Ken,

Coincidently, I just recalibrated a pair of 3456A meters I have been running 
24/7 for over a year.

10 V and 1 V ranges were no problem with adjustment pots having very good 
resolution.

1 K, 10 K, and 100 K pots were a little more fiddly.  I used 10 PLC averages 
for setting the pots, and then 100 PLC to re-check the setting was within 1 
count.  Reading stability on these ranges with 100 PLC was 1 count, or less.

-- Chuck


-----Original Message-----
From: volt-nuts [mailto:[email protected]] On Behalf Of Ken Goodhew
Sent: Thursday, April 30, 2015 1:52 PM
To: [email protected]
Subject: [volt-nuts] hp 3456a ohms calibration

I have been trying to check the calibration of my HP3456A multimeter.

I can check and set the DC volts calibration ok but when I try and set the ohms 
calibration using standard resistors I cannot get a steady reading as the 
display varies by several digits regardless of the value of the resistance 
being used to check the meter.

I have already changed most of the power supply filter capacitors to see if 
this solved the problem, which it did not.

The only capacitors I have not changed are the two larger ones that tested ok 
for capacitance and esr.

Any suggestions as what the problem may be or where I am going wrong in my 
calibration attempts?

Thanks,

Ken Goodhew.

VK2TKE.

_______________________________________________
volt-nuts mailing list -- [email protected] To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

_______________________________________________
volt-nuts mailing list -- [email protected]