TIC: SR620 with Z3805 as external reference; signal source
Nortel/Trimble GPSTM (GPSDO) 10MHz output

Enclosed two plots (SW: "Plotter"):
- one is sigma(tau) calculated from phase samples (SR620 TIME mode),
- the other one is sigma(tau) from frequency data (SR620 FREQ mode)

Whole equipment had a power up time of several days/weeks. Room
temperature was stable over both measurements (within about 2 degrees C).

The SR620 uses a bit different path through the logic when doing TI and FREQ measurements. The frequency measurement has a "feature" that means that the time error between start and stop signal needs to be calibrated out. This can be done using the calibration routines given in the manual. This should not affect the ADEV measure, but as a precaution.

Try doing a pair of noise-floor measurements. That is, feed the reference 10 MHz to the A input for the frequency noise measurement. Then, for the TI noise-floor measurement, put a T on the A input, put it in high-Z mode and then use a 1 m cable to put the signal onto the B input which is terminating.

You indeed have a higher level. Your initial shape makes me wonder. I would really like to get the TimeLab measurement files and eye-ball them closer.

If you plot the phase or frequency, it may be easier to spot systematic wobbles. TDEV would also help, as it provides a general *tau scaling to the ADEV plot.