC range / 4095 ) * ( Measurement_from_ADC - Offset) ] / Gain factor

EMG:
Offset (optimal): 2048
Gain factor (data sheet): 682
ADC range: 3.0V (fullscale, -1.5V to +1.5V)
Measurement_in_mV = [ ( ADC range / 4095 ) * ( Measurement_from_ADC - Offset) ] / Gain factor

As shown in the Youtube videos, offset and gain can be corrected accordingly. Can you please confirm that i am correct or provide any explanation if i am wrong!