Mickle writes:

Solartron 7071/7081 manuals don't have official information about the
Only one of the docs - Instruments Test Specification - have a 10 V noise 
limits:
"10 V noise - The rms noise of 10 readings at 10 V is measured. It must be less
than 3 uV". This is roughly equal to 12 uV p-p (with p=0.998, f=10).
My old measurements (7.5-digits, short, 5-10 minutes):
0.1 V = 0.2 uV
10 V = 5 uV

Correct, not quite sure what I was thinking there. Re-reading what the OP said, 
looked like he was measuring the p-p reading over 10 mins. The posted values I 
gave were 'transfer standard' worst-case stability values at full scale, which 
of course isn't noise, but is a passable average over the specified transfer 
time limit.
An averaged measurement over 10 mins should be pretty close to within those 
values, and of course you wouldn't make a precision measurement without 
averaging multiple samples.

I actually have a 7081, but that shouldn't matter for the noise measurement 
(except possibly for a selected zener). My measured 7.5 digit p-p over 5 mins 
was 3uv (+2, -1). I've noticed that the average reading variation is usually in 
that range, but looking at a histogram of a few hundred readings, there are 
certainly occasional significant glitches of multiple microvolts. It's really 
hard to determine the source of those kinds of glitches, which is why when I'm 
trying to make the best measurements I can, I capture the samples and process 
them myself, throwing our any outliers.

I also decided to check calibration, because I haven't recalibrated in over a 
year. I use dual Datron 4910's that I compare to each other, sending one out 
for Fluke calibration if they disagree by more than 2 ppm. I used to pay for 
hot-shipping, but Fluke said not to bother.

Anyway, my 7081 has only drifted by 5ppm. I'm impressed.

Question - did you find the switch to the LTC1052 to make a significant 
difference? That's an easy one to do, and I always have some on hand.

I've considered making the LTZ1000 replacement you did, since I have a spare 
ref board from a 3458A, but given the stability of my meter, not sure if this 
is really worth the effort. Any opinion?


--
Bill Ezell
----------