Ken wrote:

R6 slider to U1 varies from +15v to -15v but the output of U1 only
varies by mv's, I put this down to the loading of R9 to earth on the input
to pin3 of U1 (221 ohm),only allowing a small variation to the input on pin
3 of U1, maybe this needs further investigation!

Look at the circuit. The 100k potentiometer (R6) feeds a voltage divider composed of R8 and R9 (464k and 221 ohms, respectively). The tap of this divider feeds the reference input of U1. R8 and R9 attenuate the +/- 15v range of R6 by a factor of 210, so the divider tap voltage (i.e., the reference voltage for U1) only varies +/- 7.14 mV from ground for the full range of R6.

U1 is an integrator, so it has lots of gain at DC (Q38 switches the integrating capacitor to give two time constants). If one of these capacitors is leaky or shorted (or there is another leakage path from Pin 6 to Pin 2 of U1), it would reduce the gain of the integrator and could produce the symptom you are having.

Best regards,

Charles