I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V. It's quite stable even without the heater.
----- Original Message -----
From: "Rob Klein" <[email protected]>
To: "Discussion of precise voltage measurement" <[email protected]>
Sent: Friday, June 29, 2012 6:29 AM
Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard
Op 29-6-2012 6:34, Tom Leedy schreef:
Has anyone actually measured the internal temperature of the oven (in
degrees
C)? Since the operating temperature of the 732A is spec'd at 40 C, I
suspect
that it is about 45-50 C, or so.
According to the manual it's 48 ± 2°C
And I also suspect that temperature stability
of the oven is more important than absolute accuracy.
You'd be correct :-)
I didn't expect the large outpouring of knowledge!
*Nobody* expects the large outpouring of knowledge! (with apologies to
M. Python)
Rob.