Another update on progress:
--------------------------
Now at the stage where two of the three 3210's are working. The fault that was blowing the line fuse turned out to be a shorted reservoir cap. Managed to find an Ebay seller in Greece with exactly the same voltage, capacity and outline, though different manufacturer. The caps have a stud mount at the base of the can, so not easy to find, but bought four at just over 3 ukp each and replaced both caps in the psu, leaving two for spares.

From cold, the spec is up to 90 minutes warmup time, but if the OCXO and pump have been running for some time, both units lock up within 4 minutes. Have done some setup. For example, normalised the gains and offsets in the preamps as per the manual and both units show the expected preamp level of ~165mV and 2nd harmonic amplitude of 9-10 on the meter, so assume both tubes are in good condition. None of the settings were very far out.

Took a set of measurements from one of the good units and the faulty for comparison, with similar results, Good unit as follows, but faulty unit more or less the same:

Ion pump = EHT ok, meter = 0
Electron multiplier EHT voltage = 1800 volts, (100Mohm hv probe on o/p wire)
Cesium oven = Not open circuit, voltage = 5.5 volts
Ioniser = not open circuit, 1 volt p-p square wave at ~26KHz
Synthesiser output = 12.6317715, follows variation in OCXO and in lock
Backplane test points, tpb, tpc = 8.2 volts p-p square wave, ~137 Hz.
Multiplier 12.6... MHz input = 2v p-p (Scope)
Multiplier 137Hz input = 2.25v p-p (Scope)
Multiplier 180MHz output = +26DBm (HP 3406A rf voltmeter + 20DB attenuator)
Microwave tap on waveguide = -14.5DBm, threaded attenuator works (HP432A, 478A)

The 3406 and 432 are quite old and the sensor head is even older, but should be reasonably accurate. The 478A head was coupled to the microwave tap using a ~3 inch length of rigid coax, so not much attenuation. Hopefully, these figures may help others trying to debug these standards, but if there's anything i've missed, please let me know.

For the faulty unit, even with all the levels as expected, there is still no signal, nor 2nd harmonic, even with the preamp gain turned up to max. Connected a Fluke electrometer in series with the preamp input socket, but the current is at least 100 times down on 1nA. Looks like the tube is completely lifed, or has an internal fault. /FX: Have visions of a sad, faithful and slowly dying 3210 left for decades in a rack, in the dark, powered up, but long forgotten after the accurate reference it provided ceased to be used anywhere in the organisation. A ship adrift for thousands of years, but all systems still at least partially working and waiting to be discovered :-).

So what else am I missing in this puzzle ?...

Regards,

Chris



_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to