Time Nuts- This 5061B was working fine until a recent power failure, after which the unit would not come on line (alarm light stayed lit).
Major symptoms now: Beam I ; Second Harmonic; Ion Pump Supply are all 0 on front panel. All other front panel readings are normal. (except for battery, which th unit doe snot have). L4P2 signal is 16.832 MHz 1.7 V p-p (perfect - suggesting that A1 synthesizer is fine) Going through 5-173 A12 Cesium Beam Tube Operational Checks: 5-175 Low Frequency Coil Check. Everything is fine except that tweaking the external oscillator frequence/amplitude has no effect. Beam I is always 0. The text says "If the beam current is not restored, it indicates a problem in the the cesium beam tube power supplies (A11, A15, A18, or A19) or in the cesium beam tube itself. A11 = Cesium Oven Controller - front panel reading is fine. Removed A11 from the chassis and made the tests in 5-141. Everything fine. +- 16V and -10V all fine; Cs oven voltage fine; the thermistor tests out at 435 ohm vs. the 430 ohm on the label. A15 = Power Regulator Board - main +18.65V is fine A18 = +3.5KV supply - measured and is fine A19 = -2.5KV supply - measured and is fine I am inclined to conclude that the tube is dead. I don't understand why that occurred suddenly in conjunction with a line power failure. Can I have any hope that the tube is still o.k.? Any ideas? Thanks! Jim _______________________________________________ time-nuts mailing list -- [email protected] To unsubscribe, go to http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com and follow the instructions there.