I used a Fluke 5100B and 8920A for my SCAL calibration. I got lucky and it 
passed at Agilent. I think if you use a 3325A , you should get results that are 
close.

Todd

Sent from my iPad

> On May 20, 2015, at 15:46, Electronics and Books via volt-nuts 
> <volt-nuts@febo.com> wrote:
> 
> It is stated that the ac calibration of a 3458 is so difficult. at least the 
> equipment mentioned is so difficult to obtain. but look what you need. a hp 
> signal generator 3325 where i have a pile of (any other good generator will 
> do to), an other hp 3458 and 3 different thermocouples. very expensive 
> nowhere to be found and very delicate. hp not procduced them for 40 years or 
> ballantine which cost more then the meter your calibrating.
> but now for the accuracy. the signal generator has only 3 digits of amplitude 
> resolution. if you see the accuracy in the hp protocol what to measure it is 
> not possible to generate that.
> and for the thermocouples. they give only millivolts. you need an other 3458 
> with no offset or better a nanovolt meter to measure them.
> a far easier solution is a hp powersensor 8482a on a hp power meter. it is a 
> thermocouple of the correct frequency range and can be measured accurate 
> enough for the generator.
> the ac calibation is not an exact value as the volt or resistor but only a 
> deviation in the frequency range. it is even difficult to obtain the 
> deviation in 3 digits.
> ok it took me an hour but then i was done.
> 
> 
> Regards
> 
> Frans
> 
> 
> i...@electronicsandbooks.com
> Netherlands
> 
> Discere ne cesses
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Reply via email to