Bert,

I missed it because I do not have one or worked with one, which is not
to say it is bad or anything, I just took the examples I recall because
it is familiar to me. There is more of these for sure. I hope it was
good enough to illustrate the points with some real-life examples.

Cheers,
Magnus

On 2020-02-21 13:45, ew via time-nuts wrote:
> You missed my favorite  HP5345A  only direct 500 MHz and internal 500 MHz, 
> only recently replaced it with 53132A still use it for 40 GHz work
> Bert Kehren
>
>
>
> In a message dated 2/21/2020 7:27:32 AM Eastern Standard Time, 
> mag...@rubidium.se writes:
>
> Hi Taka,
>
> On 2020-02-21 04:45, Taka Kamiya via time-nuts wrote:
>> I was in electronics in big ways in 70s.  Then had a long break and came 
>> back to it in last few years.  Back then, if I wanted 1s resolution, the 
>> gate time had to be 1s.  So measuring ns and ps was pretty much impossible.  
>> As I understand it, HP53132A (my main counter) takes thousands of samples (I 
>> assume t samples) to arrive at most likely real frequency.  That was 
>> something I had hard time wrapping my head around. 
> It actually does two things.
>
> First, it interpolates the occurrence of a rising edge (for start and
> stop channel), so if this does not happen in perfect alignment with a
> rising edge of the reference/coarse clock. Often the OCXO/Rubidium is
> for 10 MHz,  but then a 90-500 MHz oscillator is locked to the
> reference, and this higher clock is then used instead of the 10 MHz for
> coarse-counting. Coarse-counting is counting of cycles just back in the
> good old days of counters. The resolution is increased further not by
> raising the counting frequency, but by measuring the time-error of the
> trigger channel event in relation to the coarse-counter clock-edge.
> Thus, measuring 0.000-0.999 of a coarse-counting cycle. In practice it
> becomes hard to design for that, as the shorter end has problem is gate
> delay times to be well decided, so one add one or two coarse cycles to
> do 1.000-1.999 or 2.000-2.999 cycles, but these extra cycles is only for
> the interpolator design, so once the fractional cycle is known the other
> can be ignored.
>
> Just to give you an idea of what different counters do, here is from the
> top of my head some numbers:
>
> HP5370A: Ref 10 MHz, Coarse 200 MHz, Interpolation gain 256, time
> resolution < 20 ps
> HP5328A: Ref 10 MHz, Coarse 10 MHz, Interpolation gain 1, time
> resolution 100 ns
> HP5328A with Option 040-042 and HP5328B: Ref 10 MHz, Coarse 100 MHZ,
> Interpolation gain 1 (TI-average has other interpolation means), time
> resolution 10 ns or for TI-avg 10 ps (claimed)
> HP5335A: Ref 10 MHz, Coarse 10 MHz, Interpolation gain 200, time
> resolution 1 ns
> HP5372A: Ref 10 MHz, Coarse 500 MHz, Interpolation gain 10, time
> resolution 200 ps
> HP53132A: Ref 10 MHz, Coarse 100 MHz, Interpolation gain 1000, time
> resolution 100 ps
> SR620: Ref 10 MHz, Coarse 90 MHz, Interpolation gain 512?, time
> resolution < 25 ps (don't recall details)
> PM6863: Ref 10 MHz, Coarse 500 MHz, Interpolation gain 1, time
> resolution 2 ns
> CNT-90: Ref 10 MHz, Coarse 100 MHz, Interpolation gain 512, time
> resolution 100 ps (claimed)
> CNT-91: Ref 10 MHz, Coarse 100 MHz, Interpolation gain 512, time
> resolutions 50 ps (claimed)
> SIA3000: Ref 100 MHz, Coarse, 100 MHz, Interpolation gain 50000, time
> resolution 200 fs
>
> As I write claimed above, the actual performance can be better, but the
> spec on the sheet did not overstate it more. While all the numbers may
> not be 100% correct, I think they help to illustrate the relationships
> very well. As you calculate the length of the coarse counter period from
> it's frequency, and then divide with the interpolation gain, which is by
> how many steps the period is interpolated, the raw time resolutions pops
> out.
>
> Interpolation methods differs, but typically first an error signal is
> generated and then it is stored into a capacitor which is then measured
> with some slower technique. The 5335A use a very simple technique where
> the discharge of the capacitor is done with a much lower current than
> the charging, so now the discharge time can be measured using the coarse
> clock. This is called pulse-stretching. Today the far most common
> technique is to use an ADC to digitize the voltage.
>
> The 5328 counters have a unique interpolation technique by
> phase-modulating the reference clock with noise, effectively shifting
> the reference transitions around and that way interpolate over time a
> higher resolution. It works better than claimed.
>
> Remember that this single-shot resolution is reduced by the trigger
> jitter as well as unstability of reference oscillator. In practice the
> trigger jitter or resolution dominates as a 1/tau limit as you look at
> the Allan Deviation, to fix that you need to buy a better counter or
> signal condition for better trigger.
>
> The second trick used in 53132 for measuring frequency is averaging. It
> uses an average technique originally from optical frequency measures to
> accumulate data into blocks and then subtract the time-stamps of two
> subsequent blocks. This is the same as average the output of a number of
> overlapping frequency estimations.
>
> This has advantages as white noise is supressed with a steeper slope,
> and the associated deviation is the modified Allan Deviation MDEV.
>
>>   
>>
>> I understand most of what you said, but I've never taken statistics, so I am 
>> guessing on some part.  I can see how adev goes down as tau gets longer.  
>> Basically, averaging is taking place.  But I am still not sure why at some 
>> point, it goes back up.  I understand noise will start to take effect, but 
>> the same noise has been there all along while adev was going down.  Then, 
>> why is this inflection point where sign of slope suddenly changes? 
> OK, so the trouble is that rather than only white noise as classical
> statistics deal with, we have at least 4 noise types, with different
> frequency slopes. As we try to analyze this with standard deviation, the
> standard deviation estimator (RMS estimator) does not converge, is
> simply keeps producing noise even if we add more values. To put that in
> another way, we do not gain more knowledge by doing more measurements.
> The classical white noise is what is called white phase modulation
> noise, we then have flicker phase noise, white frequency noise and
> flicker frequency noise. All these noise-types is to be expected
> according to the David Leeson model, and it is due to those that we need
> to use more advanced statistics as introduced by David Allan.
>
> The White Phase Modulation has a flat frequency response in phase noise
> spectrum, 1/tau in ADEV.
> The Flicker Phase Modulation has 1/sqrt(f) respone in phase noise
> spectrum, 1/tau in ADEV.
> The White Frequency Modulation has 1/f respone in phase noise spectrum,
> 1/sqrt(tau) in ADEV.
> The Flicker Frequency Modulation has 1/sqrt(f^3) response in phase noise
> spectrum, flat in ADEV.
>
> In addition to this, linear frequency drift creates a slope that scales
> with drift and tau, so that is an upper limit. Thermal sensitivity tends
> to lay ontop as well, so does other disturbances.
>
> Depending on details of oscillators and their sensitivity to thermal
> noise, their effective minimum shifts around.
>
>>   
>>
>> Also, to reach adev(tau=10), it takes longer than 10 seconds.  Manual for 
>> TimeLab basically says more samples are taken than just 10, but does not 
>> elaborate further.  Say it takes 50 seconds to get there, and say that's the 
>> lowest point of adev, does that mean it is the best to set gate time to 10 
>> second or 50 second?  (or even, take whatever gate time and repeat the 
>> measurement until accumulated gate time equals tau?
> The Allan Deviation takes a number of estimates to produce values, but
> remember these are stability values for a certain observationtime for
> frequency, not the frequency measure itself.
>
> Cheers,
> Magnus
>
>> --------------------------------------- 
>> (Mr.) Taka Kamiya
>> KB4EMF / ex JF2DKG
>>   
>>
>>     On Thursday, February 20, 2020, 7:54:22 PM EST, Magnus Danielson 
>> <mag...@rubidium.se> wrote:  
>>   
>>   Hi Taka,
>>
>> On 2020-02-20 19:40, Taka Kamiya via time-nuts wrote:
>>> I have a question concerning frequency standard and their Allen deviation.  
>>> (to measure Allen Dev in frequency mode using TimeLab)
>>>
>>> It is commonly said that for shorter tau measurement, I'd need OCXO because 
>>> it's short tau jitter is superior to just about anything else.  Also, it is 
>>> said that for longer tau measurement, I'd need something like Rb or Cs 
>>> which has superior stability over longer term.
>> Seems reasonably correct.
>>> Here's the question part.  A frequency counter that measures DUT basically 
>>> puts out a reading every second during the measurement.  When TimeLab is 
>>> well into 1000s or so, it is still reading every second; it does not change 
>>> the gate time to say, 1000s.
>>> That being the case, why this consensus of what time source to use for what 
>>> tau?
>>> I recall reading on TICC, in time interval mode, anything that's reasonably 
>>> good is good enough.  I'm aware TI mode and Freq mode is entirely 
>>> different, but it is the same in fact that measurement is made for very 
>>> short time span AT A TIME.
>>> I'm still trying to wrap my small head around this.  
>> OK.
>>
>> I can understand that this is confusing. You are not alone being
>> confused about it, so don't worry.
>>
>> As you measure frequency, you "count" a number of cycles over some time,
>> hence the name frequency counter. The number of periods (sometimes
>> called events) over the observation time (also known as time-base or
>> tau) can be used to estimate frequency like this:
>>
>> f = events / time
>>
>> while it is practical that average period time becomes
>>
>> t = time / events
>>
>> In modern counters (that is starting from early 70thies) we can
>> interpolate time to achieve better time-resolution for the integer
>> number of events.
>>
>> This is all nice and dandy, but now consider that the start and stop
>> events is rather represented by time-stamps in some clock x, such that
>> for the measurements we have
>>
>> time = x_stop - x_start
>>
>> This does not really change anything for the measurements, but it helps
>> to bridge over to the measurement of Allan deviation for multiple tau.
>> It turns out that trying to build a standard deviation for the estimated
>> frequency becomes hard, so that is why a more indirect method had to be
>> applied, but the Allan deviation fills the role of the standard
>> deviation for the frequency estimation of two phase-samples being the
>> time-base time tau inbetween. As we now combine the counters noise-floor
>> with that of the reference, the Allan deviation plots provide a slopes
>> of different directions due to different noises. At the lowest point on
>> the curve, is where the least deviation of frequency measurement occurs.
>> Due to the characteristics of a crystal oscillator to that of the
>> rubidium, cesium or hydrogen maser, the lowest point occurs at different
>> taus, and provide different values. Lowest value is better, so there is
>> where I should select the time-base for my frequency measurement. So,
>> this may be at 10 s, 100 s or 1000 s, which means that the frequency
>> measurement should be using start and stop measurements with that
>> distance. OK, fine. So what about TimeLab in all this. Well, as we
>> measure with a TIC we collect a bunch of phase-samples at some base
>> rate, such as 10 Hz or whatever. TimeLab and other tools can then use
>> this to calculate Allan Deviation for a number of different taus simply
>> by using three samples, these being tau in between and algoritmically do
>> that for different taus. One then collects a number of such measurements
>> to form an average, the more, the better confidence interval we can but
>> on the Allan Deviation estimation, but it does not improve our frequency
>> estimation, just our estimation of uncertainty for that frequency
>> estimation for that tau. Once you have that Allan Deviation plot, you
>> can establish the lowest point and then only need two phase samples to
>> estimate frequency.
>>
>> So, the measurement per second thing is more collection of data rather
>> than frequency estimation in itself.
>>
>> Cheers,
>> Magnus
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe, go to 
>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>> and follow the instructions there.
>>   
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@lists.febo.com
>> To unsubscribe, go to 
>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to 
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts@lists.febo.com
> To unsubscribe, go to 
> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com