Thanks Tom/Frank.
 
I eventually found some information that explained it a little more, that's 
when I had the 'light bulb moment' and understood that a buffer of 7.01 
referred to the testing of a known or preferred pH level one may be aiming for. 
 Until I found that I thought it would be alright to use when the readings 
never go under 7.0.  I only need to range from 7.0 up to 10.0, or praps higher. 
 I had a sample tested shortly after production a while ago and it went >9.0, 
so I guess their equipment was calibrated to only go up to 9.0.  I'm curious to 
find out how high that pH level would be *immediately* after 
production...that's why I was looking into getting some sort of metering 
device.  I don't really want to use those litmus strip thingo's, they'd be a 
tad too 'iffy' for my liking.
 
Pity a refractometer wouldn't work, but it appears they only read glycol and 
Specific Gravity.  Pity a simple hydrometer couldn't be invented for this stuff 
too, measuring the weight or Specific Gravity of silver in water, now *there's* 
something that would be very handy, between a TDS/PWT or EC/TEMP meter and a 
hydrometer one would pretty much have the best of everything.
 
P.S.  I just had a quick google and it appears hydrometers are used to measure 
silver in photography solutions...will go and look a little closer into that.

N.
 




From: [email protected]
To: [email protected]
Subject: RE: CS>pH calibration fluid?
Date: Sun, 20 Feb 2011 09:30:01 -0800






Hello Neville,
 
When using single point calibration you generally use a calibration solution 
that is close to what you are trying to measure.  If your solutions are close 
to a PH of 7, the 7.01 calibration solution will work well.
 
If you are measuring over a wide range of PH, you would typically do a triple 
point calibration using calibration solutions of PH 4, PH 7 and PH 10.  
 
One nice thing about the triple point calibration is that it alerts you when 
your PH probe is wearing out.
 
Tom
 
 


From: Neville Munn [mailto:[email protected]] 
Sent: Saturday, February 19, 2011 4:57 PM
To: [email protected]
Subject: CS>pH calibration fluid?
 
Can someone help me out here re: specifically a *single point* calibration?
 
If one does not need to calibrate below base 7 or below in the acidic range, is 
a buffer solution of 7.01 adequate?
 
I know one can calibrate to determine ranges between two desired figures, but 
using two buffer solutions would only be required if measuring both adicity 
*and* alkalinity would it not? 
 
If measuring *above* 7.0 only, or in the alkaline range, I would assume two 
calibration points would not be necessary as the reading would/should never go 
below 7 in this instance...Yes/No?
 
Thanks
 