I made it through the DC Cal steps this evening and I was reading about the Ohms calibration.
What resistance standards are needed? 10000 ohms and 1000 ohms? 4 wire or 2? Thanks, Joe -----Original Message----- From: [email protected] [mailto:[email protected]] On Behalf Of Jean-Louis Noel Sent: Friday, June 22, 2012 9:25 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Calibration Sources for Solartron 7081 Hi J. L. From: "J. L. Trantham" <[email protected]> > However, AC promises to be a problem. I guess I will try using a > signal generator and an amplifier to see what output level and > stability I can achieve. You have to link the amplifier to your DC standard to have the stability needed. > But still, what frequency? 60 Hz? 1 kHz? The 'basic' calibration is usually performed at 400Hz. With check at higher frequencies (40KHz and 400KHz for the 7081) to know if it's still properly compensated. Bye, Jean-Louis _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.