Did you use the CFG-TMODE svinVarLimit to set 350 and the svinMinDur
to set 32400?
The CFG-TMODE svinVarLimit is a variance in squared mm, not a standard
deviation...

On Fri, Apr 15, 2016 at 6:27 AM, Bob Stewart <[email protected]> wrote:
> I hope I haven't posted about this before, but it's something that's bugged 
> me for awhile.  When I run a survey-in on an LEA-6T, the 
> CFG-TMODE2->fixedPosAcc value seems to be wrong.  For instance, I ran a 
> survey today and told it to run for 32,400 seconds and reach at least 350mm 
> accuracy.  When it finished, TIM-SVIM reported it had run 38,126 seconds with 
> a mean 3D std deviation of 350mm.  OK, so far.  So, I go look in the 
> CFG-TMODE2 structure and it reports 122mm.  That seems a bit odd, but OK.  
> And then I squared 350, and came up with 122500.  It appears that instead of 
> just moving the 350mm value over to the TMODE structure, they've squared it 
> and divided it by 1000.  I've noticed this happening on at least 2 versions 
> of the 6T.  I suppose I should do the same test on the M8T.
>
> Bob
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.