How long had the meter been powered on for when you started the test?  Think 
days rather than hours to stabilise. 

D.
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf 
Of m k
Sent: 20 January 2013 09:06
To: [email protected]
Subject: [volt-nuts] 7081 resistance accuracy issue


Hi everyone, 

As part of my checking over of the behaviour of the 7081 I ran "mode,testk" and 
"nines=8" overnight, the result is a swing larger than I would expect!

max 166.00455
min 166.00040
this means a 25 ppm shift 

The circuit is meant to generate a constant current, so if that constant 
current can vary by 25ppm over a 20-23 degree house temperature then it is only 
accurate to 25ppm for resistance measurement except when used as a transfer 
standard.

anyone else find a similar shift?

M K 
                                          
_______________________________________________
volt-nuts mailing list -- [email protected] To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.