I did a test of the FA2 internal OCXO oscillator.  The FA2 uses this as the 
reference if you don't connect an external reference.  You can adjust the 
oscillator frequency with a small pot on the back panel.  As shipped the 
oscillator was 0.10 Hz off.  I adjusted to be within 0.003 Hz (I didn't try to 
fiddle it to be spot on).

I did the oscillator test by having it measure an HP-5071A cesium oscillator 
(10 MHz) which is assumed to be (and is) spot on and very high quality.  Any 
deviations from the noise floor test that I did earlier are mostly due to the 
FA2 internal oscillator.

The wobbles in the frequency plot are due to air conditioning.  The oscillator 
changes freq by around 0.00025 Hz / degree C.  The oscillator drifted at the 
rate of 0.00635 Hz / day, but the drift rate is slowing down (it's now around 
0.005 Hz / day).   It will be interesting to see how it ages in over the next 
few weeks.