Yes, you'll see something like: *ITMS + p NSI E full scan (400-1800)* in the
scan filter/header information on the file.

Does this help?

-jane


On Fri, Sep 18, 2009 at 4:30 PM, Matthew Chambers <
[email protected]> wrote:

>
> Hi Jane,
>
> Thanks for answering so quickly. This filter line was found on an LTQ XL
> RAW file. Does the "enhanced scan speed" you mention result in the 'E'
> in the filter line?
>
> Thanks,
> -Matt
>
>
> Science Gurl wrote:
> > If you mean "enhanced" scan speed and you're using an ion trap (linear
> > or LCQ), I'd highly recommend choosing it.  The scan speed is slightly
> > slower than the normal speed and allows for better mass accuracy and a
> > better chance at seeing the charge envelop (thus better able to
> > determine charge state).  I have an LXQ and find that when I use
> > "enhanced" I get mass accuracy approaching +/- 0.2 - 0.5 amu vs unit
> > resolution for regular.  Of course this only applies to profile data,
> > I don't know if it's even an option in centroid collection mode.  I
> > also use it for my ms/ms.
> >
> > -jane
> >
> > On Fri, Sep 18, 2009 at 4:00 PM, Matthew Chambers
> > <[email protected]
> > <mailto:[email protected]>> wrote:
> >
> >
> >     Does anybody know what this means? How does it affect the data and/or
> >     metadata?
> >
> >     Thanks,
> >     Matt
> >
> >
> >
>
> >
>

--~--~---------~--~----~------------~-------~--~----~
You received this message because you are subscribed to the Google Groups 