HP is NOT perfect.

I have a 5519A LaserInterferometer head which was acting up. They are
distinctly non-trivial to fix. Anyway, this one had a critical feedback
detector that had NEVER been soldered to the PCB.

Moral:  Look for the simple faults first.

-John

=============


> I was recently having trouble with one of the 54002A input pods on my
> 5372A.  It was occasionally being recognized as a 54001A pod.  I took it
> apart and found that 7 of the 8 solder joints between the circuit board
> and the connector at the back of the pod were bad.  I checked my other
> pods and they all had at least one bad connection..  All the pods
> worked, but gave weird results when making low-level measurements.
> After I resoldered them the strange readings disappeared.  Imagine
> that!  :-)
>
> Ed