Ben, Do you have the 4921 Current Shunt Adapter option for the 4920? I have not found any information on it other than what is listed in the 4920 user's handbook. It mentions the unit simulates the impedance of a 540B. The guide says that the instructions are included with the 4921. Do you happen to have a copy of that document?
The 4920 user's guide is on the KO4BB website. Thanks, Todd On Thu, Jul 10, 2014 at 5:17 AM, ben <[email protected]> wrote: > Hello, For my 5cents worth, I also found Fluke 8506A's to be satisfactory > - so long as they work. I only have one good remaining unit that does AC > volts as well as DC. I like the HP 3458A for AC volts, but always need to > apply corrections for serious AC volts work. A better solution I now use > is the Datron 4920 units which are an order of magnitude more accurate than > the 3458A across their full bandwidth. However, be aware of their input > impedance and selection of cables - always use short coaxial. The 4920 > impedance is around 100k to 400kohm (mostly frequency dependant) and this > also varies during its sampling period. Otherwise they are great units, in > the same ball park as a Fluke 5790 but way cheaper. For an AC source I > thoroughly recommend the Datron 4200 or 4700 series, which are also way > cheaper than the near equivalent Fluke 5700. > Ben > > > > > _______________________________________________ volt-nuts mailing list -- [email protected] To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.