Hi

Ok, You have a thermocouple junction at the + post on the DVM. You have one at 
the - post on the DVM.
You have a junction at the + connection to the board. You have a junction at 
the - connection to the board. 
There are indeed more than that, but those four are pretty much a sure thing. 
The ones near the hot OCXO
are also worth looking at. 

The metals involved are unknown, so we have to guess a bit. You *can* get some 
alarming junctions with 
very normal test lead materials. Getting low thermal EMF connections requires 
special attention. 
If your junctions have a 1 mV / K coefficient, then you need temperature data 
that is good to 0.001C at each
junction to work out what is going on at the junctions. That assumes you are 
after 1 uV on your “data” plot. 
It’s not terribly hard to get a contact at one (or more than one) junction that 
messes you up.

Bob

> On Nov 7, 2016, at 12:20 PM, Bob Stewart <[email protected]> wrote:
> 
> Hi guys,
> First of all, thanks for the additional responses.  I was a bit angry and 
> rude yesterday, and I figured this thread was over.  Thanks for staying with 
> me.
> I haven't had time to look over the data etc in your responses.  I'll do that 
> and get back to the list if appropriate.
> 
> I spoke to Attila and Azelio offline last nite and from their input, I 
> decided to hook up the 3456A and collect some data, which is in the plot 
> attached.  As usual, I've modified one of my standard plot scripts, so there 
> is some extraneous data that wasn't removed.
> 
> First for our purposes is the thin red line, which is the DAC value locked at 
> 0x734B0.  The orange trace is the temperature adjusted so that each step of 
> 10 on the right hand Y tics is one degree F.  The dark blue trace is the EFC 
> value read by the 3456A.  It has been multiplied by 100,000 and then had 
> 282600 subtracted.  This leaves just the LSD scaled at 1:1 on the right hand 
> Y tics.
> There's something interesting on the far right hand side where the 
> temperature goes low and stays there.  The DVM value follows it down, but 
> then recovers while the temperature stays down.  I'm not sure what to make of 
> this.  Either then OCXO is making up for the temperature change by increasing 
> the temperature, or the 3456 is compensating for it after the fact.  In 
> either case, the EFC seems to only follow the transient temperature changes, 
> and doesn't actually track the temperature on the board.
> 
> So, to my eye, after 14 hours, there is only a dependency on thermal 
> transients.  I'll leave it running for some time yet, but the EFC doesn't 
> seem to be drifting in any meaningful way at this point, other than in 
> relation to temperature changes.
> 
> Note:  I used a shielded twisted-pair with the usual clips attached to ground 
> and EFC in my GPSDO.  On the 3456, the two leads go to the appropriate volts 
> inputs, and the shield goes to the ground input.  The "guard" switch is out, 
> which is the off position.  There is no shield connector on the DUT side.
> 
> Bob
> 
>  -----------------------------------------------------------------
> AE6RV.com
> 
> GFS GPSDO list:
> groups.yahoo.com/neo/groups/GFS-GPSDOs/info
> 
> 
> 
> <PLL.png>_______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to