Hi

On all these counters, you need to switch to DC input for 1 pps signals. You 
also need to manualy set the trigger level to something like +1.5V. Next thing 
to fiddle is the 1 meg / 50 ohm termination. Sometimes it works better with 50 
ohms, sometimes not.

Bob

On Feb 16, 2014, at 2:03 PM, Jimmy Burrell <[email protected]> wrote:

> I've recently purchased an HP 5335A and for the most part, we're getting 
> along famously. Then I tried to measure my GPS' 1 PPS pulse. I'm not seeing 
> anything Iike what I'd expect to see, e.g. 0.000001. Doesn't matter what type 
> of measurement I select (freq, period) the measurements are nowhere near "1". 
> In fact they are all over the place and changing a lot faster than once per 
> second.
> 
> My first thought was "not enough signal" present at the input of the counter 
> causing a low SNR. But my scope says it's better than 2v on p-p. According to 
> the HP5335A manual, that should make the counter happy.
> 
> I've never tried measuring any freq. this slow so I'm out of my element. 
> Probably missing something obvious.
> 
> Any of this ringing a bell with anyone? Any clues or suggestions?
> 
> Many thanks,
> 
> Jimmy...
> N5SSPE
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.