The HP 5328A that I have uses the EFC as a fine manual adjust. There is an
accessory card that accepts the oscillator and the supporting fine
manual-adjust circuitry is located there.

Bill, K8CU

----- Original Message ----- 
From: "Rick Karlquist" <[EMAIL PROTECTED]>
To: "Discussion of precise time and frequency measurement"
<[email protected]>
Cc: <[email protected]>
Sent: Friday, December 29, 2006 8:04 PM
Subject: Re: [time-nuts] HP 5345A


> Colin Bradley wrote:
> > I just picked up a 5345A from eBay. This unit has a 10811 oscillator
> > instead of the 10544 depicted in the manuals I have. The serial number
has
> > been removed. Does anyone know if there is an EFC adjustment for this
> > counter or is the course control the only way of setting frequency?
> > Colin
>
> I don't know of any instrument in the history of HP/Agilent that ever
> used the EFC as a fine manual frequency adjustment.  The EFC
> was only ever used as a phase locking interface.  The tuning adjustment
> is adequate for annual calibration purposes (IE, you can set it
> within 1 Hz, or 1E-7 if you are careful, and the expected annual
> aging is in that ballpark).
>
> Rick Karlquist N6RK
>
>
> _______________________________________________
> time-nuts mailing list
> [email protected]
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


_______________________________________________
time-nuts mailing list
[email protected]
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Reply via email to