In message <[email protected]>, Frank Stellmach writes:
>The only unsolved problem then would again be the range transfer, which >is 
>not yet possible by cryogenic standards, afaik.
 
Change the frequency ?
 
-- Poul-Henning Kamp
| UNIX since Zilog Zeus 3.20 [email protected] | TCP/IP since RFC 956 FreeBSD 
committer | BSD since 4.3-tahoe Never attribute to malice what can adequately 
be explained by incompetence.
Poul-Henning,no that won't do it, what I meant to say:The range transfer is 
transfer, or measuring of voltages of 100V and 1000V range.Progammable JJ 
arrays deliver voltages in the range +/-12V, not much more.Therefore, 
increasing the driving frequency to 750Ghz or 7.5THz will not work, also 
serialization of 10 or 100 arrays is not feasible.So you have to divide 1kV or 
100V by 100 and 10, which has to be done by normal resistive dividers, Hammon 
type, as cryogenic / superconducting ones for such voltages do not 
exist.Therefore, that voltage range is still limited to 0.5 and 0.2ppm 
uncertainty.As the uncertainty of comparison is limited to around 10nV, the 
measurement on the 1V and 100mV ranges also are uncertain to 0.01 and 0.1ppm, 
respectively.Even if more care and effort is spent on the Null instrument for 
those low volt ranges, perhaps to 1nV uncertainty, there is still a big gap to 
a real 9 digit DMM, with complete covering of ranges and modes.Frank

---
Alle Postfächer an einem Ort. Jetzt wechseln und E-Mail-Adresse mitnehmen! 
Rundum glücklich mit freenetMail
_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to