Well that is a clue within itself. Use a can of circuit freeze very lightly on 
a few components, especially resistors and capacitors, and see if one in 
particular component quietens the noise. Sounds like you have a case of 
Johnson's noise aggravated with warm up.

Don
  ----- Original Message ----- 
  From: David C. Partridge 
  To: 'Discussion of precise voltage measurement' 
  Sent: Saturday, June 09, 2012 8:43 PM
  Subject: Re: [volt-nuts] Help with 7081 AC buffer circuit


  Argghhhh!  Changing the pot didn't fix it - once the meter had warmed up for 
> 1hr it was backto its old tricks. 

  Dave
  -----Original Message-----
  From: [email protected] [mailto:[email protected]] On 
Behalf Of David C. Partridge
  Sent: 09 June 2012 19:02
  To: 'Discussion of precise voltage measurement'
  Subject: Re: [volt-nuts] Help with 7081 AC buffer circuit

  Changing the potentiometer was indeed the solution - thanks to Ed and to John 
(whose post never made it here, though I found it on the site). 

  The DC level at TP705 is now relatively stable and close to 0uV (jumping 
around a bit but typically well less than 100uV).

  Now that I get a stable reading near 0uV here, I see another problem :(

  If I then switch from the VAC to the VAC plus VDC setting, the DC output 
level at TP705 jumps positive by about 600uV, which results in a difference in 
reading of around 60uV which I don't believe should happen.   Strangely, if I 
leave the input open circuit, I can switch between VAC and VAC+VDC modes 
without this shift occurring.

  I re-wired the circuitry around the relays using PTFE feed-through insulators 
and direct point to point wiring to eliminate leakage problems between the 
relay power lines (-10V, -15V) and the signals before I baked the PCB.

  Regards,
  David Partridge


  _______________________________________________
  volt-nuts mailing list -- [email protected]
  To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
  and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to