Mike.
One concern I have with active components as mixer is noise. For an SA I
designed only a passive DB diode mixer had low enough output noise. Would a
PF detector as being an active component, not create more noise as output?
Erik

On Tue, Jul 5, 2022, 18:20 Mike Monett via time-nuts <
time-nuts@lists.febo.com> wrote:

> You stated:
>
> Mike,
> The phase detector is an ADE-1 mixer, the IF output of the mixer goes
> into a loop filter that has a corner frequency of about 0.2Hz to enable
> Phase noise measurements down to 1Hz offset
>
> That is your problem. A double balanced mixer is an exclusive-or phase
> detector. The lock range is determined by the loop bandwidth, as you have
> found.
>
> The phase-frequency detector is completely different. It will lock to any
> signal in the lock range, independent of loop bandwidth. You can have a
> bandwidth of 0.001 Hz, and it will still lock. Think of what this could do
> for your phase measurements.