Hi

The change is suspiciously close to the electrical tuning range of a typical HP 
OCXO. 
The answer may be a failure of the bias on the EFC line …..

Bob

> On Nov 2, 2018, at 2:51 PM, Walter Shawlee 2 <walt...@sphere.bc.ca> wrote:
> 
> I have several of these as the -010 high stability timebase options
> in my various HP counters, and generally they work very well, with
> usual errors under 0.01Hz and very minor drift over time.
> 
> a few months ago, I had a power interruption, and black out for a few hours,
> along with the usual erratic restart from the power company. shortly 
> afterwards, I built a nice little
> homemade TM500 plug-in OCXO unit to fit on my bench, and without giving it 
> much thought, used
> my bench standard (an hp 435B-K26 power/frequency reference with an internal 
> 10544A) to cal it.  all seemed good.
> 
> soon after, I was working on an hp 5334A counter, and added a 10811A as as 
> upgrade from my
> spares and suddenly, I had a big 1.3Hz error at 10Mhz when I cross checked it 
> to my bench references AND a rubidium. I brought over my recently cal'd 
> rubidium from the upstairs lab, and yes, there was now clearly a big step 
> error. my upstairs 5335A with a 10811A had the same step effect!
> 
> these 2 units are always on for stability, so both got cycled the same way 
> during the power failure.
> 
> it seems that the power failure cycle had bumped the internal 10544A 
> oscillator inside the 435 by that amount, as well as the 10811A inside the 
> 5335A. I have never seen that effect before, both the ovenized osicllators 
> from hp have been very reliable for me, so I thought I would put that info 
> out in case anyone else has seen this effect and knows the cause.
> 
> using the rubidium (which I keep off until I need it, and wait for at least 2 
> hours for best settling), I reset everything back to a flat 10Mhz, and all 
> was well, except that the first 10811A I put in the 5334A conked out (oven 
> still fine, but the oscillator went dead, giving the dreaded "no osc" message 
> on the counter). another spare fixed that, and two days of drift testing to 
> get everything back where it belongs.  anybody want the bad 10811A?
> 
> anyway, just thought the information might be handy for others.  the EFC 
> range on the 10811A/10544A is *only 1Hz*, so such a big jump is unusual to 
> say the least. it required the main coarse adjustment to fix.
> The 435B-K26 is a pretty remarkable widget if you ever see one, it makes a 
> great 10Mhz reference and 1mW power reference in one little box, very useful 
> for an RF bench. One of hp's rare and forgotten treasures.
> 
> all the best,
> walter
> 
> -- 
> Walter Shawlee 2
> Sphere Research Corp. 3394 Sunnyside Rd.
> West Kelowna, BC, V1Z 2V4 CANADA
> Phone: +1 (250-769-1834 -:- http://www.sphere.bc.ca
> We're all in one boat, no matter how it looks to you. (WS2)
> All you need is love. (John Lennon)
> But, that doesn't mean other things don't come in handy. (WS2)
> 
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to 
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.


_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Reply via email to