Just as a note - re-education was done for staff throughout the institution.  
One of the reasons this was discovered was because the pediatric ICUs were 
undergoing an intensive PI related to infection control practices and in every 
other area were these efforts were undertaken, a reduction in infection rates 
was seen.  In these units, it was going up - which didnt' make any sense.  One 
of the Infection Control staff had just attended a conference, where this had 
been presented.  So, the Hopkins staff began to look at the timing of the 
increase in infections and discovered that it coincided with our change to the 
positive pressure valve.  As soon as we took it away, the infection rate 
dropped.   Hopkins is a fairly large facility so the thought is -  that if 
training and retraining does work for the masses, then the device is probably 
not user friendly and not a good choice for this environment.  From the info 
that I have read, (and I also believe Lynn stated this on the list!
  serv recently) that this increase in infection rates has been contributed to 
several different valves - it is not isolated to one valve.

Beverly Reynolds, RN, MBA
Nurse Manager, Vascular Access Team
The Johns Hopkins Hospital
410-614-1895

>>> <[EMAIL PROTECTED]> 01/16/06 11:03 PM >>>
 
 
Well Peter you hit the question exactly. Drs Maki and Jarvis did not  mention 
a specific brand. The word from valve manufacturers seems to be blaming  
weaknesses in the fluid path seal that allow liquid contaminates in. I heard  
practioners talking about poor staff education and/or compliance with swabbing  
as 
a possible cause. 
 
But truly we don't know that there is a problem. There was no data given as  
to type of valve or how it was used or how staff was educated. There was no 
data  about a trailing a different valve to see if there was a difference. and 
no  mention was made as to the reeducation of staff and a trail to follow that. 
 
So I feel we have many more questions to answer before we start tossing our  
valves in the can. But certainly it is interesting and worthy of note. I hope  
that everyone will begin taking a close look at the numbers in their 
facilities.  Perhaps a few quick small trials to see if any trends emerge.
 
Tony
 
In a message dated 1/15/2006 8:14:56 P.M. Central Standard Time,  
[EMAIL PROTECTED] writes:

Thank you Tony for keeping us all up to date with the conference!! A  
question though is burning in my mind. I don't recall, from the disscussion we  
had 
on the list this spring/summer, about a brand  specific needless  connector 
being responsible for for the increase in CRBSI. So could you please  tell us 
which connector(s) are in question? What is the reason that  belies a 
particular 
valve design, that makes it prone to causing  CRBSI? 
Peter Marino RN BSN 
----- Original Message ----- 
From: [EMAIL PROTECTED] (mailto:[EMAIL PROTECTED])  
Date: Sunday, January 15, 2006 12:06 am 
Subject: AVA conference, DAY 3"The topic was needleless  connectors. They 
discussed the qualities of a needleless device that would  contribute to 
maintaining a safe fluid pathway. The premise is that the  association of 
increased 
catheter related infections with valves is likely a  brand specific problem and 
not a wholesale problem of the device category as  Drs Maki and Jarvis's data 
might have led one to believe." 
> Tony West, RN, CRNI 
> Healix, Inc. 
> Email:   [EMAIL PROTECTED] or [EMAIL PROTECTED] 
> SMS: [EMAIL PROTECTED]  
> Cell:  214-674-4848  





Reply via email to