Luciano wrote:

    The CLC409 datasheet show all the test response up to +12dBm (around 1Vpp 
on 75/50Ohm load)
    see: pdf.datasheetcatalog.com/datasheet/nationalsemiconductor/DS012748.PDF

I'm reporting what I have observed with the CLC409 (and LMH6702) when operated on the "absolute max" supply rails of +/- 6v. Of course, I would never design a commercial product using abs max specs without having a serious conversation and understanding with the manufacturer (if then).

Clipping commences at just over +/- 2v (a touch over +16dBm). This is with a 50 ohm build-out or "back termination" resistor -- you can get an additional 6dBm if you do without the build-out resistor and terminate only the load end of the cable in 50 ohms, which actually works fine as long as the output loads are matched at 50 ohms. I terminate both ends "just in case" they aren't. Harmonic distortion does not rise until there is visible clipping.

    I agree that harmonic distortion is an important factor to keep in mind 
however -45dBc is a very interesting value considering the market alternatives. 
In my document I have enclosed on page 10 a little research I did online on 
similar products and as you can see the -45 are positioned at a decidedly 
higher level than other commercial solutions
    see [table]:

IMO, pretty much all commercial distribution/isolation amplifiers fail miserably to hold distortion down to tolerable levels -- one reason why I have always designed my own. Interestingly, some older, vacuum-tube models did better.

    An extremely important factor not to be overlooked is the maximum output 
power / level of the amplifier. Measurement systems such as the HP3048A for 
phase noise measurement use mixers that require minimum signal levels of + 
15dBm at higher PO1 and the CLC409 will never meet these requirements.

As noted above, it will if it is operated on abs max supplies, even with a 50 ohm build-out resistor. Without the build-out resistor, it will supply +22dBm (+20dBm on +/- 5v supplies).

    In your modification document: Residual phase noise of <-165dBc at 10 MHz
    I suppose you mean @100kHz, The buffer I propose is -170dBc @100KHz

I normally cite 10kHz measurements, but I don't recall if I cited 100kHz or 10kHz measurements (my bad for the incomplete spec). One or the other.

    In the end I don't think this should be a competition among Distribution 
Amplifiers.

Nor I. I built a few dozen of the modified Extron DAs, which I've passed on to other time nuts over the years. I still have one around, but it is not in daily use. I currently use two different designs, both scratch-built, one IC-based and one discrete. As I've said before, I distinguish between distribution amplifiers (lots of outputs) and isolation amplifiers (where only 2 or maybe 3 outputs are needed, but distortion, residual PN, and transit delay must be truly state-of-the-art for sensitive measurements).

On re-reading my original post, I see the tone could be taken as a bit confrontational. That was certainly not my intent -- I just wanted to defend IC designs against some "received wisdom" (*NOT* from you) that they are inferior. Thank you for giving me the benefit of the doubt and responding as a gentleman.

    For a "ready" solution at low cost like DA the Extron I think is a great 
solution.

Me, too, which is why my first efforts as a time-nuts newbie were made in that direction.

Best regards,

Charles



_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Reply via email to