Hi,

yes, it is my fault - my set up (connections)  for TI mode was not correct. 

According your advices I went back to the frequency mode and set gate / sample 
time to 1 second for CNT-91 and TimeLab.
Then I switched on "Smart Frequency" and "Interpolator Calibration" option on 
CNT-91

        1. Setting -> Misc -> Smart Meas -> Smart Frequency -> ON 
        Smart Frequency (valid only if the selected measurement function is 
Frequency or Period Average). 
        By means of continuous timestamping and regression analysis, the 
resolution is increased for measuring times between 0.2 s and 100 s.

        2. Setting -> Misc -> Smart Meas -> Interp Calib ON
         Interpolatator Calibration - By switching off the interpolator 
calibration, you can increase the measurement speed at the expense of accuracy.

Please see the result  with and without these CNT-91's options (Smart Frequency 
and Interpolator Calibration)
https://dl.dropboxusercontent.com/u/21338179/hamradio/OCXO_Adev/morion%20ocxo%20adev%20freq%20mode%20smartfreq%20sample_1s%20interpcalib_on.jpg
 . 

This result is for the internal CNT-91 reference source. I will check with 
external reference second Morion OCXO shortly.  

Thanks,
Karen